  • fanli
    Senior Member
    • Jul 2014
    • 197

    #16
    I think low diversity would only cause jaggedness in the intensity profile, but maybe you should check w/ tech support.

    Our 16S runs have 1st cycle intensities ~150.
    Attached Files

    Comment

    • RickC7
      Member
      • Feb 2010
      • 31

      #17
      strange

      hi fanli,

      hmmm, the run summary you posted on 5-11-2015 shows read 1 intensity at 17 and read 4 intensity at 53. The other run summary from the OP also shows low intensity 1st cycle. This also fits with what I see in 16s libraries on miseq. So, 3 different machines, 3 different places, 3 similarly low 1st cycle intensities...
      Tech support is telling me this is the reason I am having issues with run completion. Last 3 runs are terminating randomly, run1- cycle 385, run2-cycle 60, run3 - cycle 587. Tech support has been helpful in replacing kits, but miseq is still out of commission. Have arrange for libraries to be sequenced on another miseq, Qc all checks out so I have little to no concern about the libraries. One comment that came out was " your 1st cycle intensities are very low..." I get a stopped run and funky z-stage errors, z-stage replaced but same issue persists.

      Comment

      • fanli
        Senior Member
        • Jul 2014
        • 197

        #18
        My bad - that last screenshot is Called Int. Yeah, you're right - the Cycle 1 intensities for my last run are 30 and 39 for read 1 and read 3, respectively.

        We haven't had any issues with runs terminating randomly for 16S libraries though. Although now that I think about it, we did have one bacterial WGS run that died on cycle 515 or so. Something about a .NET framework error and tech support said they hadn't seen it before.

        Comment

        • BioGenomics
          Member
          • Apr 2009
          • 25

          #19
          Hi all,

          what is the minimum Q-value you would suggest for a 16s read/merged amplicon Trimming/clipping ?

          thanks

          Comment

          • thermophile
            Senior Member
            • Apr 2015
            • 243

            #20
            I don't trim based on qscore, I use the qscores to merge reads. I use mothur which impliments pandaseq for it's read merging
            Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
              by SEQadmin2




              NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
              Today, 06:46 AM
            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 10-28-2026, 09:57 AM
            0 responses
            22 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-28-2026, 09:51 AM
            0 responses
            15 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 11:13 AM
            0 responses
            68 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 10:48 AM
            0 responses
            48 views
            0 reactions
            Last Post SEQadmin2  
            Working...