Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts

  • super0925
    replied
    strange result in outlier

    Originally posted by dpryan View Post
    That's the common way to do things in edgeR. Realistically, the results shouldn't change much if one compares filtering like this before testing or doing so after (and using the same threshold). I think DESeq2's method makes more sense, but that's me.



    DESeq2 will just give a bunch of NA pvalues for those it filters. The results will be there, just not a test statistic. This makes life a bit easier if you ever need to look at multiple experiments together, since then you don't have to deal with genes being in one results file but not the other (plus, you get an idea of what's getting filtered out).
    Hi D
    The strange is that after I use the count outlier detection by DESeq2 (if I don't filter before DE analysis)
    Why outliers have fold change but not outliers have NA fold change and P-value?
    and why the first two genes have P-value but not have adjust P-value?
    Command in R are bold
    W<-res$stat
    maxCooks<-apply(assays(dds)[["cooks"]],1,max)
    idx<-!is.na(W)


    head(res[which(idx=="TRUE"),],100)

    DataFrame with 100 rows and 6 columns
    baseMean log2FoldChange lfcSE stat pvalue
    <numeric> <numeric> <numeric> <numeric> <numeric>
    ENSBTAG00000000005 1.0597616 0.6512851 0.4803849 1.3557569 0.1751765
    ENSBTAG00000000008 0.8508167 0.2153765 0.4600089 0.4682007 0.6396411
    ENSBTAG00000000010 2.9979379 0.2802807 0.5104307 0.5491062 0.5829326
    ... ... ... ... ... ...
    ENSBTAG00000000191 1.4469456 0.03588896 0.4931081 0.07278113 0.9419803
    ENSBTAG00000000195 0.4348506 -0.15788350 0.3280363 -0.48129882 0.6303041
    ENSBTAG00000000197 3.0236299 -0.08213013 0.5141906 -0.15972703 0.8730961
    padj
    <numeric>
    NA
    NA

    0.9626562
    ... ...
    NA
    NA
    0.9892849


    head(res[which(idx=="FALSE"),],100)

    DataFrame with 100 rows and 6 columns
    baseMean log2FoldChange lfcSE stat pvalue
    <numeric> <numeric> <numeric> <numeric> <numeric>
    ENSBTAG00000000003 0 NA NA NA NA
    ENSBTAG00000000011 0 NA NA NA NA
    ENSBTAG00000000020 0 NA NA NA NA
    ... ... ... ... ... ...
    ENSBTAG00000000437 0 NA NA NA NA
    ENSBTAG00000000438 0 NA NA NA NA
    ENSBTAG00000000441 0 NA NA NA NA
    padj
    <numeric>
    NA
    NA
    NA
    ... ...
    NA
    NA
    NA
    Last edited by super0925; 05-27-2014, 05:21 AM.

    Leave a comment:


  • super0925
    replied
    Originally posted by dpryan View Post
    That's the common way to do things in edgeR. Realistically, the results shouldn't change much if one compares filtering like this before testing or doing so after (and using the same threshold). I think DESeq2's method makes more sense, but that's me.



    DESeq2 will just give a bunch of NA pvalues for those it filters. The results will be there, just not a test statistic. This makes life a bit easier if you ever need to look at multiple experiments together, since then you don't have to deal with genes being in one results file but not the other (plus, you get an idea of what's getting filtered out).

    Let me summarize what I got:

    1. DESeq2 don't requrie for filter before DE analysis. If I insist, I could do it as well.

    2. Outlier list comparison steps:
    step1. After DESeq2 processing, we could run the command in section 4.3 , select the genes flagged as 'TRUE' to build the outlier gene list (list 1).
    step2. We get another list (list 2), which is the genes predicted as DE in edgeR or Cuffdiff while not in DESeq2.
    step3: We check the how many genes in list 2 while also in the list 1. The proption is bigger, the method (edgeR or Cuffdiff) is more unreliable.

    3. So far we cannot use "outlier list" to judge DESeq2 is good or not because only DESeq2 package give our function detecting the outliers, but we could observe the reliability of edgeR or Cuffdiff.

    Am I right?
    Last edited by super0925; 05-27-2014, 05:00 AM.

    Leave a comment:


  • dpryan
    replied
    Originally posted by super0925 View Post
    Hi D, I found that after I filter and save the tags that are expressed in at least one of condition in edgeR, there are only 9000 genes out of 25000 genes in my dataset.
    The command:
    keep <- rowSums(cpm(countstable )>1) >= 3 # mine data is 2 condition, each condition has 3 samples.
    countstable <- countstable[keep,]
    (question 1) Is my command OK?
    That's the common way to do things in edgeR. Realistically, the results shouldn't change much if one compares filtering like this before testing or doing so after (and using the same threshold). I think DESeq2's method makes more sense, but that's me.

    however,the DESeq2's result still has 25000 genes.
    So (question 2) Do I need to filter the countstable before DE analysis in DESeq2??
    because I think when I am doing the pipeline comparison , I cannot separately dectect the outliers on 9000 genes vs 25000 genes.
    DESeq2 will just give a bunch of NA pvalues for those it filters. The results will be there, just not a test statistic. This makes life a bit easier if you ever need to look at multiple experiments together, since then you don't have to deal with genes being in one results file but not the other (plus, you get an idea of what's getting filtered out).

