  • MiSeq Error stopped run

    I started a run on the MiSeq on Friday and by Monday morning I noticed that an error caused the MiSeq to stop just less than 2 hrs into the run. The error saids "No useable signal found in the images; it is possible that clustering has failed." Has anyone recieved this error before? I was running a few Nextera XT libraries. I have used this procedure before with success. This error is a first.

  • #2
    Have you checked images in the "Thumbnail_Images/L001/C1.1" directory to see if there are clusters?

    Comment


    • #3
      If you spike in PhiX you will be able tell if it's a chemistry or instrument error. Without spiking it is difficult to troubleshoot.

      Comment


      • #4
        No thumbnail images were created and I didn't use PhiX. What I can tell you is that the machine went under rountine maintenance and a replacement of the motor that controls the focus mirror. The machine worked very well after the service job, but we used a non-expired flowcell and reagents. For this run we used an expired flowcell (june 6th expiry) and expired reagent cartridge. Do you think that the expired flowcell may have caused this level of failure?

        Comment


        • #5
          It may be best to open a ticket with Illumina tech support so they can remote-in to look at this run.

          If there was a hardware failure it should show up in the "Logs" directory. You could look through those files (*.log) while you wait to hear back from Illumina.

          Expired kit (unless it was mishandled in terms of storage) should not cause a total failure.

          Comment


          • #6
            The reagents should still have been fine even though they were "expired".

            I had this problem once with a run, and the issue was ultimately chalked up to a "bad flowcell". That excuse gets thrown around a bit by Illumina when they don't have any other explanation, but they'll usually replace the kit so you can re-run the libraries.

            As was noted above, you should always have phiX spiked into your samples so you can troubleshoot more effectively. Right now you don't know if your libraries or the kit or the MiSeq were at fault, although my suspicion would be the libraries.

            Did you do any library QC before trying to run them or did you just follow through the XT protocol? We use XT extensively (no pun intended), and we've had issues when following the denaturing and normalization parts of the protocol. We've since switched to stopping after the post-PCR cleanup and doing Qubit and Bioanalyzer to QC the library, followed by standard pooling and denaturing like these were standard libraries. Since we've switched to doing things this way, we've been getting more even pooling and have more control over clustering.

            Comment


            • #7
              I contacted Illumina and they had me run a number of diagnostic tests to make sure there was nothing wrong with the MiSeq. All tests passed. Therefore I was told that these errors are caused by the presence of bubbles at the wrong place and time. For example, a bubble present near the inlet port, where the optical system attempts to find focus, of the FC can cause focusing problems and lead to a premature run stop. Apparently, these errors do occur from time to time.

              Thanks everyone for your help.

              Comment

              Latest Articles

              Collapse

              • seqadmin
                Understanding Genetic Influence on Infectious Disease
                by seqadmin




                During the COVID-19 pandemic, scientists observed that while some individuals experienced severe illness when infected with SARS-CoV-2, others were barely affected. These disparities left researchers and clinicians wondering what causes the wide variations in response to viral infections and what role genetics plays.

                Jean-Laurent Casanova, M.D., Ph.D., Professor at Rockefeller University, is a leading expert in this crossover between genetics and infectious...
                09-09-2024, 10:59 AM
              • seqadmin
                Addressing Off-Target Effects in CRISPR Technologies
                by seqadmin






                The first FDA-approved CRISPR-based therapy marked the transition of therapeutic gene editing from a dream to reality1. CRISPR technologies have streamlined gene editing, and CRISPR screens have become an important approach for identifying genes involved in disease processes2. This technique introduces targeted mutations across numerous genes, enabling large-scale identification of gene functions, interactions, and pathways3. Identifying the full range...
                08-27-2024, 04:44 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by seqadmin, Today, 06:25 AM
              0 responses
              13 views
              0 likes
              Last Post seqadmin  
              Started by seqadmin, Yesterday, 01:02 PM
              0 responses
              12 views
              0 likes
              Last Post seqadmin  
              Started by seqadmin, 09-18-2024, 06:39 AM
              0 responses
              14 views
              0 likes
              Last Post seqadmin  
              Started by seqadmin, 09-11-2024, 02:44 PM
              0 responses
              14 views
              0 likes
              Last Post seqadmin  
              Working...
              X