    high 260/280 ratio (more than 3)

    I isolated RNA form human samples with concentration ~ 500 ng/ul , 260/280 Ratio = 2 and RIN = 7.3 then mRNA isolation with concentration ~ 5 ng/ul and 260/280 =2.2 but in mRNA fragmentation step, the concentration became 20 ng/ul and 260/280 ratio more than 3, why ?
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    That depends on how you quantify. Quantification in presence of salts and other reagents from buffers will less likely to be correct because most quantification assays are designed for nucleic acids resuspended in particular buffer. If you have used NanoDrop for 260/280 then you need to blank it with the same buffer that is present in your RNA reaction.

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #3
      Originally posted by Amr Elguoshy View Post
      I isolated RNA form human samples with concentration ~ 500 ng/ul , 260/280 Ratio = 2 and RIN = 7.3 then mRNA isolation with concentration ~ 5 ng/ul and 260/280 =2.2 but in mRNA fragmentation step, the concentration became 20 ng/ul and 260/280 ratio more than 3, why ?
      How about posting the actual UV absorbance spectrum for your samples? Based on the peaks we see there we might be able to give you a guess as to what substance is that is confounding your concentration/prep quality assay?

      Here is the spectrum of some of them:

      Techniques and protocol discussions on sample preparation, library generation, methods and ideas


      --
      Phillip

      Comment

      • Amr Elguoshy
        Junior Member
        • Feb 2015
        • 2

        #4
        Thank you So much. I thunk that my fragmented mRNA sample may be contaminated with ethanol because UV absorbance is high at 230 nm . but How I eliminate ethanol contamination from my sample?
        amr_150222_fragmented_mRNA.pdf

        Comment

        • pmiguel
          Senior Member
          • Aug 2008
          • 2328

          #5
          Originally posted by Amr Elguoshy View Post
          Thank you So much. I thunk that my fragmented mRNA sample may be contaminated with ethanol because UV absorbance is high at 230 nm . but How I eliminate ethanol contamination from my sample?
          [ATTACH]3617[/ATTACH]
          Ethanol doesn't absorb UV that I am aware of. Acetate does though and sodium acetate is frequently used as a co-precipitant in ethanol precipitation. Did you fragmentation buffer have a bunch of Mg-acetate in it?

          Looks more like a guanidium salt though and that is not typically used in fragmentation buffers.

          --
          Phillip

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
            by SEQadmin2



            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
            ...
            10-14-2026, 09:55 AM
          • SEQadmin2
            New Genomics Technologies Take Aim at Long-Standing Limits
            by SEQadmin2


            Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

            We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
            ...
            09-28-2026, 10:25 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 10-22-2026, 11:13 AM
          0 responses
          27 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-22-2026, 10:48 AM
          0 responses
          20 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-15-2026, 10:52 AM
          0 responses
          36 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-14-2026, 05:39 AM
          0 responses
          43 views
          0 reactions
          Last Post SEQadmin2  
          Working...