    Leave a comment:


  • dpryan
    replied
    Originally posted by super0925 View Post
    Hi D, I still have some questions:

    1.I haven't totally get your meaning. Do you mean firstlly I detect the outliers (by section4.3 in DESeq2 vignette). If the genes found to be DE with edgeR but not DESeq2 were flagged 'TRUE ' as outliers, the edgeR results would be less reliable. and Vise versa?
    For example, suppose there 10 DE genes detected by edgeR but not DESeq2, and 20 DE genes detected by DESeq2 but not edgeR. and for first 10 genes, 8 are flagged as outlier, and for second 20 genes, 10 are flagged as outlier. So we could say edgeR (8/10=80%) is less reliable than DESeq2(10/20=50%). Am I right?
    More or less, yes. edgeR doesn't do outlier detection (as far as I recall, at least), so you just see if any of its DE genes were flagged by you'd just want to know that 8/10 of edgeR's DE genes are likely unreliable.

    2. Besides this outlier finding method, are there any other mothods (the more the better) could do the comparison in computer (not qPCR) as well? Currently I did the (1) overlap among the piepelines, and (2)outlier decetion as you know. I found a paper talking about the spearman correlation among the DE methods, is that also good? or the best ways are overlap and outlier.
    I can't think of anything other than the ways you mentioned off-hand. Using the spearman correlation is an interesting idea. Presumably one would rank the union of DE genes by their spearman correlation coefficient, perhaps applying some threshold.

    3. Before I do the DE analysis, I didn't use your genefilter package but use edgeR mannual recommend command which filter and save the tags that are expressed in at least one of condition in edgeR (but not DESeq2, cause you said the lastest version of DESeq2 could filter the counts automatically), is that OK?
    That's fine, you just have less power with those results (BTW, genefilter isn't my package, I just recommend its usage).

    4. Are these outlier genes useful/meaningful for DE analysis? I mean are these interesting results (large counts gene is good?) or bad effect on the DE analysis we must remove them?
    Sometimes the outliers are still useful, but usually not. It's often good to just have a look at the underlying data to see what's going on.

    Leave a comment:


  • super0925
    replied
    Originally posted by dpryan View Post
    There are two kinds of filtering that can occur. Previously, I was referring to "independent filtering", wherein filtering is performed to maximize power (see the genefilter package and accompanying paper). The second type of filtering, that I was alluding to just above, uses cook's distance (basically, if, when looking at a given gene, a single sample has too much leverage (in the statistical sense), then the fit is unreliable and we ignore statistical tests) to find genes where there may be outlier samples. DESeq2 won't normally (you can disable this, which turns out to be necessary on occasion) test these, so that can also produce a difference in the results between the various packages. If a number of the genes found to be DE with edgeR but not DESeq2 were flagged and excluded in this manner, then that'd be good to know, since then the edgeR results would be less reliable (likewise with cuffdiff).

    See section 4.3 in the DESeq2 vignette.
    Hi D, I found that after I filter and save the tags that are expressed in at least one of condition in edgeR, there are only 9000 genes out of 25000 genes in my dataset.
    The command:
    keep <- rowSums(cpm(countstable )>1) >= 3 # mine data is 2 condition, each condition has 3 samples.
    countstable <- countstable[keep,]
    (question 1) Is my command OK?

    however,the DESeq2's result still has 25000 genes.
    So (question 2) Do I need to filter the countstable before DE analysis in DESeq2??
    because I think when I am doing the pipeline comparison , I cannot separately dectect the outliers on 9000 genes vs 25000 genes.
    Last edited by super0925; 05-27-2014, 02:21 AM.

    Leave a comment:


  • super0925
    replied
    Originally posted by dpryan View Post
    There are two kinds of filtering that can occur. Previously, I was referring to "independent filtering", wherein filtering is performed to maximize power (see the genefilter package and accompanying paper). The second type of filtering, that I was alluding to just above, uses cook's distance (basically, if, when looking at a given gene, a single sample has too much leverage (in the statistical sense), then the fit is unreliable and we ignore statistical tests) to find genes where there may be outlier samples. DESeq2 won't normally (you can disable this, which turns out to be necessary on occasion) test these, so that can also produce a difference in the results between the various packages. If a number of the genes found to be DE with edgeR but not DESeq2 were flagged and excluded in this manner, then that'd be good to know, since then the edgeR results would be less reliable (likewise with cuffdiff).

