    a problem about SHRiMP2 solving AB SOLID

    Hi, everyone. I want to do the mapping about AB SOLID data. They are paired-end sequencing. The format of .fastq files show below:
    pair1:
    @SRR586064.1ugc_357_358_MatePair_2x50bp_solid0032_20100528_MP_ugc_357_854_13_97/1
    T30..01121.12.1032100213131122200031222022101302313
    +
    !AB!!?:>@<!@B!;AB?AA@@<2<@?@@<?AB>A?9?:;@@>;-=>>7=@
    pair2:
    @SRR586064.1ugc_357_358_MatePair_2x50bp_solid0032_20100528_MP_ugc_357_854_13_97/2
    G10330000122222033220201000000220002000000000000000
    +
    !@BBABBBB@>?@@(.))35.%-.3((%1+%((82-'.*-*/3*14*'696
    I try use the tool:SHRiMP2, but it is not friendly to solve my datasets. I run the command:
    gmapper-cs -1 pair1.fastq -2 pair2.fastq -L ./test/Sscro --pair-mode opp-in -Q --trim-first --qv-offset 33 >pair12.sam 2>pair12.log
    And I got the info in the end of file .log:
    note: detected fastq format in input file [pair1.fastq]
    - Processing read files [pair1.fastq , pair2.fastq]
    note: quality value format not set explicitly; using PHRED+33
    done r/hr r/core-hr
    There has been a problem reading in the read "SRR586064.1", the quality length exceeds the sequence length!
    Are you using the right executable? gmapper-cs for color space? and gmapper-ls for letter space?


    I have tried some parameters of SHRiMP2, but failed. And I can't find any similar example from SHRiMP2 website. Does anybody get the same problem and have you solve it ? Or maybe I should use another tool, any suggestion?
  • Brian Bushnell
    Super Moderator
    • Jan 2014
    • 2709

    #2
    Looks like the problem is that there are more qualities than colors. There are 50 colors (the leading 'T' or 'G' are fixed) and 51 qualities, but the first quality (!) is a placeholder. I can't remember if that's normal but I haven't worked with Solid in a while. It should be fine if you eliminate the first quality value for every quality line, and leave everything else unchanged.

    Comment

    • Lv Ray
      Member
      • Jun 2014
      • 42

      #3
      Originally posted by Brian Bushnell View Post
      Looks like the problem is that there are more qualities than colors. There are 50 colors (the leading 'T' or 'G' are fixed) and 51 qualities, but the first quality (!) is a placeholder. I can't remember if that's normal but I haven't worked with Solid in a while. It should be fine if you eliminate the first quality value for every quality line, and leave everything else unchanged.
      Thanks, Brian Bushnell. I will try your suggestion.

      Comment

      • Lv Ray
        Member
        • Jun 2014
        • 42

        #4
        Originally posted by Brian Bushnell View Post
        Looks like the problem is that there are more qualities than colors. There are 50 colors (the leading 'T' or 'G' are fixed) and 51 qualities, but the first quality (!) is a placeholder. I can't remember if that's normal but I haven't worked with Solid in a while. It should be fine if you eliminate the first quality value for every quality line, and leave everything else unchanged.
        Brian Bushnell, Thank you for your suggestion, and I have got the main point to solve it. Maybe now the core problem is to try some parameters of SHRiMP2.

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
          by SEQadmin2



          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
          ...
          10-14-2026, 09:55 AM
        • SEQadmin2
          New Genomics Technologies Take Aim at Long-Standing Limits
          by SEQadmin2


          Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

          We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
          ...
          09-28-2026, 10:25 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 10-22-2026, 11:13 AM
        0 responses
        28 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 10:48 AM
        0 responses
        20 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-15-2026, 10:52 AM
        0 responses
        38 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-14-2026, 05:39 AM
        0 responses
        44 views
        0 reactions
        Last Post SEQadmin2  
        Working...