Hi All,
We've seen a lot of read 2 failures on our HiSeq and on externally supplied data. The quality of the second index is terrible and we lose most of the reads as they cannot be demultiplexed.
Having looked into this we are now confident the root-cause is acidification of NaOH during the run, and that this means the template removal step before index reads is compromised. When we check NaOH on long-read (PE100) runs the pH has dropped to less than 10.
Has anyone esle seen this? I'd lie to get an idea of how widely this problem is being seen as NaOH has been a fly in Illumina's ointment for ever! Such a simple reagent should not be the casuse of a £5000 run failing after a weeks seqeuncing on a £450,000 machine!!!
Thanks.
James.