    Oh, sorry, the command I gave you was for interleaved reads, so those are all false positive merges. For pairs in separate files, the command would be:

    bbmerge.sh in1=R1.fq in2=R2.fastq ihist=ihist.txt reads=100000

    As for whether you merge the reads before assembly, that depends on the assembler. AllPathsLG does its own merging; Ray seems to do well with merged reads; SoapDenovo creates worse assemblies with merged reads; there's no single rule. I've never used Geneious, so I don't know if it would help. As with many options, such as trimming and subsampling, sometimes the only way to get the best assembly is to try both ways.

    If you do merge the reads, I suggest using the default settings:

    bbmerge.sh in1=r1.fq in2=r2.fq out=merged.fq outu1=unmerged1.fq outu2=unmerged2.fq

    ...then feed the assembler both the merged and unmerged reads. Many or most assemblers will accept both paired and unpaired reads; merging should not be done for assemblers that don't allow you to feed them both paired and unpaired reads simultaneously, as low-complexity genomic regions will not merge as well.

    Comment

    • Marisa_Miller
      Member
      • Aug 2010
      • 34

      #17
      Originally posted by Brian Bushnell View Post
      The fraction joined and the position of the peak in the graph will make it clear what the real distribution is like. If the graph is still rising then abruptly drops to zero just before 2x(read length) then the insert sizes are generally too long for merging.
      Hi Brian,
      I re-ran bbmerge with the correct command, and it looks like the distribution shows the insert sizes are too big for merging. Although, the fraction joined is around 60% for most libraries, not sure if this is good or bad.
      Attached Files

      Comment

      • Brian Bushnell
        Super Moderator
        • Jan 2014
        • 2709

        #18


        For a 2x300bp library, getting 60% merging and a median of 400bp is pretty optimal for merging, actually. Again, whether merging is a good idea depends on the assembler, but this library is a good. You never see 90%+ merging unless the insert sizes came out way too short.
        Attached Files

        Comment

        • Marisa_Miller
          Member
          • Aug 2010
          • 34

          #19
          Originally posted by Brian Bushnell View Post


          For a 2x300bp library, getting 60% merging and a median of 400bp is pretty optimal for merging, actually. Again, whether merging is a good idea depends on the assembler, but this library is a good. You never see 90%+ merging unless the insert sizes came out way too short.
          I think I misunderstood earlier about what to look for in a library to see if it can be merged. I will go ahead and merge them and give the assembly a shot with both merged and unmerged. Thanks again!

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
            by SEQadmin2



            CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

            Despite this, “CRISPR helped turn genome editing from a specialized technique into
            ...
            07-31-2026, 11:01 AM
          • SEQadmin2
            Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
            by SEQadmin2


            Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

            The systematic characterization of the human proteome has
            ...
            07-20-2026, 11:48 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, Today, 10:05 AM
          0 responses
          8 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 08-13-2026, 12:22 PM
          0 responses
          32 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 08-11-2026, 10:35 AM
          0 responses
          27 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 08-06-2026, 07:41 AM
          0 responses
          38 views
          0 reactions
          Last Post SEQadmin2  
          Working...