new posts

  • hlu
    replied
    Originally posted by boss_hoss View Post
    hlu,

    what do you mean by primer dimers in the sequencing reaction? wouldnt these be removed by SPRI during LC?

    Sometimes, the content of reads were very short, 40bp, mostly formed from a pair of primer sequences.

    This happened in the past in GS-FLX, and GS-20 times.

    Leave a comment:


  • boss_hoss
    replied
    hlu,

    what do you mean by primer dimers in the sequencing reaction? wouldnt these be removed by SPRI during LC?

    Leave a comment:


  • lionline
    replied
    Hi Bia,

    Thanks for sharing your emPCR experience. We got problem of too low % filter passed reads -- about 25%. One question, were the % filter passed reads differ or similar between your results from Large vs. Small emPCR?

    Leave a comment:


  • bia
    replied
    Hi everybody,
    we recently upgraded to the Titanium system. What we observed is that the breaking step of the Large Volume emPCR could be tricky.
    Sometimes we found some oil above the beads pellet (usually more visible at the end of the procedure when the beads are transferred to a 1,5 mL tube).
    In the Small Volume preparation (where the breaking is performed by using the filters), performed to titrate the library, we didn't find oil on the beads.
    Both the beads from the Small and the Large preparation were sequenced and we obtained good results for the "Small" (expected number of reads, median length 450 bp) and very bad results for the "Large" ones (40-50% of the expected reads, median length 260 bp).
    The output graph of the "oily sample" was very similar to that obtained by Andeas, while for the other beads it was as expected.
    As the library used for both preparations was the same, we hypothesize the oil could hamper something during the enrichment steps or during the sequencing reaction.

    Leave a comment:


  • Tom Bair
    replied
    Following hlu's suggestion (thanks did not realize the sff file had complete reads) I dug into a run that had a similar profile to andpet's

    In this case the length is there it is just low quality, ie the untrimmed lengths have a mode right around 500.

    Looking at some of the trimmed sequences they are trimmed for a reason, low quality scores are evident. The question still is though why do we get reads with low quality?

    Anybody know how to dig out the signals for a particular read, I guess I have the pixel coordinates ...
    Attached Files

    Leave a comment:


  • joa_ds
    replied
    well, we did amplicon sequencing and had a pcr reaction prior to the emPCR, which of course amplified problems that were already there.

    We added a purification step prior to the emPCR. I don't know the details, as I am not in the lab, but i know they are using some sort of chromatographic technique to remove short sequences 50-100bp long. This simple technique improved our mapping efficiency from around 60% to around 95%.

    We add MID tags and detection of an MID is also something we use as a quality check to verify if sequences are ok. We expect very short sequences to have a reverse complement of the MID at the end of a sequence.

    Another thing we do is map and check if the mapped sequence is just the primersequence, they appear ok at first sight, but something is wrong of course then. After using the chromatography, and doing some minor tweaking, the distribution looks like what it should be

    Leave a comment:


  • Tom Bair
    replied
    We are seeing some similar runs, intermittent.

    Something else we tried to eliminate library prep issues was to do a filter template and in <qualityFilter> add a <doPrimerTrimming>false</doPrimerTrimming>


    This will allow you to see the B-adaptor sequence if present. We would expect library issues to present B-adaptors on short sequences.

    joa_ds what did you do for PCR optimization?

    Leave a comment:


  • hlu
    replied
    This looks like a shape of poor Titanium run.

    For good Titanium shotgun runs, the peak length usually at 500bp, and average length is around 400bp. Short length regions may have a hump, but should not have a big peak. In other word, the length distribution should be mainly one peak around 500bp.

    Besides mapping method suggested above, I would say use output of quality filter metrics of 454 pipeline to make judgement on the run.

    Run metrics prints out lots of information, particuarlly useful is those metrics of raw reads, filtered reads, dots, mixed etc breakdown. Use sffinfo (offline package 454 software) to see whether the short reads was trimmed by quality filter. Just run" sffinfo sffile" you will get all filtered reads information with non-trimmed full length there.

