I am afraid something has gone wrong with this run. Either specific index cycles failed (where you see the . ) or more likely (hate to say this) your libraries may have failed. Either scenario would make this data unusable.
Are you sure the adapters you used were good and/or there was no issue with the run itself (either hardware or software related)? You could submit a ticket to Illumina tech support. You seem to have access to full run folder so they will help you diagnose the problem. Even if you don't have a maintenance contract the diagnosis should be covered under regular tech support.
The file you attached above has an odd format that I have not seen before. Contact Illumina tech support and see what they have to say.
Are you sure the adapters you used were good and/or there was no issue with the run itself (either hardware or software related)? You could submit a ticket to Illumina tech support. You seem to have access to full run folder so they will help you diagnose the problem. Even if you don't have a maintenance contract the diagnosis should be covered under regular tech support.
The file you attached above has an odd format that I have not seen before. Contact Illumina tech support and see what they have to say.