Was this ever resolved?
I am encountering similar issues with focus errors.
I know it's not my library because it was quantified by qPCR and we've sequenced this before.
My flow cell stage still looks quite dirty even after cleaning it, but Illumina says it's fine.
Unconfigured Ad
Collapse
X
-
Is a problem of focus alignment.
You need to realign the Focus position.
Leave a comment:
-
Hi ikripp,Originally posted by ikripp View PostMy experience with "best focus" issues is that it does come down to an over-clustered run, although we have a very over-clustered run on the machine at the moment and it hasn't failed - it just has poor QC metrics. I have had runs with lower clustering (still over-clustered) that have failed yet this one has persisted.
I've never seen an error of no usable signal. What do the thumbnail images look like?
in the case of "no usable signal", we didn't even have thumbnail images, but we do have folder "Images", and there's a "Focus" -> "C1" folder, with .tif images file.
Here an example of name: FocusImage_0_exp_200_Focus T_IM_zpos_-0.15_method_CoarseFocus.tif
It looks like a huge black square. Literally no signal
Leave a comment:
-
My experience with "best focus" issues is that it does come down to an over-clustered run, although we have a very over-clustered run on the machine at the moment and it hasn't failed - it just has poor QC metrics. I have had runs with lower clustering (still over-clustered) that have failed yet this one has persisted.Originally posted by ArciMol View PostHi everyone!
Here I was searching about failed runs on MiSeq, and found this thread
In my case, we do runs on MiSeq each month, and recently we've had problems with 2 runs (MiSeq v3, 600c). In the first failed run, it was a problem of focus ("Best focus is too near edge of range(...)"); on the 2nd failed run, it has no clustering ("No usable signal found in the images"). I'll post lot number, in case it helps someone else:
MiSeq Reagent v3 600c box 1 of 2: LOT 20183241 and LOT 20270202
MiSeq Reagent v3 box 2 of 2: LOT 20178686 and LOT 20262956
We've also had the same answer from our local Illumina manager ... something like "it's your libraries fault" or "did you even load library on cartridge?". As we're no newbies, these answers doesn't help
Has someone else had failed runs with MiSeq, or reagents v3?
I've never seen an error of no usable signal. What do the thumbnail images look like?
Leave a comment:
-
Hi everyone!
Here I was searching about failed runs on MiSeq, and found this thread
In my case, we do runs on MiSeq each month, and recently we've had problems with 2 runs (MiSeq v3, 600c). In the first failed run, it was a problem of focus ("Best focus is too near edge of range(...)"); on the 2nd failed run, it has no clustering ("No usable signal found in the images"). I'll post lot number, in case it helps someone else:
MiSeq Reagent v3 600c box 1 of 2: LOT 20183241 and LOT 20270202
MiSeq Reagent v3 box 2 of 2: LOT 20178686 and LOT 20262956
We've also had the same answer from our local Illumina manager ... something like "it's your libraries fault" or "did you even load library on cartridge?". As we're no newbies, these answers doesn't help
Has someone else had failed runs with MiSeq, or reagents v3?
Leave a comment:
-
What is the loading concentration ( ng/ul) you got based on tape station?
Leave a comment:
-
We do approximately a run each week and vary rarely do we have what is considered an unsuccessful run. I base the loading concentration off of the pool concentrations as determined by a HSD1000 tape run on a tapestation 4200, again previous run pools. So its hard to say what loading concentration I have gone with but our standard PhiX is 10% which changes when we have a unique pool. I work with 16S amplicon pools for reference.Originally posted by chuanwu View Postikripp, have you had a successful run recently? What is your loading concentration with how much % of PhiX? Thanks
Leave a comment:
-
ikripp, have you had a successful run recently? What is your loading concentration with how much % of PhiX? Thanks
Leave a comment:
-
KAPA is the most accurate, provided you calculate it correctly. We usually run librarys on BioA or AA to get the correct size distribution and use it to do the calculation with KAPA results.
Leave a comment:
-
All the system checks passed.
The only issue for me why the quant of library measured by KAPA qPCR, Qubit, BioAnalyzer and LabChip GX are not consistent with each other. Which platform is reliable?
Thanks
Leave a comment:
-
I had the same issue - "Best focus too near edge of range" and after contacting tech support and being directed to run these tests:
Y Stage test
M3 mirror test
Z Stage test
Compensator test
Conduct volume test
Thermal Ramp test
Prime Reagent Lines
Optics test
I was told the run was significantly overclustered and to drop my loading concentration and reload.
Leave a comment:
-
I had the same issue twice in a row at Miseq platform. How did you guys resolve the problem. I called Illumina and they just questioned a lot about my library.
Thanks
Leave a comment:
-
We have had this error several times as well. The run just stops without images to show whether there is under- or over-clustering. I am very doubtful if this is just a clustering issue. I think it is a hardware problem.
Leave a comment:
-
There's no restarting a run with focus problems. It's unlikely to be a camera problem, more likely cluster failure, lens or z stage issue. Absolutely call tech support
Leave a comment:
Latest Articles
Collapse
-
by SEQadmin2
Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?
Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating...-
Channel: Articles
10-14-2026, 09:55 AM -
ad_right_rmr
Collapse
News
Collapse
| Topics | Statistics | Last Post | ||
|---|---|---|---|---|
|
Genome Sequencing Reveals Large, Repeat-Rich DNA in Trapdoor Spider Named for Stephen Colbert
by SEQadmin2
Started by SEQadmin2, Yesterday, 09:57 AM
|
0 responses
20 views
0 reactions
|
Last Post
by SEQadmin2
Yesterday, 09:57 AM
|
||
|
Started by SEQadmin2, Yesterday, 09:51 AM
|
0 responses
11 views
0 reactions
|
Last Post
by SEQadmin2
Yesterday, 09:51 AM
|
||
|
Started by SEQadmin2, 10-22-2026, 11:13 AM
|
0 responses
67 views
0 reactions
|
Last Post
by SEQadmin2
10-22-2026, 11:13 AM
|
||
|
Started by SEQadmin2, 10-22-2026, 10:48 AM
|
0 responses
47 views
0 reactions
|
Last Post
by SEQadmin2
10-22-2026, 10:48 AM
|
Leave a comment: