    Lower quality in reverse reads only. Is it normal?

    We did 100bp PE sequencing (RNA-Seq) on 8 samples and we see a distinct systematic drop in quality only for reverse reads (attached screenshots). We contacted the sequencing center and they said it "passed" their QC and also Illumina's spec of ~83% bases > q30. However, when we look at the % of sequences below q30, there is a remarkable difference (~7% in fwd reads vs ~24% in reverse reads).

    I wonder if this is okay. Similar problem has been discussed before in this forum or elsewhere.

    Bridged amplification & clustering followed by sequencing by synthesis. (Genome Analyzer / HiSeq / MiSeq)
    Attached Files
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    Could you please post FastQC plot and which chemistry was used.

    Comment

    • GenoMax
      Senior Member
      • Feb 2008
      • 7142

      #3
      It is "normal" (if one wants to call it that) to see reduction in Q-scores in later R2 cycles (and the difference can be similar to what you found). I assume this is MiSeq data. As the sequencing center pointed out the data technically satisfies published spec for MiSeq Q-scores.

      One reason that can lead to rapid decrease in Q-scores is inserts shorter than the read length.

      Comment

      • drramki
        Junior Member
        • Dec 2010
        • 3

        #4
        Here is the link for fastQC plots. And I suppose this is MiSeq. We have never seen this pattern in HiSeq data.

        Comment

        • nucacidhunter
          Jafar Jabbari
          • Jan 2013
          • 1250

          #5
          If the read numbers are over 30M pairs/lane, it should be from HiSeq. In any case the quality drop is more than expected in any of Illumina systems and indicates an issue most likely with sequencing reagents. As GenoMax has pointed out short inserts also can be the cause where in later cycles adapter sequences are being read. It will be interesting to know output read number and % of reads which were demultiplexed.
          Last edited by nucacidhunter; 07-31-2014, 06:55 AM.

          Comment

          • drramki
            Junior Member
            • Dec 2010
            • 3

            #6
            Thanks GenoMax and nucacidhunter.

            I was wrong when I said MiSeq. This is actually from HiSeq. We got the data twice from the sequencing center and both looked same.

            Comment

            • GenoMax
              Senior Member
              • Feb 2008
              • 7142

              #7
              This may be a "not so good" quality library, if the result has been the same two times in a row on two runs. You would want to do some additional QC/cleanup (trimmomatic, bbduk etc) to figure out what is causing this drop (adapters?).

              Also ask the provider if they had seen something similar for other lanes on those runs to eliminate possibility of technical problems with the runs.

              Comment

              • nucacidhunter
                Jafar Jabbari
                • Jan 2013
                • 1250

                #8
                In addition to above suggestions you can have a look at Adapters plot in FastQC report to see the level of adapter sequences and if that correlates with drop in quality. Also you can check size distribution of the library by examining the Bioanalyser or TapeStation trace to work out insert sizes. If high percentage of library inserts is below 100 base (fragment sizes lower than 230 bp) in this case, the reads will run into adapters causing quality issues that can be fixed by trimming. If none of those seems to be likely cause it could be the chemistry specially if it was Rapid run.

                Comment

                • N311V
                  Member
                  • Jul 2013
                  • 34

                  #9
                  With pretty much all of our HiSeq data we do see a drop-off in quality towards the end of the second read but nothing that extreme. I tend to agree with previous comments that something has gone wrong here. In my experience sequencing centres etc. have rather soft QC metrics.

                  Comment

                  Latest Articles

                  Collapse

                  • SEQadmin2
                    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                    by SEQadmin2



                    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                    ...
                    10-14-2026, 09:55 AM
                  • SEQadmin2
                    New Genomics Technologies Take Aim at Long-Standing Limits
                    by SEQadmin2


                    Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                    We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                    ...
                    09-28-2026, 10:25 AM

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by SEQadmin2, 10-15-2026, 10:52 AM
                  0 responses
                  17 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-14-2026, 05:39 AM
                  0 responses
                  24 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-13-2026, 09:56 AM
                  0 responses
                  22 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-09-2026, 11:14 AM
                  0 responses
                  27 views
                  0 reactions
                  Last Post SEQadmin2  
                  Working...