  • Dr. Hybe
    Member
    • Jun 2009
    • 14

    #1

    Question About Illumina Paired end Sequencing

    Quick question? Would it be possible to reposition the flowcell between read 1and 2 on the GA2x? i.e. take the flowcell/prism off and clean it in between reads? Also would it be possible to load the flowcell and do a manual fluidics inspection b/w R1 and R2? What about recalibrating b/w R1 and R2?

    Thanks in advance for your response,

    Dr. H
  • kmcarr
    Senior Member
    • May 2008
    • 1181

    #2
    Removing a flow cell between read 1 and 2, no you absolutely can not do this IF you want to maintain the pairing between the reads which I presume is a requirement. The location of clusters relative to the 0,0 point of the flow cell coordinate system is determined during the first 5 cycles of read 1 and is thus set for the remainder of the run. It's not within the realm of human ability to reposition the flow cell to within the tolerances required (at least speaking for me, I shouldn't assume for anyone else I suppose )

    Comment

    • GW_OK
      Senior Member
      • Sep 2009
      • 411

      #3
      I think you can do it on a Hiseq but on a GA, as kmcarr said, don't do it if you want to keep the pairings.

      Comment

      • kmcarr
        Senior Member
        • May 2008
        • 1181

        #4
        Originally posted by GW_OK View Post
        I think you can do it on a Hiseq ...
        Yesterday I would have believed there was no way you could do this on the HiSeq either, under the same theory as the GA. Our FSE just happened to be here working on our GA so I put the question to him. Yes, you can remove and replace the flow cell on a HiSeq at any time during the run. Naturally Illumina does not recommend this and should only be done in extreme circumstances. The HiSeq flow cells have registration marks on them which the software can use to provide a precise relative positioning. If a flow cell is removed mid run, when you remount the flow cell the software locates the registration marks and resets the 0,0 position according to those.

        Learn something new everyday. Thanks SeqAnswers and GW_OK.

        Comment

        • Dr. Hybe
          Member
          • Jun 2009
          • 14

          #5
          Thanks for the replies - Also I learned to day from FSE that it is possible on the GA to reload the calibration specifications in case you need to Load flowcell to check fluidics or shut down the software for some reason. Again not recommended, but possible if you have to. It will be interesting to see if they put markers on newer versions of the GA flowcells...

          Comment

          • Santosh
            Member
            • Feb 2011
            • 11

            #6
            Focus Issue

            Hey all, here we have a new problem. We are running a 101 bp Paired End run. The Read 1 was done and the Read 2 had also started. After reading through 117 bases, there was a power shut down and everything including the machine, stopped working. It happened while the 117th Base Scanning was half way through. The run was anyway resumed after 10 minutes and the scan mix was pumped. The scanning started from the 1st tile of the 117th base. But then, only the right half of the tile was showing clusters and the exact left half was completely dark. Could that be some kind of Focus Issue?

            Comment

            • Santosh
              Member
              • Feb 2011
              • 11

              #7
              Here I have another question. Just imagine, the Paired End run is going on and the machine is somewhere in Read 2 and ran out of focus. If u can stop the run and then go back to the calibration tab before the Read 1 FBR, and then resume the run from Read 2 after the Calibration, will it impact the coordinates on the flow cell? Just out of curiosity, will it have the the same calibration metrics as earlier? Because essentially here, we dont have to move the flow cell from its stage. Its just the Recipe we have to play with.

              Comment

              • GW_OK
                Senior Member
                • Sep 2009
                • 411

                #8
                How did you resume your run? Did you see if the camera did any alignments to the flowcell? Also, a total power outage is quite likely to cause vacuum loss on the flowcell, I'd check that out pronto.

                PS, the same thing happened to us but we were out for 30 minutes. Total loss of vacuum.

                Comment

                • HESmith
                  Senior Member
                  • Oct 2009
                  • 512

                  #9
                  We've been able to recover the run after an "intential" power outage (b/c of port comm. issues, we had to restart the computer and HiSeq). We resumed the run from the start screen, and went through the normal checks of vacuum (which had been lost) and flow. The data (paired-end 76bp) were fine.

                  Comment

                  • Santosh
                    Member
                    • Feb 2011
                    • 11

                    #10
                    Hi,
                    Thanx fr the reply.
                    But referring to what GW_OK said about the vacuum loss, is it applicable to GA2X? I thought vacuum operates only in Hiseq. And also, after resuming the run, we straight away resumed the run by pumping 75 uL of SMX. Even the camera did not do any kind of alignment on the flow cell. It straight away started imaging from the first tile of the cycle where it stopped.

                    Comment

                    • GW_OK
                      Senior Member
                      • Sep 2009
                      • 411

                      #11
                      Oh, yeah, that vacuum loss is only for the Hiseq. It might be some sort of focus issue with the GA2, or the stage might have moved minutely out of position, I don't know. I think there is a way to start the scan over completely for a step but I can't remember, that way it wouldn't start right at the tile it left off.
                      Anyway, how did the rest of the run look?

                      Comment

                      Latest Articles

                      Collapse

                      • SEQadmin2
                        16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                        by SEQadmin2




                        NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                        Today, 06:46 AM
                      • SEQadmin2
                        Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                        by SEQadmin2



                        Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                        Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                        ...
                        10-14-2026, 09:55 AM

                      ad_right_rmr

                      Collapse

                      News

                      Collapse

                      Topics Statistics Last Post
                      Started by SEQadmin2, 10-28-2026, 09:57 AM
                      0 responses
                      22 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-28-2026, 09:51 AM
                      0 responses
                      14 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-22-2026, 11:13 AM
                      0 responses
                      67 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-22-2026, 10:48 AM
                      0 responses
                      47 views
                      0 reactions
                      Last Post SEQadmin2  
                      Working...