  • ABI 3130xl 50 cm Array NOS Lifetime

    How long is an ABI 3130xl 50 cm array if its never been opened and/or used? In other words, what's its shelf-life?

  • #2
    One should remember, that ABI capillary array is made of glass capillaries coated with polymer on the outside, so provided they are kept clean from POP7 on the inside, they may outlast the electrolytic capacitors inside the instrument in storage.... (I would say decades for shelf life).
    Obviously the polymer coating on the outside may degrade/crumble first (especially if stored in a cupboard, that is emitting formaldehyde vapours from cheap particle board).

    When changing the capillary array, make sure to perform the water wash before removal of the old array, followed by capillary rinse with MiliQ water
    (Manual control->Polymer delivery pupm -> fill 80cm array), and do this fill for 20 - 200 times:

    https://genomics.lsu.edu/genomics-sa...ay_longevity-1

    Make sure to store old array ONLY in water filled state!
    PS: I do hope that some after-market manufacturer(s) starts making compatible arrays for ABI machines.

    Comment


    • #3
      Originally posted by Markiyan View Post
      One should remember, that ABI capillary array is made of glass capillaries coated with polymer on the outside, so provided they are kept clean from POP7 on the inside, they may outlast the electrolytic capacitors inside the instrument in storage.... (I would say decades for shelf life).
      Obviously the polymer coating on the outside may degrade/crumble first (especially if stored in a cupboard, that is emitting formaldehyde vapours from cheap particle board).

      When changing the capillary array, make sure to perform the water wash before removal of the old array, followed by capillary rinse with MiliQ water
      (Manual control->Polymer delivery pupm -> fill 80cm array), and do this fill for 20 - 200 times:

      https://genomics.lsu.edu/genomics-sa...ay_longevity-1

      Make sure to store old array ONLY in water filled state!
      PS: I do hope that some after-market manufacturer(s) starts making compatible arrays for ABI machines.
      Thanks for the advice!

      Comment

      Latest Articles

      Collapse

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by seqadmin, 05-26-2023, 09:22 AM
      0 responses
      8 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 05-24-2023, 09:49 AM
      0 responses
      12 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 05-23-2023, 07:14 AM
      0 responses
      30 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 05-18-2023, 11:36 AM
      0 responses
      115 views
      0 likes
      Last Post seqadmin  
      Working...
      X