Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts

  • mareen_engel
    replied
    Thansk austinso,

    I think denaturation at a higher temperature sounds like a very good idea to tackle denaturation problems. (we did the standard NaOH denaturation 5min at RT, then mix with cold HT1).
    Just to be safe: for the TruSeq RNA libraries stored as dsDNA, you suggest to mix with NaOH, 5min RT, ice, mix with cold HT1, 2min 95deg, ice, pool, load.
    Woudl you suggest doing the NaOH before storage in later library preps as they suggest to do with the SGP plate?

    Thanks, Mareen

    Leave a comment:


  • austinso
    replied
    Originally posted by mareen_engel View Post
    Thank you Phillip.
    While I agree that we may sometimes tend to overrate the problem of duplicates in RNA-Seq, I think it's a slightly different thing here, where I do see 3 Mio duplicates exactly next to each other on the flowcell with overall duplicate numbers which in my eyes are not high enough to render this statistically possible? Usually we see some 0-low hundreds of those optical duplicates even in our low-complexity runs?
    For me this sounds like a problem in library prep/denaturation/clustering, would you disagree?

    Best, Mareen
    We've seen similar things, and tended to ignore dedup'ing for our particular application, which drove everybody (well...just the keyboard warriors) apesh*t.

    Doing library denaturation with NaOH for 2 min at 96C as recommended for TruSeq libraries (http://support.illumina.com/content/...15027983-c.pdf, p.36) did the trick in removing these duplicates.

    A.
    Last edited by austinso; 02-05-2015, 05:12 PM. Reason: clarification

    Leave a comment:


  • mareen_engel
    replied
    Thank you Phillip.
    While I agree that we may sometimes tend to overrate the problem of duplicates in RNA-Seq, I think it's a slightly different thing here, where I do see 3 Mio duplicates exactly next to each other on the flowcell with overall duplicate numbers which in my eyes are not high enough to render this statistically possible? Usually we see some 0-low hundreds of those optical duplicates even in our low-complexity runs?
    For me this sounds like a problem in library prep/denaturation/clustering, would you disagree?

    Best, Mareen

    Leave a comment:


  • pmiguel
    replied
    I don't know -- I don't really trust programs that claim to be able to find optical or PCR "duplicates". How do these programs know they are not just biological duplicates? RNA will have a highly non-normal distribution with the possibility of just a few transcripts making up a large percentage of the library. If there is any fragmentation/ligation bias in library construction then you will see lots of the same sequence popping up irrespective of any instrument/PCR duplication.

    --
    Phillip

    Leave a comment:


  • mareen_engel
    started a topic High number of optical duplicates on MiSeq

    High number of optical duplicates on MiSeq

    Dear all,

    we've run a Illumina TruSeq mRNA nonstranded library (mouse brain lower end of input recommendation) on a MiSeq for quality control with v3 Reagent Kit, 2*75 PE. The run has very good QCs, good cluster density and some 22 Mio reads.

    After pairing the reads, Picard MarkDuplicates reports more then 3 Mio optical duplicates (i.e. duplicates less then 100pixel apart on the flow cell, similar results for 10pixel) next to roughly the same amount of "real" PCR duplicates (50% of QC20 reads after throwing the optical duplicates).
    If analyzing the SE data without pairing, the number of optical duplicates is reduced to 0, thus it seems not to be a cluster-read failure.

    For me, this indicates that there's 3 Mio clusters on the flowcell next to a cluster with their reverse complement strand which seems to be much above chance level even with a high PCR-duplicate RNA-Seq?

    Potential explanations put forward by the representative contacted are incomplete library denaturation prior to loading (thus partially dsDNA library molecules hybridize to the flowcell and build two reverse complement clusters) or low complexity of the library (however, 3 Mio sounds really above chance level for me, the comeplexity of the library doesn't seem to be that bad?)

    Has anybody ever seen this? Any ideas what is causing this high number of optical duplicates?
    Or is this number of duplicates simply expected on a standard RNA-Seq on low input (~200ng)?

    Any ideas or recommendations are very much appreciated!
    Many thanks!

    Regards,
    Mareen



    Edit1:
    Ps.: Not sure which heading this thread should go to, please move if you have a better idea. Thanks!
    Edit2:
    library PCR was 12 cycles
    Last edited by mareen_engel; 02-05-2015, 04:13 AM.

Latest Articles

Collapse

  • SEQadmin2
    16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
    by SEQadmin2




    NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
    10-30-2026, 06:46 AM
  • SEQadmin2
    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
    by SEQadmin2



    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
    ...
    10-14-2026, 09:55 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, 10-28-2026, 09:57 AM
0 responses
34 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-28-2026, 09:51 AM
0 responses
22 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-22-2026, 11:13 AM
0 responses
72 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-22-2026, 10:48 AM
0 responses
48 views
0 reactions
Last Post SEQadmin2  
Working...