If you're worried that the NaOH concentration is too high, I would recommend neutralizing the NaOH using a bit of HCl. You can add it just after the first dilution with HT1.
Seqanswers Leaderboard Ad
Collapse
Announcement
Collapse
No announcement yet.
X
-
Originally posted by kcchan View PostIf you're worried that the NaOH concentration is too high, I would recommend neutralizing the NaOH using a bit of HCl. You can add it just after the first dilution with HT1.
Out of curiousity I just went and ran ILMN's standard protocol (as I posted above) and measured pHs...
10ul TrisHCl pH 8.5 (sample)
10ul 0.1N NaOH (diluted fresh)
980ul HT1 (initially pH 7)
Final pH was ~12.5, but this protocol works great. So pH isn't the only issue.
We've been using a neutralization protocol which brings the final pH back to 7, and haven't noticed any real increase in template requirements for v2 MiSeq kits. It's really funny/sad that people are finding they need higher template concentrations for v2 kits...when their final NaOH concentrations are 100% higher if they are following the "new/revD" manuals.
Comment
-
Originally posted by ECO View PostWe do this, but pH isn't the only factor here, I suspect library complexity and temp (effecting renaturation) are also important here.
Out of curiousity I just went and ran ILMN's standard protocol (as I posted above) and measured pHs...
10ul TrisHCl pH 8.5 (sample)
10ul 0.1N NaOH (diluted fresh)
980ul HT1 (initially pH 7)
Final pH was ~12.5, but this protocol works great. So pH isn't the only issue.
We've been using a neutralization protocol which brings the final pH back to 7, and haven't noticed any real increase in template requirements for v2 MiSeq kits. It's really funny/sad that people are finding they need higher template concentrations for v2 kits...when their final NaOH concentrations are 100% higher if they are following the "new/revD" manuals.
Tris has a 0.3 pH unit/10 degrees Celsius pKa delta, though. So if HT1 is tris-buffered, you might have some major systematic errors in your pH calculations if everything is not being measured at the same temp.
Now that we are using a v2 upgraded instrument with v2 chemistry, the issues seem to be gone. Although I did switch to using higher concentration initial library to avoid higher final NaOH concentration problems. (If any.)
--
Phillip
Comment
-
We've been having a hell of a time getting the desired cluster density on our system. The initial run with our FAS (PhiX only) loaded at 12pM and yielded 1068k/mm.
I ran a couple of genome runs subsequently. The first I loaded at 10pM, but I think the lab that prepared it fouled their library quant and it overclustered at around 1700k/mm. Still got 8.8Gb usable data from that run (67% PF).
The next run (genome) I loaded at 7pM at the advice of our FAS. The same lab prepared the library using the same method (Kapa qPCR), and it clustered at only 391k/mm.
I ran a few amplicon runs next (bacterial community amplicons) spiking PhiX between 25 and 40%, but all runs clustered around 400k despite loading between 6 and 8pM each time.
Then another genome run (for the same lab), loading at 10pM and still at 400k. I was checking this lab's quant by now, and it seemed pretty accurate this time. I have been bugging my FAS about this and she sent me a 50cycle kit I am running right now to check the system. Using PhiX only, loading at 10pM, I still only clustered at 600k/mm.
This is really frustrating and we need to be able to get our clustering right every time. I'm ready to start increasing the loading concentration, but those first two runs still have me a little gun shy.
Our instrument was delivered with the upgrade and we have only ever run v2 kits. Ideas?
AK
Comment
-
Illumina just posted a new technical bulletin regarding sample loading. They now use 4nM of DNA and 0.2N NaOH at 5uL each for the denaturation. This would result in a lower final NaOH concentrations than previous v2 protocols and more in line with the earliest MiSeq and cBot protocols.
Comment
-
Hi AKrohn -
At install (Oct 2012) our FAS said we should aim for cluster densities between 850 and 1200K/mm2 and load 12.5pM. This generally works for us when running balanced samples; low diversity samples are another issue.
Illumina's website indicates you should be able to get max instrument output with cluster densities between 880-965k/mm2 that pass filtering (based on PhiX data, of course).
Definitely follow the new denaturing/dilution protocol kcchan referenced to ensure the final NaOH conc is not more than 1mM in your final dilution. Higher than this can inhibit cluster formation...
Comment
-
Originally posted by genomeseeker View PostHi AKrohn -
At install (Oct 2012) our FAS said we should aim for cluster densities between 850 and 1200K/mm2 and load 12.5pM. This generally works for us when running balanced samples; low diversity samples are another issue.
Illumina's website indicates you should be able to get max instrument output with cluster densities between 880-965k/mm2 that pass filtering (based on PhiX data, of course).
Definitely follow the new denaturing/dilution protocol kcchan referenced to ensure the final NaOH conc is not more than 1mM in your final dilution. Higher than this can inhibit cluster formation...
According to my math, even with the "old" denature protocol, NaOH still should have been lower than 1mM in the final library. Also, our first PhiX (12pM) clustered at almost 1100k/mm. But yes, I have downloaded the new protocol and will follow the revisions therein.
Comment
Latest Articles
Collapse
-
by seqadmin
The sequencing world is rapidly changing due to declining costs, enhanced accuracies, and the advent of newer, cutting-edge instruments. Equally important to these developments are improvements in sequencing analysis, a process that converts vast amounts of raw data into a comprehensible and meaningful form. This complex task requires expertise and the right analysis tools. In this article, we highlight the progress and innovation in sequencing analysis by reviewing several of the...-
Channel: Articles
Yesterday, 07:48 AM -
-
by seqadmin
The field of epigenetics has traditionally concentrated more on DNA and how changes like methylation and phosphorylation of histones impact gene expression and regulation. However, our increased understanding of RNA modifications and their importance in cellular processes has led to a rise in epitranscriptomics research. “Epitranscriptomics brings together the concepts of epigenetics and gene expression,” explained Adrien Leger, PhD, Principal Research Scientist...-
Channel: Articles
04-22-2024, 07:01 AM -
ad_right_rmr
Collapse
News
Collapse
Topics | Statistics | Last Post | ||
---|---|---|---|---|
Started by seqadmin, Today, 06:57 AM
|
0 responses
4 views
0 likes
|
Last Post
by seqadmin
Today, 06:57 AM
|
||
Started by seqadmin, Yesterday, 07:17 AM
|
0 responses
13 views
0 likes
|
Last Post
by seqadmin
Yesterday, 07:17 AM
|
||
Started by seqadmin, 05-02-2024, 08:06 AM
|
0 responses
19 views
0 likes
|
Last Post
by seqadmin
05-02-2024, 08:06 AM
|
||
Started by seqadmin, 04-30-2024, 12:17 PM
|
0 responses
20 views
0 likes
|
Last Post
by seqadmin
04-30-2024, 12:17 PM
|
Comment