    Another way to easy tell something is wrong or not is to check reads count. Titanium should generate 1 million plus per run filtered reads passing quality filter. If you don't get this number of reads per run, then lots of raw reads were rejected (and trimmed) by quality filter, meaning something is wrong on library, EMPCR or whatever upstream procedure.

    If the short reads were not result of trimming and rejecting by quality filter, then the story may change. The sample may have massive amount of short reads (primer dimers etc) and 454 machine would favor those short fragement and sequencing them in junk amount, in a high quality manner.

    Leave a comment:


  • joa_ds
    replied
    we had the same profile doing amplicon sequencing and the amplicons were selected to be 200-300 bp (it was before we had the titanium upgrade).

    After further analysing the results we discovered that the graph profile in fact was a combination of a peak around 60 and the normal peak around 200-300. The 60bp peak was caused by priming dimers and priming mismatches. After removing everything from the dataset that had 2 short priming matches, the profile looked ok.

    In a later run the pcr was optimised and there were 95% less priming mismatches etc and the profile was ok then.

    When i look at your graph, it looks like 95% the same to the graph we had... So i guess there must be something wrong with the emulsion PCR? Are there primer contaminations in the room or something else that would cause priming dimers to appear?

    Are you doing de novo sequencing or just a resequencing experiment to validate the techniques? It might be useful to map the ok reads on an existing genome and see if the 60bp peak has a different 'mapping behaviour'. We have discovered that the down stream analysis can greatly improve the upfront library preparation. It takes some time to analyse your data, or write some scripts to automate it. But future experiments can be QC-ed in 1h and you can always use the results to optimize the upfront work.

    Leave a comment:


  • andpet
    replied
    Shotgun sequencing. We tried to sequence three insects and all runs showed the same read profile.

    We always measure the size distribution of our DNA before sequencing (Agilent Chip) and all three cases showed good results (500 - 2000 bp).

    Andreas

    Leave a comment:


  • joa_ds
    replied
    what type of sequencing was it? amplicon or shotgun?

    Leave a comment:


  • andpet
    replied
    Titanium kit - short reads

    Okay here is the attached sequence length histogram ...

    ANdreas
    Attached Files

    Leave a comment:


  • ECO
    replied
    Andreas, try attaching it to the post, rather than using the [IMG] tags. As far as I know one cannot in-line display images with FTP.

    Leave a comment:


  • andpet
    replied
    Hmm,

    insert picture does not work with ftp ... why not ?

    Andreas

    Leave a comment:


  • andpet
    started a topic Titanium kit - short reads

    Titanium kit - short reads

    Hi,

    we upgraded to 454 Titanium in january 2009 and the first runs seemed to be fine (although we see some small decline in average read length).

    Now we startet a new Titanium sequencing kit and all runs show massive amounts of short reads. The read length is almost equal distributed among the reads.

    I have attached a histogram of the read lengths:

    [IMG]ftp://genome.imb-jena.de/pub/andpet/titanium_read_length_histo.jpg[/IMG]


    Has anyone an idea about this or experienced the same before ?

    Thanks a lot,

    Andreas

Latest Articles

Collapse

  • seqadmin
    An Introduction to the Technologies Transforming Precision Medicine
    by seqadmin


    In recent years, precision medicine has become a major focus for researchers and healthcare professionals. This approach offers personalized treatment and wellness plans by utilizing insights from each person's unique biology and lifestyle to deliver more effective care. Its advancement relies on innovative technologies that enable a deeper understanding of individual variability. In a joint documentary with our colleagues at Biocompare, we examined the foundational principles of precision...
    01-27-2025, 07:46 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by seqadmin, Yesterday, 10:34 AM
0 responses
14 views
0 likes
Last Post seqadmin  
Started by seqadmin, 02-03-2025, 09:07 AM
0 responses
19 views
0 likes
Last Post seqadmin  
Started by seqadmin, 01-31-2025, 08:31 AM
0 responses
32 views
0 likes
Last Post seqadmin  
Started by seqadmin, 01-24-2025, 07:35 AM
0 responses
81 views
0 likes
Last Post seqadmin  
Working...
X