@melop: It is still not clear if you have talked with the facility about your data/concerns and what the result of that conversation was? Do you know if they contacted Illumina tech support about this run?
Based on your post in #29 it seems that other 6 lanes on this flowcell were very close to acceptable spec (Illumina spec only specifies 75% bases > Q30 for 2 x 150bp, which seems to refer to entire sequence output from that flowcell).
It is certainly possible that Illumina refused to provide free replacements. As with any customer service issue the kind of response one gets (in this case free replacements) may entirely depend on the support person you reach.
Ultimately your result may be due the characteristic of the library. A re-run may produce a very similar result. Only option then would be to remake the library and try again.
Based on your post in #29 it seems that other 6 lanes on this flowcell were very close to acceptable spec (Illumina spec only specifies 75% bases > Q30 for 2 x 150bp, which seems to refer to entire sequence output from that flowcell).
It is certainly possible that Illumina refused to provide free replacements. As with any customer service issue the kind of response one gets (in this case free replacements) may entirely depend on the support person you reach.
Ultimately your result may be due the characteristic of the library. A re-run may produce a very similar result. Only option then would be to remake the library and try again.