Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • renuka22
    Member
    • Mar 2017
    • 22

    #1

    polyA library preparation and RIN values

    Hi there,

    I am new to this forum and had some questions about the RNA bioanayzer data and nanodrop readings. I got a good concentration of 200-300ng/ul when I analysed my samples by nanodrop whereas on the bioanalyzer, the compay was not able to obtain any RIN value (Nobands obtained, not even a smear, only some clear area seen on the gel). They concluded saying all the RNA is degraded. Also, they say they are not able to make polyA library as the concentration after polyA shows only around 0.1-0.2 ng/ul and it is very low to get any sequencing results. Anyone can suggest something regarding this, coming from a totally inexperienced person..

    Thanks
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    Nanodrop QC is only good to detect contaminants not for quantification. RNA quantification can be done using Qubit or similar method and integrity by gel or Bioanalyzer-like instruments. Quantification by Bioanalyzer is Ok for most application though it is not very accurate. I tend to agree with your service provider.

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #3
      Has anyone bought a "Nanodrop One"? It is supposed to be able to analyze your spectrum of your sample and deconvolute the spectra of nucleic acids, protein, guanidine and phenol. That is, supposedly it can correct for the presence of the latter three common spectral contaminants of nucleic acid preps/methods and give you an accurate nucleic acid concentration.

      I don't imagine it will be able to distinguish between RNA and DNA, so genomic DNA prep concentrations will still usually be wrong. But for RNA preps, DNA usually isn't a major contaminant.

      If it works and renuka22 had one, it would have saved him or her some troubles...

      --
      Phillip
      Last edited by pmiguel; 05-19-2017, 11:12 AM.

      Comment

      • renuka22
        Member
        • Mar 2017
        • 22

        #4
        No, I have never used this. Would have been a good idea though..

        Comment

        • renuka22
          Member
          • Mar 2017
          • 22

          #5
          Does doing polyA treatment lowers RIN values? My polysomal RNA values showed good RIN values of 9.5..but when I did the polyA, it was in the range of 5-6. The concentration could not be detected by qubit high sensitivity kit. Is the polyA treatment degrading the samples?

          Thanks

          Comment

          • nucacidhunter
            Jafar Jabbari
            • Jan 2013
            • 1250

            #6
            RIN number is valid only for total RNA and one of the main factors in its calculation is 18S/28S ratio. With a successful polyA selection rRNA would be excluded so RIN number is not meaningful. If after polyA selection you still see 18S or 28S bands then it has been less successful. See Agilent application note about RIN number: https://www.agilent.com/cs/library/a...989-1165EN.pdf

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM
            • SEQadmin2
              New Genomics Technologies Take Aim at Long-Standing Limits
              by SEQadmin2


              Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

              We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
              ...
              09-28-2026, 10:25 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, Yesterday, 11:13 AM
            0 responses
            19 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, Yesterday, 10:48 AM
            0 responses
            18 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-15-2026, 10:52 AM
            0 responses
            36 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-14-2026, 05:39 AM
            0 responses
            42 views
            0 reactions
            Last Post SEQadmin2  
            Working...