    Funny ReadPrep1 Images GAIIx

    Hi all,

    I'm new to the SEQanswers community; from what I have seen from browsing this is an awesome resource for all those dealing with NGS. So, thank you in advance for any advice you can throw my way.

    I run an Illumina Genome Analyzer IIx, with a Paired End module. It seems everytime I start a run, another issue crops up. I started a run early this month but was unsatisfied with Q30 scores, so I stopped it. Turns out the machine needed a new green laser motor. Once installed, I ran a PhiX flow cell using a 2x33 recipe. Our Q30 metric for the entire run was above 95% which is awesome...

    So, from here I rehyb'd the original flow cell and loaded it onto the GA with full confidence. I applied the oil, checked for leaks, wiped down the peltier, the whole nine yards. The first base report was alarming however, with standard deviations above 50,000 for cluster densities. I contacted the FAS who did a share-my-desktop; he pumped the reagents manually and reran the calibration, find edge, and read prep 1. The bottom row of tiles that are imaged showed something very strange. One tile would be extremely blurry, then the tile directly adjacent to it would be clear as day. The blurry images only occurred on the bottom row, but once again, it was not every tile or even both tiles in a lane. Has anyone out there experienced this before? Of course I am in contact with the company, but any advice or hypothesis is appreciated.

    Many thanks

    JB
  • GenomicsJB
    Junior Member
    • Nov 2011
    • 2

    #2
    This problem is not novel, from what my FAS told me. It's potentially a dying auto focus laser.

    Comment

    Latest Articles

    Collapse

    • SEQadmin2
      Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
      by SEQadmin2



      Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

      Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
      ...
      10-14-2026, 09:55 AM
    • SEQadmin2
      New Genomics Technologies Take Aim at Long-Standing Limits
      by SEQadmin2


      Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

      We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
      ...
      09-28-2026, 10:25 AM

    ad_right_rmr

    Collapse

    News

    Collapse

    Topics Statistics Last Post
    Started by SEQadmin2, Yesterday, 11:13 AM
    0 responses
    19 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, Yesterday, 10:48 AM
    0 responses
    18 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 10-15-2026, 10:52 AM
    0 responses
    36 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 10-14-2026, 05:39 AM
    0 responses
    42 views
    0 reactions
    Last Post SEQadmin2  
    Working...