    problem with Illumina PE RNA seq alingment

    Hi All,
    We recently started to do PE RNA seq and we encountered a problem the cause of which have eluded us so far. In summary, following cycle 30 we see rapid increase in alignment error. The error rates per cycle go up to 20% and the % reads with 1 or more mismatches goes up to 35-50%.
    What we have found is:
    1) The error is alingner-nonspecific ( we tested eland, bowtie and even commercial software).


    2) The error is not due to contamination with amplified adapters or low quality scores for the bases in the second half of the amplicons.

    3) The problem is RNA-specific. Both PhiX and genomic DNA samples had normal error and alignment plots.

    4) The problem seems to depend on both the length of the insert and the length of the run. During the same run, shorter insert libraries have more misaligned reads than libraries consisting of longer fragments. A given insert size library will have more misaligned reads during a longer run.

    5) There is nothing visibly and strongly wrong with the IVC plots of the affected samples. The only striking observation is that both the % basecall as well as (even more strongly) the % all intensities from the affected samples form some type of ‘fork’ structures in the later cycles of each read.

    Has anybody heard about such a problem or encountered it? Any suggestions are welcomed.

    We know that the problem is related to the sample prep and taking longer reads ( >200bp insert size) will help alleviate it but we wonder what is the cause of it?

Latest Articles

Collapse

  • SEQadmin2
    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
    by SEQadmin2



    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
    ...
    10-14-2026, 09:55 AM
  • SEQadmin2
    New Genomics Technologies Take Aim at Long-Standing Limits
    by SEQadmin2


    Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

    We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
    ...
    09-28-2026, 10:25 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, Yesterday, 11:13 AM
0 responses
16 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, Yesterday, 10:48 AM
0 responses
18 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-15-2026, 10:52 AM
0 responses
35 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-14-2026, 05:39 AM
0 responses
42 views
0 reactions
Last Post SEQadmin2  
Working...