    Are my DNA extracts pure enough?

    Hello,

    I have a problem getting clean DNA from my DNA extractions, I have tried multiple times and consulted the extraction kit manufacturer without any particular luck. According to the Nanodrop spectrophotometer, the yields and absorption values are as follows:

    Sample_ID; Nucleic_Acid_Conc; A260; A280; 260/280; 260/230
    1; 7.6 ng/µl; 0.152; 0.086; 1.76; 0.71
    2; 24.0 ng/µl; 0.479; 0.249; 1.93; 1.16
    3; 21.9 ng/µl; 0.439; 0.238; 1.84; 1.15
    4; 18.6 ng/µl; 0.371; 0.190; 1.95; 1.29
    5; 18.5 ng/µl; 0.371; 0.200; 1.85; 1.44
    6; 6.9 ng/µl; 0.138; 0.069; 2.00; 0.70
    7; 7.9 ng/µl; 0.157; 0.092; 1.72; 0.78
    8; 35.2 ng/µl; 0.703; 0.363; 1.94; 1.67
    9; 9.6 ng/µl; 0.192; 0.099; 1.94; 0.85
    extraction_blank; 1.6 ng/µl; 0.031; 0.006; 4.84; 0.19
    collection_blank; 3.9 ng/µl; 0.078; 0.036; 2.19; 0.42

    Settings at "DNA", factor 50. Total extraction volume 100 µl, so total DNA yields are the concentrations multiplied by 100.

    The question now is whether this is safe to proceed with. According to ThermoFisher "good" values are between 1.8-2.0 for the 260/280 ratio, and 2.0-2.2 for the 260/230 ratio. My samples seem generally ok for the 260/280 ratios (exluding the blanks), but the 260/230 ratios are consistently lower by quite a bit. Is this a problem? I want to do DNA metabarcoding through Illumina MiSeq sequencing of a couple of marker genes. This means I will do PCR following the DNA extraction, so on the one hand I won't need so much DNA extract. On the other hand, the PCR may be the critical step, where contaminants can interfere. Maybe it even risks spoiling the sequencing run? What are your opinions? How do you treat your own samples based on spectrophotometer values?
    Last edited by hjarnek; 01-08-2024, 02:54 AM.

Latest Articles

Collapse

  • SEQadmin2
    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
    by SEQadmin2



    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
    ...
    10-14-2026, 09:55 AM
  • SEQadmin2
    New Genomics Technologies Take Aim at Long-Standing Limits
    by SEQadmin2


    Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

    We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
    ...
    09-28-2026, 10:25 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, Yesterday, 10:52 AM
0 responses
9 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-14-2026, 05:39 AM
0 responses
19 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-13-2026, 09:56 AM
0 responses
16 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-09-2026, 11:14 AM
0 responses
27 views
0 reactions
Last Post SEQadmin2  
Working...