    Problems with displaying reads in SAM format in IGV - Integrative Genome Viewer

    Hi!
    I have my doubts about the IGV displaying reads in sam-format correctly.

    The reads in sam-format all have the same length (90) but appear like being of different lengths when I upload them in IGV. The same happend with the same file in BAM format.
    I converted the SAM-file to a BED-file using Pyicos. Uploading this BED-file results into a correct display of the reads. See uploaded SAM-file (upper track) and uploaded BED-file (lower track) in attachment.
    But when BED is displayed, I can obviously not see the actual sequences of the reads (to look for polymorphisms).

    I am using the latest version of IGV: IGV_2.2.5

    Did anyone see this issue as well and knows a way around it?

    Below I give the SAM-file and the BED-file, as well as the command for the conversion.
    Any help is highly appreciated!
    Thanks!
    Sonja






    # here the command to convert sam to bed:
    pyicos convert input.sam output.bed -f sam -F bed

    # here the SAM-file
    ERR045703.5732207 99 19 16149684 60 75M15S = 16150110 508 AACGGGCACCCAGTGAGCACTCGAGGATGACCCTCCTCGGGCAGCTGCCGCCCACCCAGTAGCGACTGTCCCCAAGTCAGCAGGGAGGGA B5@9FGGIFHHIIEFAFFHHKI@IKJIHG@EJJKIJKI?GGB>CGBHGJ@HEABAD?B@DCDC898?>GCGGBH################ X0:i:1 X1:i:0 XC:i:75 MD:Z:75 RG:Z:ERR045703 AM:i:37 NM:i:0 SM:i:37 MQ:i:60 XT:A:U BQ:Z:@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@
    ERR045704.58587335 163 19 16149690 60 37M53S = 16150123 494 CACCCAGGGAGCACTCGAGGATGACCCTCCTCGGGCAGCTGCCGCCCACCCTGTAGCGACTGTCCCCAAGTGAGCAGGGAGGGAAGCGAG @EC3<B=',EF/GBFE:7E<C->*416@A@8B;+$@###################################################### X0:i:1 X1:i:0 XC:i:37 MD:Z:7T29 RG:Z:ERR045704 AM:i:37 NM:i:1 SM:i:37 MQ:i:60 XT:A:U BQ:Z:@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@
    ERR045703.17024081 147 19 16149699 60 8S82M = 16149269 -511 ACCCAGTGAGCACTCGAGGATGACCCTCCTCGGGCAGCTGCCGCCCACCCAGTAGCGACTGTCCCCAAGTCAGCAGGGAGGGAAGAGAGC #########A=??=AFGD>?@37E=?9=:=4@@BBC=:=DA?CGADADGJKHJKJAFJJJFBGJGCGHEKIJIIGHHAHGGFGHDGCFC@ X0:i:1 X1:i:0 XC:i:82 MD:Z:82 RG:Z:ERR045703 AM:i:37 NM:i:0 SM:i:37 MQ:i:60 XT:A:U BQ:Z:@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@
    ERR045704.7021863 147 19 16149705 60 7S83M = 16149294 -493 GACCACTCGAGGATGCCCCTCCTCGGGCAGCTGCCGCCCACCCAGTAGCGACTGTCCCCAAGTCAGCAGGGAGGGAAGAGAGCAGGTCAC ########EEHGE?E?D<EFD>EEFDGEDFE?<BF?BFGIJKEIGG>IJJIGFHKHHFKHLJKJJIIIIJGIIJJHGGHGGDGCFEH@ X0:i:1 X1:i:0 XC:i:83 MD:Z:8A74 RG:Z:ERR045704 AM:i:37 NM:i:1 SM:i:37 MQ:i:60 XT:A:U BQ:Z:@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@
    ERR045704.32772245 163 19 16149726 60 90M = 16150187 499 AGCTGCCGCCCACCCAGTAGCGACTGTCCCCAAGTCAGCAGGGAGGGAAGAGAGCAGGTCACGCTCTCCTAAGTCTGATCAAGCAGCCGT @EDFFFG?GEHIFGHIJHHIH@FFKJIHJJJJLIEFIKJKJKI?FHHJIFJEGEEGFCECHB=EEDFEFHFHHGFIH,:CGIHH>@EC<F X0:i:1 X1:i:0 MD:Z:90 RG:Z:ERR045704 AM:i:37 NM:i:0 SM:i:37 MQ:i:60 XT:A:U BQ:Z:@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@A
    ERR045704.57426661 163 19 16149741 60 90M = 16150165 492 AGTAGCGACTGTCCCCAAGTCAGCAGGGAGGGAAGAGAGCAGGTCACGCTCTCCTAAGTCTGATCAAGCAGCCGTGCAGAGATGTGCCTC @EFEFF>HFHHGHGHIJ@GGGIIJJKKJGIBB??CHEJFFJHG@@DC>HIEGIJFC@FDCGFFHEIBFFGFGG=BBAEDJHHGIH;C?HD X0:i:1 X1:i:0 MD:Z:90 RG:Z:ERR045704 AM:i:37 NM:i:0 SM:i:37 MQ:i:60 XT:A:U BQ:Z:@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@BB
    ERR045703.16563758 83 19 16149748 60 90M = 16149339 -498 ACTGTCCCCAAGTCAGCAGGGAGGGAAGAGAGCAGGTCACGCTCTCCTAAGTCTGATCAAGCAGCCGTGCAGAGATGTGCCTCTCACCTA EAGDDFGEHBIFIGHHGIGGEJDIEHIHIHIDHKJHHHHAKHJKJKKJKKELKJIHJJKKGJIJJAHHAJKIJIJIGIH@HIHGCCACB6 X0:i:1 X1:i:0 MD:Z:90 RG:Z:ERR045703 AM:i:37 NM:i:0 SM:i:37 MQ:i:60 XT:A:U BQ:Z:@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@
    ERR045704.56912904 147 19 16149756 60 13S77M = 16149309 -523 TAGCGACCGTCCCCAAGTCAGCAGCGAGGGAAGAGAGCAGGTCACGCTCTCCTAAGTCTGATCAAGCAGCCGTGCAGAGATGTGCCTCTC ##############AIEEFI>>43)<H@8FA9-GA02BD:46?E7ED<.6,.>1.=ECC,97E4C8ECA2'2-'EG<9,,;BHG=?GGD; X0:i:1 X1:i:0 XC:i:77 MD:Z:11G65 RG:Z:ERR045704 AM:i:37 NM:i:1 SM:i:37 MQ:i:60 XT:A:U BQ:Z:@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@
    ERR045704.58477016 147 19 16149772 60 37S53M = 16149322 -502 CCACCCCGCAGCGACGGTCCCCAAGTCAGCAGGGAGGGAAGAGAGCAGGTCACGCTCTCCTAAGTCTGATCAAGCGGCCGTGCAGAGATG ######################################?F@?4DD9@IDG=5';GEC=(7C4HB8GHCIF<(C=&%%$4*[email protected]/ X0:i:1 X1:i:0 XC:i:53 MD:Z:38A14 RG:Z:ERR045704 AM:i:37 NM:i:1 SM:i:37 MQ:i:60 XT:A:U BQ:Z:@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@@




