  • LejzerowiczF
    replied
    for diversity analysis of communities, using long primer constructs
    such as [P5][Seq Primer]NNNNN[Target Primer]
    revives the long-known primer bias (see Berry et al. 2011 http://aem.asm.org/content/77/21/7846.short)
    its best to design short, anti-complementary tags (the shortest the easiest to satisfy anti-complemetarity).

    short primers + combinatorics + 1 lib prep => cheap solution

    moreover, adding tags in amplicons make their first, hypervariable bases tobe sequenced in the first miseq cycles... => high signal entropy => good sequencing quality.

    to be acounted: short tags will be in the sequences so you only loose few postions over the read length...

    for all recommendation + how to avoid tags swiching cross-contamionations resulting (mostly) from the lib- prep: see Esling et al. 2015 http://nar.oxfordjournals.org/conten...ar.gkv107.full

    good luck with your data!

    F

    Leave a comment:


  • pmiguel
    replied
    Modulo any special issues with v2 hardware, the "spike in 50% phiX" seems to be a cure-all. Using v1 hardware we did a v3 loop sequence where the first 20 bases were 100% identical. But the addition of a genomic DNA library allowed this to work fine. So just use that method initially and you should be fine.

    --
    Phillip

    Leave a comment:


  • vl80
    replied
    Apologies for reactivating an old thread. I posted a similar question yesterday and thought may be asking here may get more answers.
    I plan to sequence fragments ranging from 90 bp - 450 bp using Miseq as a pilot study. Per sample, I would have 2 amplicons. My PCR is a multiplex with > 10 primer pairs per reaction. I am completely new to this field and my theorising was to use the Trueseq kit, skip the fragmentation and end repair steps and proceed with the other steps. However, reading this thread may me doubt the appropriateness of that method. I am worried about suggetion to modify the PCR primers with an N sequence as this may cause mispriming in my multiplex PCR. Any suggestions are welcome. Thanks a lot

    Leave a comment:


  • bbeitzel
    replied
    groad,

    If you have incorporated the Illumina-specific sequences (sequence read primer binding sites, bridge PCR sequences, index read primer site) into your primers, then the 12 Ns would come between the Illumina stuff and the gene-specific sequence. If you are adding the Illumina adapters after PCR (ie. running amplicons through the TruSeq protocol), then you would put the 12 Ns at the 5' ends of both PCR primers. They have to be on both primers since TruSeq will not distinguish one end from the other.

    Hope this helps.

    Leave a comment:


  • ramsemm
    replied
    Hi!

    I am doing some similar sequencing - 30 samples of 1 amplified region. I was wondering if anyone has a citation or comment on what level is the limit for assigning taxonomy to 100 bp sequences from bacterial 16S (V4) amplicons?

    Thank you in advance,
    Molli

    Leave a comment:


  • groad
    replied
    Originally posted by bbeitzel View Post
    I don't think balanced barcodes will work. The index read occurs after read 1 so the instrument will still have issues calling clusters.

    I am doing some tag sequencing that requires a custom library prep protocol, and I have the same issue with homogeneous nucleotides for the first 10 bases of read 1. Illumina tech support recommended adding a stretch of 12 Ns (not 6) right after the read 1 primer binding site. It seems like that would be fairly straightforward to add to primers for amplicon sequencing.

    I am pretty sure (but don't quote me on this) that you can specify indeces of different lengths on the sample sheet. I think that the 2 index reads use a 6 base and a 7 base index.
    Hi bbeitzel,

    Do you mean to add 12 degenerated nucleotides to the 5' end of gene specific primer? Thanks.

    Jim

    Leave a comment:


  • gnomers
    replied
    Originally posted by bbeitzel View Post
    I don't think balanced barcodes will work. The index read occurs after read 1 so the instrument will still have issues calling clusters.

    I am doing some tag sequencing that requires a custom library prep protocol, and I have the same issue with homogeneous nucleotides for the first 10 bases of read 1. Illumina tech support recommended adding a stretch of 12 Ns (not 6) right after the read 1 primer binding site. It seems like that would be fairly straightforward to add to primers for amplicon sequencing.
    Thanks for your answer. I should have clarified that I am not using truseq/nextera indexing reads; my barcodes compose the very first bases of read1. So, it sounds like I could use balanced barcodes, in lieu of the stretch of 12 Ns they recommended to you.

    Leave a comment:


  • bbeitzel
    replied
    I don't think balanced barcodes will work. The index read occurs after read 1 so the instrument will still have issues calling clusters.

    I am doing some tag sequencing that requires a custom library prep protocol, and I have the same issue with homogeneous nucleotides for the first 10 bases of read 1. Illumina tech support recommended adding a stretch of 12 Ns (not 6) right after the read 1 primer binding site. It seems like that would be fairly straightforward to add to primers for amplicon sequencing.

