  • Why Can't SureSelect work in TruSeq

    Hi. I'm a member of a experimental reagent company. I want to know

    "Why Can't SureSelect work in TruSeq" .

    Please tell me about this.

    Thanks in advance.

  • #2
    A key bit of the SureSelect kits are blocking oligos -- it is critical in these procedures to prevent library molecules from "daisy chaining" -- hybridizing through complementarity of their adapter sequences. Such chaining will result in pull-down of off target fragments.

    The blocking oligos are complementary to the adapter sequences. Hence, any change in the library oligos will lead to incompatibility / poor performance.

    Comment


    • #3
      It can also depend on what read length you are using. Sureselect capture is designed for ~150bp insert size, while the new Truseq reagents are designed for longer read lengths. You don't want to be doing PE100 reads on a 150bp insert sample size. The Truseq capture reagents are designed to take longer read lengths into account.

      Comment


      • #4
        Thank you for your easily comprehensive answer. I got it.

        Therfore Agilent can't sell SureSelect for Illumina Platform unless they develop new product for it.

        Comment


        • #5
          Even for Roche you have to synthesize the blocking oligos for TruSeq. They have the sequences listed on the Nimblegen web site for the blocking oligos.

          Comment


          • #6
            Originally posted by GW_OK View Post
            It can also depend on what read length you are using. Sureselect capture is designed for ~150bp insert size, while the new Truseq reagents are designed for longer read lengths. You don't want to be doing PE100 reads on a 150bp insert sample size. The Truseq capture reagents are designed to take longer read lengths into account.
            Do you (or anyone) know how reagents would be changed simply for longer insert sizes?

            If the full reason for why they aren't compatible is simply the blocking oligos, could you synthesize your own blocking oligos and then have SureSelect be compatible with TruSeq? Has anybody tried this?

            Comment


            • #7
              In my experience, the only reason this wouldn't work is that you n eed the correct blocking oligos. Length of the pull down in this range and higher doesn't seem to be much of a factor.

              Comment


              • #8
                Agilent Sureselect and Truseq

                I've done a few dozen Sureselect captures with the Illumina Truseq Prep. My on-target sequencing suffers quite a bit. My on target % went from 60-70% to 30%, unfortunately, but it still worked. I used the standard hybridization protocol without any different blocking primers, like what is recommended by Roche. For my next hybridization, I'll try the Roche blocking primers and see if it helps my on target capture.

                Comment


                • #9
                  Hi All,

                  Thanks for all the info. I was just wondering for the SureSelect capture following TruSeq library prep, do you still use all the SureSelect blocking oligos (#1-3), in addition to the ones specific for the TruSeq adaptors? I'm thinking that I might be able to leave out the blocks specific to their adaptors (instead of the TruSeq ones).

                  Thanks in advance for any insights!

                  Comment


                  • #10
                    I've done this recently and it can be modified to work great with TruSeq libraries.

                    First....Read this protocol in detail and make sure you understand every step. http://openwetware.org/wiki/Image:Geller_exome.pdf It's great. I made a small modification in that this paper uses blocking oligos that will immediately anneal to each other...not really a great strategy. I'm just blocking one strand on each adapter...eliminating any complementarity between library molecules.

                    What I've done...leave out Agilent's block #3 (the 0.6ul per capture), and add the following mixtures of blocking oligos (all at 50uM).

                    With the modified blocking oligos the PCT_SELECTED_BASES was 73%...which is within range of normal SureSelect in our hands (here our positive control (stock sureselect) was 80%.

                    I'm still working on the split blocking strategy as it's an unnecessary workflow complication to have an index specific block. I was not very surprised that this split adapter strategy didn't perform well as the Tm's are too low to be very stable in the hyb, but it was worth a shot. My second attempt will fix this issue I hope!

                    Anyway I hope this helps....let me know if you have any questions...hopefully I didn't make any mistakes or typos below!

                    Attached Files

                    Comment


                    • #11
                      Hi Eco,

                      Thanks for posting your results of your experiment testing various blocking oligo strategies. Have you tried using a generic oligo with deoxyinosine substituted for all six variable nucleotides of the blocking index oligo? I've been using this approach with a custom NimbleGen sequence capture experiment. The PCT_SELECTED_BASES is usually 55-65%. It's not as good as specific blocking oligos but good enough for my needs.

                      DoubleA

                      Comment


                      • #12
                        AA, I haven't tried that, that's a great suggestion. My next approach is to increase the Tm of the split blockers by using modified bases (2MO and LNA).

                        Comment


                        • #13
                          If your are interested in the specifics check the small print on page 4 of the attached document.
                          Attached Files

                          Comment


                          • #14
                            Thanks again. Asked my Nimblegen rep this same question and it would've been nice to see this year old document. Heh.

                            Comment


                            • #15
                              I'm not sure why but NimbleGen removed this document from their website.

                              Comment

                              Latest Articles

                              Collapse

                              • seqadmin
                                Strategies for Sequencing Challenging Samples
                                by seqadmin


                                Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
                                03-22-2024, 06:39 AM
                              • seqadmin
                                Techniques and Challenges in Conservation Genomics
                                by seqadmin



                                The field of conservation genomics centers on applying genomics technologies in support of conservation efforts and the preservation of biodiversity. This article features interviews with two researchers who showcase their innovative work and highlight the current state and future of conservation genomics.

                                Avian Conservation
                                Matthew DeSaix, a recent doctoral graduate from Kristen Ruegg’s lab at The University of Colorado, shared that most of his research...
                                03-08-2024, 10:41 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by seqadmin, Yesterday, 06:37 PM
                              0 responses
                              11 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, Yesterday, 06:07 PM
                              0 responses
                              10 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 03-22-2024, 10:03 AM
                              0 responses
                              51 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 03-21-2024, 07:32 AM
                              0 responses
                              68 views
                              0 likes
                              Last Post seqadmin  
                              Working...
                              X