    RRBS Library 4 peaks

    Dear everyone

    I have just prepared my multiplexed reduced representation bisulfite sequencing library. It seems like I have 4 peaks on my bianalyzer results arround 180, 235, 300 and 360 base pairs.

    Have any of you experienced these peaks before?

    I do multiplexed RRBS with Msp1 digestion, TruSeq Nano adapter ligation and cleanup with dynabeads AMPure XS.

    Myself Im thinking that it could be that the Msp1 has a high cutting efficiency at these peaks. Do you have any suggestions?

    Do you think that these libraries is good to go for sequencing?



    Kind regards from Emil.
    Attached Files
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    Few points:
    1- HS Chip is overloaded. To see real profile Chip should be loaded within specification, otherwise profile will be distorded
    2- Is this from human DNA?
    3- What was input and how many PCR cycles were done?
    4- Peaks are most likely fom microsattelite fragments that contain MspI site
    5- Profile looks a bit different from usual but could be due to overloading or overcycling

    Comment

    • Andersen
      Member
      • Oct 2015
      • 15

      #3
      Hi Nucacidhunter.

      1 - I will try and run a new chip when we are running a full load.
      2 - It is from human stromal vascular fraction
      3 - Input was 55 ng DNA originally from 12 samples pooled in one library. I did 20 cycles.
      4 - Thanks, I will try and look into the microsattelite fragments. I am quite new to sequencing and library prep so had not heard about it before.
      5 - Good, sounds like I will need to run another bioanalyzer before going to sequencing.

      Thank you for your help!

      Comment

      Latest Articles

      Collapse

      • SEQadmin2
        Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
        by SEQadmin2



        Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

        Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
        ...
        10-14-2026, 09:55 AM
      • SEQadmin2
        New Genomics Technologies Take Aim at Long-Standing Limits
        by SEQadmin2


        Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

        We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
        ...
        09-28-2026, 10:25 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by SEQadmin2, Yesterday, 11:13 AM
      0 responses
      16 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, Yesterday, 10:48 AM
      0 responses
      16 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-15-2026, 10:52 AM
      0 responses
      35 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-14-2026, 05:39 AM
      0 responses
      42 views
      0 reactions
      Last Post SEQadmin2  
      Working...