We spiked in a good 20% freshly diluted PhiX library, as our library's first 6 bases are the same due to restriction enzyme processing.
As for your question 1, I'd like to know how do you do your library quantification? We used KAPA's library quantification kits. You might want to do it the old school way, clone the library into T-vector or something and Sanger sequence it to see if the P5 and P7 are sequence correct and in good shape. I suspect your P5 and P7s are not working properly.
as for question 2, again, would like to know if it's possible for you to Sanger sequence your library (let's say 50 clones would be good.). Illumina's P5 and P7 sequences are actually close to the annealing sites for the index primers. I suspect your P5 and P7 sequences and nearby sequences are off, hence the failure for the indexing primer to anneal and subsequently you get dark spots on the cluster (a.k.a. GGGGGG)
Hope this helps! Good luck and may the bases be with you!