Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • wzombie
    Junior Member
    • Dec 2011
    • 3

    #1

    Strange Bioanalyzer results

    Hi,

    we have some plant RNA samples obtained from a phenol-chloroform extraction followed by LiCl precipitation; we would like to sequence these RNA samples with Illumina Hiseq platform, however RIN ratios are all between 5.5 and 6.5 (I’ve attached a typical bioanalyzer result, Nanodrop values for this sample are 2.14 for 260/280 ratio and 2.41 for 260/230 ratio)

    Does anyone know what kind of contamination we have in our samples and to what kind of problems this could lead if we tried to sequence this sample? purification with phenol-choloform and ethanol precipitation does not lead to better results, any idea if there's another possibility to purify the samples?

    Thanks a lot for comments - suggestions!
    Attached Files
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    Did you do a DNAse treatment of your RNA preps? DNA might be causing your elevated baseline.

    About the extraneous peaks, what tissue and types of plants are these?

    You will often see mitochondrial and chloroplast ribosomal RNA peaks in plant RNA. That might be what you are seeing. Also, some organisms have different cleavage variants of rRNA. This is common in insects and, I think, some bacteria.

    --
    Phillip

    Comment

    • kmcarr
      Senior Member
      • May 2008
      • 1181

      #3
      Did you use the plant specific rRNA assay for the 2100? Plant chloroplasts have their own set of rRNAs which may be highly abundant in certain tissues. See this AppNote from Agilent about their plant RNA integrity assay.

      Comment

      • wzombie
        Junior Member
        • Dec 2011
        • 3

        #4
        Phillip, yes we did a DNAse treatment, then we checked by PCR for DNA contamination with several microsatellites and with an 18S rDNA marker, no fragment was amplified..
        These RNA samples were extracted from peach leaves, so yes there might be a lot of mitochondrial and chloroplast rRNA in there, but could they explain the peak at 49 seconds or the low RIN ratios?

        kmcarr, yes we also used the plant specific rRNA assay for the 2100, which concluded to very similar electropherograms and very similar RIN values...

        Comment

        • pmiguel
          Senior Member
          • Aug 2008
          • 2328

          #5
          Past a certain point it is hard to tell. There may just be a highly abundant non-rRNA RNA at 49 seconds. If you know anyone else who has isolated RNA from this tissue, you would want to ask them.

          You sample was denatured prior to running, right?

          --
          Phillip

          Comment

          • wzombie
            Junior Member
            • Dec 2011
            • 3

            #6
            thanks for your very fast comments!!

            well, a few weeks ago i thought that denaturation was responsible for bad results so i also loaded samples without denaturation, but no difference was observed between denaturated and not-denaturated samples..

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM
            • SEQadmin2
              New Genomics Technologies Take Aim at Long-Standing Limits
              by SEQadmin2


              Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

              We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
              ...
              09-28-2026, 10:25 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 10-22-2026, 11:13 AM
            0 responses
            26 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 10:48 AM
            0 responses
            20 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-15-2026, 10:52 AM
            0 responses
            36 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-14-2026, 05:39 AM
            0 responses
            43 views
            0 reactions
            Last Post SEQadmin2  
            Working...