  • GenoMax
    replied
    Originally posted by pmiguel View Post
    Is this v2 kits on v2 hardware, or v1 hardware?

    --
    Phillip
    This is v.2. kits on v.2 hardware.

    Leave a comment:


  • pmiguel
    replied
    Originally posted by GenoMax View Post
    Philip,

    Disclaimer: I do not work in the lab. I am merely passing the following info along.

    We are not seeing this problem with v.2. kits but the inserts in question are not long.
    Is this v2 kits on v2 hardware, or v1 hardware?

    --
    Phillip

    Leave a comment:


  • pmiguel
    replied
    Originally posted by Bucky View Post
    Hey Philip,

    We are using v1 hardware. I should also mention that I pooled my sample at 4pm and then diluted it out more than usual in order to get a lower NaOH final concentration. Maybe that helped as well..
    Not sure I follow you. Normal protocol is to mix 10 ul of 0.2 M NaOH with 10 ul of 2 nM (total concentration of all libraries), incubate, then neutralize with 980 ul (50xdilution) of HT1. That gives you 1 ml of 20 pM denatured libraries, that you can dilute further, as needed.

    Could you elaborate, please?

    --
    Phillip

    Leave a comment:


  • GenoMax
    replied
    Philip,

    Disclaimer: I do not work in the lab. I am merely passing the following info along.

    We are not seeing this problem with v.2. kits but the inserts in question are not long.

    Leave a comment:


  • Bucky
    replied
    Hey Philip,

    We are using v1 hardware. I should also mention that I pooled my sample at 4pm and then diluted it out more than usual in order to get a lower NaOH final concentration. Maybe that helped as well..

    Leave a comment:


  • pmiguel
    replied
    Thanks Bucky,
    I have been stripping my gears trying to figure out what was going on. That makes it seem like v2 performs more like a HiSeq with respect to raw cluster density[1].

    Are you running v1 hardware, or has your instrument been upgraded to v2?

    Anyone else see this or not see this?




    [1]Of course as a matter of efficiency the cBot blows the MiSeq away -- 120 ul/lane for the HiSeq gets you the same density that 600 ul gets you for the MiSeq.

    --
    Phillip

    Leave a comment:


  • Bucky
    replied
    We had the same problem here. We used to cluster at 8pM and that resulted in about 700k/mm2. Switching to v2, we ended up with 200k/mm2. Loaded the latest run at 15pm and that seemed to do the trick (1000k/mm2). Will probably set up the next run at 13 or 14pM.

    Leave a comment:


  • pmiguel
    started a topic MiSeq v2 cluster density 1/2?

    MiSeq v2 cluster density 1/2?

    Did a run last week on a v1 hardware MiSeq using a v2 chemistry 500 cycle kit on qPCR titrated libraries. Clustered at 10 pM. Instead of the expected 800-1000 clusters/mm^2 we got 425.

    Again these were with KAPA qPCR titrated libraries using the same protocol that usually allows us to nail cluster density. These TruSeq RNAseq libraries were a little odd in that the insert sizes were a mean of 700-800 bp to get the full value of the 250 base reads. But we have gotten dead-on accurate clustering results for libraries with much larger mean inserts sizes than these in the past. (Obviously this does require compensation for the larger amplicon sizes compared to the KAPA standards.)

    This is the first time we have used v2 chemistry, however. But Illumina Tech support says there should be no major difference in cluster densities with v2. I have another run to do with another set of similar libraries, but I don't want to waste the reagents getting 1/2 the amount of data I should. On the other hand clustering at 20 pM seems too aggressive.

    Anyone seeing much lower than expected cluster densities running v2 MiSeq chemistry? Especially with fairly long amplicons?

    --
    Phillip

Latest Articles

Collapse

  • SEQadmin2
    16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
    by SEQadmin2





    NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification,...
    10-30-2026, 06:46 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, 11-17-2026, 06:27 AM
0 responses
35 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 11-12-2026, 07:59 AM
0 responses
45 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 11-05-2026, 09:25 AM
0 responses
57 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 11-03-2026, 11:54 AM
0 responses
80 views
0 reactions
Last Post SEQadmin2  
Working...