    PCR duplicates

    Hi all. Sorry to bother you. After reading several threads about PCR duplicates, I'm totally confused.

    The last step of library preparation in illumina RNA-Seq is PCR which means a fragment will have many copies in the library. Are these copies the so-called PCR duplicates? People suggest to remove PCR duplicates during RNA/ChIP-Seq analysis. I'm confused that why the PCR step is added to the library preparation if the info of these copies will be finally removed in the analysis. I think I must misunderstand what is PCR duplicate. Could anyone help me out?

    Thanks very much! I will appreciate it.

    Best
  • ECO
    --Site Admin--
    • Oct 2007
    • 1360

    #2
    PCR is typically done to increase yield to more tractable levels (ie visible on gel, bioanalyzer, nanodrop, etc)...and does indeed generate exact copies of fragments (thus paired end reads will both have the exact same start positions). This would manifest as bias in any quantitation/read-counting experiment as different fragments may have different amplification efficiencies.

    If one quantitates their libraries by qPCR, has a fairly complex library (in relation to the number of reads collected), and is confident in the size distribution, one can force the levels of PCR duplicates very very low (<1%).

    Sounds like you understand it perfectly. =]

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #3
      Blithely removing "PCR duplicates" from a data set is bogus.

      There is no way to determine whether a group of sequence reads come from a single fragment of DNA/RNA or result from over-amplification of a single initial target strand in a typical data set.

      I think this is especially the case for ChIP seq or RNAseq as these would tend to have the highest initial start-site bias of commonly constructed libraries.

      --
      Phillip

      Comment

      • aquleaf
        Member
        • Mar 2010
        • 38

        #4
        Thanks for your reply.

        According to Illumina manual, 15 cycle PCR is performed. So there are about 30,000 copies for each fragment. How could the PCR duplicate level be so low? I guess only a small part of PCR product is added to the hybridization cell, isn't it? I'm lack of this information.


        Originally posted by ECO View Post
        PCR is typically done to increase yield to more tractable levels (ie visible on gel, bioanalyzer, nanodrop, etc)...and does indeed generate exact copies of fragments (thus paired end reads will both have the exact same start positions). This would manifest as bias in any quantitation/read-counting experiment as different fragments may have different amplification efficiencies.

        If one quantitates their libraries by qPCR, has a fairly complex library (in relation to the number of reads collected), and is confident in the size distribution, one can force the levels of PCR duplicates very very low (<1%).

        Sounds like you understand it perfectly. =]

        Comment

        • aquleaf
          Member
          • Mar 2010
          • 38

          #5
          Thanks for your reply. I agree with you. Especially our data are single reading, it is difficult to mark PCR duplicates.


          Originally posted by pmiguel View Post
          Blithely removing "PCR duplicates" from a data set is bogus.

          There is no way to determine whether a group of sequence reads come from a single fragment of DNA/RNA or result from over-amplification of a single initial target strand in a typical data set.

          I think this is especially the case for ChIP seq or RNAseq as these would tend to have the highest initial start-site bias of commonly constructed libraries.

          --
          Phillip

          Comment

          • pmiguel
            Senior Member
            • Aug 2008
            • 2328

            #6
            Originally posted by aquleaf View Post
            Thanks for your reply.

            According to Illumina manual, 15 cycle PCR is performed. So there are about 30,000 copies for each fragment. How could the PCR duplicate level be so low? I guess only a small part of PCR product is added to the hybridization cell, isn't it? I'm lack of this information.
            Well if you cluster at 10 pM, that amounts to roughly 6 million amplicons/ul. So you can do the math. 600 ul is added the MiSeq whereas 120 ul/lane is added to the HiSeq by the cBot.

            I think 10 pM on the MiSeq would give you about 12-16 million clusters. Whereas you would be at around 100 million in a HiSeq lane. (These are pretty rough figures.)

            --
            Phillip

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM
            • SEQadmin2
              New Genomics Technologies Take Aim at Long-Standing Limits
              by SEQadmin2


              Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

              We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
              ...
              09-28-2026, 10:25 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, Yesterday, 11:13 AM
            0 responses
            19 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, Yesterday, 10:48 AM
            0 responses
            18 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-15-2026, 10:52 AM
            0 responses
            36 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-14-2026, 05:39 AM
            0 responses
            42 views
            0 reactions
            Last Post SEQadmin2  
            Working...