  • HESmith
    replied
    random (adj.) - lacking a definite plan, purpose, or pattern (emphasis added).

    The poster stated that the "first four bases were completely random". I assumed (not "presumed") he meant what he wrote :-).

    Semantically, "high complexity" does not mean base composition diversity, which is the relevant issue for cluster calling. A library that consists solely of AAAAA, CCCCC, GGGGG, or TTTTT starts (in roughly equal amounts) would suffice, yet (almost) no one would argue that this constitutes high complexity.

    Apologies if this message comes across as cranky, Phillip. I was just trying my best to help the poster, and don't see how your comments contribute to the solution.

    -Harold

    Leave a comment:


  • pmiguel
    replied
    Originally posted by HESmith View Post
    If the first four bases are random, then subsequent low complexity should not adversely affect cluster calling or data quality. Excessive cluster density is a possible culprit: what are your raw and PF values?
    Semantically "random" does not mean "high complexity". See.

    Point being we presume "random" here means all four bases were a random mix of ACGT and therefore an even mixture of all 256 possible sequences. But we would need to know what the method of generating these was to know.

    --
    Phillip

    Leave a comment:


  • greigite
    replied
    For what it's worth we sequenced two lanes on a HiSeq (v2 flow cell) containing 11 7 bp inline barcodes with a spike-in of 5% phiX. Despite the low diversity visible in fastqc plots for the first 7 bp our cluster density and % of clusters passing filter was comparable to other lanes on the same run that did not have any low-complexity issues.

    Leave a comment:


  • csquared
    replied
    We do similar things to what you are describing all the time. A 20-30% PhiX spike (or any other library) should do the trick. PhiX is easy since it can be easily removed without an index and you can monitor the percent alignment as the run is going.

    Our most common condition is a HiC or 5C library where we need to get through some T3 and T7 sequences that are common to all of the samples. We have used both ChIPseq libraries as well as PhiX spikes with very good results. The use of the ChIPseq libraries just allows those reads to be used for something useful where PhiX is just data thrown away.

    Add in a spike and lower cluster density on the HiSeq to the 500k to 600k range and you should be fine. If you want to avoid the spike altogether, lowering clusters to about 200k also works but with more variable results.

    Leave a comment:


  • HESmith
    replied
    If the first four bases are random, then subsequent low complexity should not adversely affect cluster calling or data quality. Excessive cluster density is a possible culprit: what are your raw and PF values?

    Leave a comment:


  • HMorrison
    replied
    Originally posted by HESmith View Post
    There's an alternative approach, assuming that you have not yet constructed the libraries. Design them so the junction is at the opposite end of the insert, and perform paired-end sequencing. Cluster calling is based only on the first five cycles of read one, so you'll avoid the low-complexity issue.

    I have a sample of 96-plex low diversity amplicon libraries running now and clusters were found just fine--but the low diversity is causing a tremendous discrepancy between the blue and the green box-and-whiskers plot--raw clusters and clusters passing filter. I hope those data are recoverable at the end. Nothing in my primer design, barcoding, indexing scheme can change the fact that it's "low complexity". First four bases were completely random and followed by eight different in-line bar codes.

    This is PE sequencing.

    Yet I know labs are making this work.

    Leave a comment:


  • pmiguel
    replied
    I have seen bad batches of phiX that had fairly high (a few percent, I think) adapter dimers levels in them. Maybe you should make your own genomic DNA library to make sure your "diluent" is of high quality.

    Also you could obtain your "diluent" by sub-contracting a sequencing job. Send an email out to a prospective department (maybe one with a high level of plant or fungal sciences being done) and offer a one time only discount genome sequence. Our diluent was a sorghum genomic DNA library.

    --
    Phillip

    Leave a comment:


  • fkrueger
    replied
    Just as another thought, if you could afford to spike in 90-95% gDNA, couldn't you also find an external sequencing facility who still run GAIIx's and use the methods which work well on these?

    Leave a comment:


  • BIG_SNP
    replied
    We have had great success using the NuGen library prep. Their adapters have inline barcodes which adds to the diversity for the first cycles and allows sequences to pass filter. After passing filter the HiSeq can sequence the no or low diversity samples without any problems.

    Leave a comment:


  • Simcom
    replied
    Originally posted by pmiguel View Post
    We got it to work on our HiScanSQ -- which uses the same chemistry as the HiSeq, but only scans the top of the flowcell. Not an identical situation, but we had some SMART cDNAs that we sheared and ligated TruSeq adapters on. So about 1/2 of them had the same 50 nt of SMART primer at the beginning. We mixed them 1:1 with a genomic DNA library. Cluster registration went fine.

    --
    Phillip
    OK, that is good to hear. I'm not sure why my core was having trouble spiking 1:1 gDNA.

    Leave a comment:


  • Simcom
    replied
    Originally posted by HESmith View Post
    Yes, adapter reads are present.

    Just remember that you'll also need to include the standard sequencing primer for the gDNA library (or construct the gDNA library with custom adapters to match your custom primer).
    Yep, I planned on including both primers. Thank you so much for your help, I really appreciate it.

    Leave a comment:


  • HESmith
    replied
    Yes, adapter reads are present.

    Just remember that you'll also need to include the standard sequencing primer for the gDNA library (or construct the gDNA library with custom adapters to match your custom primer).

    Leave a comment:


  • pmiguel
    replied
    We got it to work on our HiScanSQ -- which uses the same chemistry as the HiSeq, but only scans the top of the flowcell. Not an identical situation, but we had some SMART cDNAs that we sheared and ligated TruSeq adapters on. So about 1/2 of them had the same 50 nt of SMART primer at the beginning. We mixed them 1:1 with a genomic DNA library. Cluster registration went fine.

    --
    Phillip

    Leave a comment:


  • Simcom
    replied
    Originally posted by HESmith View Post
    If, as you say, you can tolerate discarding 90-95% of the reads, then spiking in a gDNA library at that level will definitely solve your problem. After all, adapter dimers are often present at 5-10% in many libraries (the same % as your desired samples), and sequencing them is not a problem!
    Thanks, this gives me confidence. Just to be sure though: do the adapter-adapter ligation reads come back in the data, or does the machine throw them out and not include them in sequencing results? If you actually get the adapter-adapter reads from the machine, I should be golden.

    Leave a comment:


  • HESmith
    replied
    If, as you say, you can tolerate discarding 90-95% of the reads, then spiking in a gDNA library at that level will definitely solve your problem. After all, adapter dimers are often present at 5-10% in many libraries (the same % as your desired samples), and sequencing them is not a problem!

    Leave a comment:

Latest Articles

Collapse

  • SEQadmin2
    16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
    by SEQadmin2




    NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
    Yesterday, 06:46 AM
  • SEQadmin2
    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
    by SEQadmin2



    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
    ...
    10-14-2026, 09:55 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, 10-28-2026, 09:57 AM
0 responses
29 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-28-2026, 09:51 AM
0 responses
20 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-22-2026, 11:13 AM
0 responses
71 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-22-2026, 10:48 AM
0 responses
48 views
0 reactions
Last Post SEQadmin2  
Working...