    I am pretty sure (but don't quote me on this) that you can specify indeces of different lengths on the sample sheet. I think that the 2 index reads use a 6 base and a 7 base index.

    Leave a comment:


  • gnomers
    replied
    I am looking for an alternative to spiking in 50% or more PhiX while sequencing low diversity amplicon libraries.

    I am wondering if using longer barcodes with balanced representation of bases (up to 24 bases) would be sufficient.

    Additionally, does anyone know if I can specify barcodes of different lengths on a MiSeq sample sheet?

    Leave a comment:


  • NextGenSeq
    replied
    Yes, you just buy the PhiX control from Illumina and spike it in.

    Leave a comment:


  • ShiveringFire
    replied
    no more deferred cluster calling

    Thanks Philip,

    I came across this paper formally defining the problem and a possible solution:


    Sadly I was told that MiSeq doesn't save raw image files so a "deferred cluster calling" is not possible anymore. This sounds like getting rid of the evidence for a dirty job...

    So spiking a lot of pHiX seems like the only solution for low complexity. Can one spike phiX even in custom made libraries?

    Leave a comment:


  • pmiguel
    replied
    Originally posted by ShiveringFire View Post
    Can you educate me more on "cluster caller"? I recently got results from MiSeq Paired-End run (150bp) and I suspect there's a double read problem. Qscores are rather wavy at the beginning.

    My samples are multiplexed targeted re-sequencing of exomes from 16 genotypes. Could this be due to a low complexity issue discussed on this thread?

    I also heard rumors that if the first 4bp of a read is identical (which is very likely in targeted re-sequencing) it will be assigned to the same cluster. Is this true?
    I am thinking about this a lot. Especially with respect to possibly streamlining our operations by dumping our GS-FLX.

    Our MiSeq is apparently 6 weeks from being delivered, so I can only extrapolate from the performance of our HiScanSQ (which is similar to a HiSeq). There are two issues masquerading as a single "problem".

    (1) The first is that the actual instrument focusing software goes nuts and misfocuses if you have an empty tile (or set of tiles?) early in the run. Probably mainly during the first cycle. For our HiScanSQ this means if you don't have a fair number of clusters in the G&T channel and the A&C channel, you have a good chance that the instrument will pick the wrong focal plan to image.
    Doesn't anyone have connections inside Illumina they could point out this blunder to? If focus is good for one channel and bad for the other, why not just use the good focal point?
    Anyway, I think this is only an issue during the first few cycles (maybe only the first) of a read. Then the focal points seem to be carried along with little (or no) adjustments.
    The effect? Once you are out of focus, you are hosed for that tile for the run, I think. (Or "psuedo-tile" -- but I'll just presume that everyone reading this realizes that no masonry is involved in this process and leave it at "tile".) This applies no matter how perfect your cluster spacing is.

    (2) If the instrument did not hose up its focal plane during the first cycle, you can still have diminished results if you have a low complexity of base calls. This is the one everyone thinks about -- two adjacent clusters give the same base calls for 4 bases. If they are very close together, then neither of them may "register" (if I use the terminology correctly), well at all. This is a little more understandable. Except that Illumina foreclosed the use of methodology that could circumvent this issue on the GA-IIx for the HiSeq by making the early layers of data processing possible only on the instrument console.

    So if issue 1 doesn't kill you, issue 2 probably will. Work-arounds are required.

    --
    Phillip

    Leave a comment:


  • ShiveringFire
    replied
    Cluster caller?

    Originally posted by krobison View Post
    A related trick I saw published was to use a variable length sequence between the Illumina primers & targeting sequence; this way the targeting sequences aren't all in phase & the complexity is significantly increased in the eye of the cluster caller.
    Can you educate me more on "cluster caller"? I recently got results from MiSeq Paired-End run (150bp) and I suspect there's a double read problem. Qscores are rather wavy at the beginning.

    My samples are multiplexed targeted re-sequencing of exomes from 16 genotypes. Could this be due to a low complexity issue discussed on this thread?

    I also heard rumors that if the first 4bp of a read is identical (which is very likely in targeted re-sequencing) it will be assigned to the same cluster. Is this true?

    Leave a comment:


  • nancysch
    replied
    is it this one?

    Kindle et al., Detection and Quantification of Rare Variants with Massively Parallel Sequencing. PNAS doi:10.1073 (April 2011)

    Leave a comment:


  • CoastsideOldRider
    replied
    Do you have the reference for the paper mentioned in your post?

    Thanks!
    Marc

    Leave a comment:

Latest Articles

Collapse

  • SEQadmin2
    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
    by SEQadmin2



    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
    ...
    10-14-2026, 09:55 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, 10-28-2026, 09:57 AM
0 responses
22 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-28-2026, 09:51 AM
0 responses
14 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-22-2026, 11:13 AM
0 responses
67 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-22-2026, 10:48 AM
0 responses
47 views
0 reactions
Last Post SEQadmin2  
Working...