    # here is the BED-file
    19 16149684 16149773 ERR045703.5732207 0 +
    19 16149690 16149779 ERR045704.58587335 0 +
    19 16149699 16149788 ERR045703.17024081 0 -
    19 16149705 16149794 ERR045704.7021863 0 -
    19 16149726 16149815 ERR045704.32772245 0 +
    19 16149741 16149830 ERR045704.57426661 0 +
    19 16149748 16149837 ERR045703.16563758 0 -
    19 16149756 16149845 ERR045704.56912904 0 -
    19 16149772 16149861 ERR045704.58477016 0 -
    Attached Files
  • dpryan
    Devon Ryan
    • Jul 2011
    • 3478

    #2
    It looks like IGV is just not showing the soft-clipped regions. That would be the appropriate behaviour.

    Comment

    • sonja
      Member
      • Aug 2010
      • 10

      #3
      soft-clip

      Thanks for your answer!
      what are soft-clipped regions, please?

      Comment

      • Richard Finney
        Senior Member
        • Feb 2009
        • 701

        #4
        Can't find your 'S" ? 'S' operator documentation is here
        ...
        See CIGAR 'S' (soft clip) operation in sam/bam format documentation at sourceforge : http://samtools.sourceforge.net/SAM1.pdf

        As far as that recently added operator, don't 'X' me.

        Comment

        • sonja
          Member
          • Aug 2010
          • 10

          #5
          does it basically mean that the base quality is very low in these regions?

          Comment

          • dpryan
            Devon Ryan
            • Jul 2011
            • 3478

            #6
            Originally posted by sonja View Post
            does it basically mean that the base quality is very low in these regions?
            In the case of your reads, at least, yes. You can see that from the quality scores (often #).

            Comment

            • sonja
              Member
              • Aug 2010
              • 10

              #7
              yes, that´s exactly what I see!
              Thanks a lot!!!

              Comment

              • Jim Robinson
                Member
                • May 2009
                • 75

                #8
                You can optionally display the soft-clipped reads, its a user preference (View > Preferences > Alignments tab).

                Comment

                • sonja
                  Member
                  • Aug 2010
                  • 10

                  #9
                  great! I was looking for that... Thanks!!!

                  Comment

                  Latest Articles

                  Collapse

                  • SEQadmin2
                    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                    by SEQadmin2



                    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                    ...
                    10-14-2026, 09:55 AM
                  • SEQadmin2
                    New Genomics Technologies Take Aim at Long-Standing Limits
                    by SEQadmin2


                    Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                    We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                    ...
                    09-28-2026, 10:25 AM

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by SEQadmin2, Yesterday, 11:13 AM
                  0 responses
                  16 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, Yesterday, 10:48 AM
                  0 responses
                  15 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-15-2026, 10:52 AM
                  0 responses
                  35 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-14-2026, 05:39 AM
                  0 responses
                  42 views
                  0 reactions
                  Last Post SEQadmin2  
                  Working...