  • Agilent TapeStation vs. Qubit: RNA concentration.

    SCENARIO:

    After RNA extraction (RNA-seq is my aim), I check the concentration of the extracted RNA with Qubit, and then run the extracted RNA on Agilent TapeStation 2200 with HSRNA ScreenTape. The concentrations are listed below and the RINe values from the TapeStation are shown in the picture.

    Qubit concentrations vs. TapeStation concentrations:
    A1 = 14.9 ng/uL vs. 40.6 ng/uL
    B1 = 19.4 ng/uL vs. 60.2 ng/uL
    C1 = 24.4 ng/uL vs. 135 ng/uL
    D1 = 21.5 ng/uL vs. 110 ng/uL
    E1 = 32.3 ng/uL vs. ?

    QUESTIONS:

    1. Which concentration values should I trust: Qubit or TapeStation?

    2. Should I trust the RINe values of the TapeStation, even though the error message says "RNA concentration outside the recommended range for RINe", or should I dilute the sample according to the concentration given by the TapeStation and run them again?

    3. I've been told to use samples with approximately the same concentration on the TapeStation because the bad concentration of one sample (either too high or too low) can affect the results of the other samples. Is that true? How can one sample affect the result of another?

  • #2
    Originally posted by evobio View Post
    SCENARIO:

    After RNA extraction (RNA-seq is my aim), I check the concentration of the extracted RNA with Qubit, and then run the extracted RNA on Agilent TapeStation 2200 with HSRNA ScreenTape. The concentrations are listed below and the RINe values from the TapeStation are shown in the picture.

    Qubit concentrations vs. TapeStation concentrations:
    A1 = 14.9 ng/uL vs. 40.6 ng/uL
    B1 = 19.4 ng/uL vs. 60.2 ng/uL
    C1 = 24.4 ng/uL vs. 135 ng/uL
    D1 = 21.5 ng/uL vs. 110 ng/uL
    E1 = 32.3 ng/uL vs. ?

    QUESTIONS:

    1. Which concentration values should I trust: Qubit or TapeStation?
    Probably Qubit. But which dye were you using? As long as you were using the RNA-specific dye, then trust the Qubit.
    Originally posted by evobio View Post
    2. Should I trust the RINe values of the TapeStation, even though the error message says "RNA concentration outside the recommended range for RINe", or should I dilute the sample according to the concentration given by the TapeStation and run them again?
    I don't know. Do you want to for some reason? What will you do with the RIN score?
    Originally posted by evobio View Post
    3. I've been told to use samples with approximately the same concentration on the TapeStation because the bad concentration of one sample (either too high or too low) can affect the results of the other samples. Is that true? How can one sample affect the result of another?
    Seems crazy to me, but there are crazy aspects to the Agilent Bioanalyzer. So maybe there are also goofy issues with the TapeStation as well.

    --
    Phillip

    Comment


    • #3
      1. I am using the appropriate Qubit dye (Qubit RNA HS Assay Kit).

      2. The extracted RNA will be used in the preparation of RNA-seq libraries, so I'm using the RINe scores from the TapeStation as a proxy of the quality of the RNA. So I guess I need to run the samples again with the new concentrations in order to trust completely the RINe scores of the TapeStation.

      3. To me that doesn't make sense, because that means one bad sample can give you bad scores for all other samples (even if they are good). If that really is the case, each run should have only one sample, which is absurd.
      Last edited by evobio; 08-03-2015, 03:25 AM.

      Comment


      • #4
        2. Okay.
        3. Actually, Agilent has decent tech support. Maybe give them a call about this issue. See if it is true...

        --
        Phillip

        Comment

        Latest Articles

        Collapse

        • seqadmin
          How RNA-Seq is Transforming Cancer Studies
          by seqadmin



          Cancer research has been transformed through numerous molecular techniques, with RNA sequencing (RNA-seq) playing a crucial role in understanding the complexity of the disease. Maša Ivin, Ph.D., Scientific Writer at Lexogen, and Yvonne Goepel Ph.D., Product Manager at Lexogen, remarked that “The high-throughput nature of RNA-seq allows for rapid profiling and deep exploration of the transcriptome.” They emphasized its indispensable role in cancer research, aiding in biomarker...
          09-07-2023, 11:15 PM
        • seqadmin
          Methods for Investigating the Transcriptome
          by seqadmin




          Ribonucleic acid (RNA) represents a range of diverse molecules that play a crucial role in many cellular processes. From serving as a protein template to regulating genes, the complex processes involving RNA make it a focal point of study for many scientists. This article will spotlight various methods scientists have developed to investigate different RNA subtypes and the broader transcriptome.

          Whole Transcriptome RNA-seq
          Whole transcriptome sequencing...
          08-31-2023, 11:07 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by seqadmin, Yesterday, 09:05 AM
        0 responses
        12 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 09-21-2023, 06:18 AM
        0 responses
        10 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 09-20-2023, 09:17 AM
        0 responses
        12 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 09-19-2023, 09:23 AM
        0 responses
        26 views
        0 likes
        Last Post seqadmin  
        Working...
        X