    Illumina Error rate

    Hello everyone,

    That's must be a silly question but there is something I don't understand : the error rate of illumina sequencing (calculating using phix genome) is around 1%. But most of the read have a mean quality >= Q30 which is 0.1%.

    So what does the error rate actually mean ?
  • Brian Bushnell
    Super Moderator
    • Jan 2014
    • 2709

    #2
    You can't estimate error rates with a linear average of log-transformed values. If you map to phiX with BBMap, like this:

    bbmap.sh in=reads.fq ref=phix.fa mhist=mhist.txt qhist=qhist.txt qahist=qahist.txt

    ...then graph qhist with something like Excel or R, you will see the difference between the linear and logarithmic averages, as well as the actual error rate. The actual error rate will much more closely resemble the lines of the logarithmic averages.

    qahist.txt (quality accuracy histogram), on the other hand, will show the actual measured quality values for each quality score, which will tell you how accurate their quality scores are. Normally they're not too far off, but it highly depends on the specific platform and software version.

    Comment

    • ClemBuntu
      Member
      • Dec 2014
      • 37

      #3
      Sorry but I don't understand.

      What's the point of qhist histogram ?

      And about qahist, am I supposed to see that most of my reads have a quality greater or equal than 35 ?
      Attached Files

      Comment

      • Brian Bushnell
        Super Moderator
        • Jan 2014
        • 2709

        #4
        The qhist shows you, per position in the read, what the expected error rate is (that's read1_log) and what the actual average error rate is (that's read1_measured). As you can see, those completely disagree so the quality scores are very inaccurate (almost meaningless) for this library, and that the error rates are high - the average error rate starts at Q20 (1%) and drops to Q11 (8%) by the end. The X axis is read position.

        The qahist is plotted incorrectly - you need to plot the "Quality" column as the X-axis, and the "TrueQuality" column as the Y-axis; discard the other columns.

        Comment

        • ClemBuntu
          Member
          • Dec 2014
          • 37

          #5
          Originally posted by Brian Bushnell View Post
          The qhist shows you, per position in the read, what the expected error rate is (that's read1_log) and what the actual average error rate is (that's read1_measured). As you can see, those completely disagree so the quality scores are very inaccurate (almost meaningless) for this library, and that the error rates are high -
          What do you mean 'expected' ?


          I'm not sure to understand the scale.
          Originally posted by Brian Bushnell View Post
          the average error rate starts at Q20 (1%) and drops to Q11 (8%) by the end.
          So I guess the Y axe is the log quality value, 20,000 is Q20 and you say it regarding to the orange curve right ?



          Originally posted by Brian Bushnell View Post

          The qahist is plotted incorrectly - you need to plot the "Quality" column as the X-axis, and the "TrueQuality" column as the Y-axis; discard the other columns.
          I have drawn the qahist again, does it show the quality is from Q8 to Q20 as well ?
          Attached Files

          Comment

          • HESmith
            Senior Member
            • Oct 2009
            • 512

            #6
            Originally posted by ClemBuntu View Post
            What do you mean 'expected' ?
            The quality scores produced by the sequencer are expected (i.e., predicted or calculated) scores; see http://www.illumina.com/documents/pr...ity_scores.pdf for an explanation.

            Originally posted by ClemBuntu View Post
            I'm not sure to understand the scale.

            So I guess the Y axe is the log quality value, 20,000 is Q20 and you say it regarding to the orange curve right ?
            Yes, 20,000 is Q20. The relevant curves are the orange one (expected scores) vs. the gray one (actual scores based on your phiX data). These two curves should overlap if the expected scores accurately reflect the true error rate. They do not. As Brian indicated, the true error rate begins at Q20 and declines to Q11 at the end of the read. These quality scores are very low.

            Comment

            • Brian Bushnell
              Super Moderator
              • Jan 2014
              • 2709

              #7
              Originally posted by ClemBuntu View Post
              I have drawn the qahist again, does it show the quality is from Q8 to Q20 as well ?
              Each point on the qahist indicates the claimed quality (from the quality scores) versus the measured quality (based on the alignment match/mismatch rate). So, for example, you have a point at X=22, Y=13. That means that if you take all bases with a stated quality score of Q22 (roughly 99.3% accuracy), on average, they have an error rate indicating Q13 (roughly 95% accuracy).

              Comment

              • HESmith
                Senior Member
                • Oct 2009
                • 512

                #8
                Hey Brian, what's the point at (0,60)? It looks like all the Q0s are 99.9999% accurate!

                [ClemBuntu, ignore this message. It's a joke. The lowest quality produced by Illumina is Q2]

                Comment

                • Brian Bushnell
                  Super Moderator
                  • Jan 2014
                  • 2709

                  #9
                  That's actually a valid point! The quality scores vary by platform and software version (I'm guessing this is NextSeq V1 chemistry, or HiSeq 4000). Normally, for non-binned quality scores (like HiSeq 2000), there is 0 (for N), 2, then 5-41. Quite often Q2 bases are more accurate than Q5 bases, as 2 has a special meaning. But sometimes, called bases (A, C, G, T) are produced with Q0 assigned. It's normally very few, under 100. I suspect it's a bug in Casava. But, due to the fact that there are so few, it's not uncommon for them to all match the reference. To keep the axes finite, I cap all quality values at 60, but technically, in this case, the Q0 bases are 100% accurate (Q infinity). I wouldn't count on it in general, though

                  Comment

                  • HESmith
                    Senior Member
                    • Oct 2009
                    • 512

                    #10
                    Thanks, Brian. As always, you're a fount of knowledge.

                    Comment

                    Latest Articles

                    Collapse

                    • mylaser
                      Reply to Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                      by mylaser
                      The world of online gaming has grown tremendously over the past few years, giving players access to exciting sports, casino games, and interactive entertainment from the comfort of their homes. Among the platforms gaining attention, Kheloyaar has become a trusted destination for users seeking a fast, secure, and engaging gaming experience.
                      Whether you're a first-time visitor or an existing user, understanding the features of Kheloyar and the Kheloyaar login process can help you enjoy everything...
                      Today, 12:33 AM
                    • SEQadmin2
                      Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                      by SEQadmin2


                      Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                      The systematic characterization of the human proteome has
                      ...
                      Yesterday, 11:48 AM
                    • SEQadmin2
                      Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                      by SEQadmin2



                      Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                      ...
                      07-09-2026, 11:10 AM

                    ad_right_rmr

                    Collapse

                    News

                    Collapse

                    Topics Statistics Last Post
                    Started by SEQadmin2, Yesterday, 11:10 AM
                    0 responses
                    8 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 07-13-2026, 10:26 AM
                    0 responses
                    30 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 07-09-2026, 10:04 AM
                    0 responses
                    39 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 07-08-2026, 10:08 AM
                    0 responses
                    25 views
                    0 reactions
                    Last Post SEQadmin2  
                    Working...