    See section 4.3 in the DESeq2 vignette.
    Hi D, I still have some questions:

    1.I haven't totally get your meaning. Do you mean firstlly I detect the outliers (by section4.3 in DESeq2 vignette). If the genes found to be DE with edgeR but not DESeq2 were flagged 'TRUE ' as outliers, the edgeR results would be less reliable. and Vise versa?
    For example, suppose there 10 DE genes detected by edgeR but not DESeq2, and 20 DE genes detected by DESeq2 but not edgeR. and for first 10 genes, 8 are flagged as outlier, and for second 20 genes, 10 are flagged as outlier. So we could say edgeR (8/10=80%) is less reliable than DESeq2(10/20=50%). Am I right?

    2. Besides this outlier finding method, are there any other mothods (the more the better) could do the comparison in computer (not qPCR) as well? Currently I did the (1) overlap among the piepelines, and (2)outlier decetion as you know. I found a paper talking about the spearman correlation among the DE methods, is that also good? or the best ways are overlap and outlier.

    3. Before I do the DE analysis, I didn't use your genefilter package but use edgeR mannual recommend command which filter and save the tags that are expressed in at least one of condition in edgeR (but not DESeq2, cause you said the lastest version of DESeq2 could filter the counts automatically), is that OK?

    4. Are these outlier genes useful/meaningful for DE analysis? I mean are these interesting results (large counts gene is good?) or bad effect on the DE analysis we must remove them?
    Last edited by super0925; 05-27-2014, 01:47 AM.

    Leave a comment:


  • dpryan
    replied
    There are two kinds of filtering that can occur. Previously, I was referring to "independent filtering", wherein filtering is performed to maximize power (see the genefilter package and accompanying paper). The second type of filtering, that I was alluding to just above, uses cook's distance (basically, if, when looking at a given gene, a single sample has too much leverage (in the statistical sense), then the fit is unreliable and we ignore statistical tests) to find genes where there may be outlier samples. DESeq2 won't normally (you can disable this, which turns out to be necessary on occasion) test these, so that can also produce a difference in the results between the various packages. If a number of the genes found to be DE with edgeR but not DESeq2 were flagged and excluded in this manner, then that'd be good to know, since then the edgeR results would be less reliable (likewise with cuffdiff).

    See section 4.3 in the DESeq2 vignette.

    Leave a comment:


  • super0925
    replied
    Hi D
    (1) I will try the latest Cufflinks to repeat the work. because the Cuffdiff result is too liberal!
    (2) I haven't totally got what you said.
    if DESeq2 "flagged" a number of the edgeR/cuffdiff only genes as having outlier samples

    what does it mean and how to do it?

    Before In did the DE analysis, I filter and save the tags that are expressed in at least one of condition in edgeR (but not DESeq2, cause you said the lastest version of DESeq2 could filter the counts automatically). Is that filter the outliers?

    Sorry for the naive question.

    Leave a comment:


  • dpryan
    replied
    There's no way to judge accuracy from a Venn diagram. Which version of cufflinks did you use? Lately it tends to be more conservative than the others, so that seems off. What often happens is that the differences (e.g., DESeq2 vs. edgeR) are toward the margins of significance, where you get an adjusted p-value of 0.08 in DESeq2 and 0.11 in edgeR (or vice versa), which isn't surprising. One thing to check is if DESeq2 flagged a number of the edgeR/cuffdiff only genes as having outlier samples. This is a really nice feature and can help avoid false-positive findings.

    Leave a comment:


  • super0925
    replied
    Originally posted by dpryan View Post
    Base what you do next on the biological goals. Sometimes it makes sense to just offer an annotated list. If the list of DE genes is rather long, it'd be nice to mine out some links in pubmed for each of the genes, so that the researcher him/herself doesn't have to do it. If this experiment is simply fishing for changes, then GO and pathway analysis might be useful. If there's an expected effect, then doing those in a more targeted way (cf. camera() and roast() in limma/edgeR) would make more sense.
    So far I use Q-value<0.05 as threshold.
    If the overlap is like this (just based from Venn plot)
    which method or methods do you recommend? If we use 145 sharing ones I don't know is it too conservative?
    Attached Files

    Leave a comment:


  • dpryan
    replied
    Base what you do next on the biological goals. Sometimes it makes sense to just offer an annotated list. If the list of DE genes is rather long, it'd be nice to mine out some links in pubmed for each of the genes, so that the researcher him/herself doesn't have to do it. If this experiment is simply fishing for changes, then GO and pathway analysis might be useful. If there's an expected effect, then doing those in a more targeted way (cf. camera() and roast() in limma/edgeR) would make more sense.

    Leave a comment:


  • super0925
    replied
    Originally posted by dpryan View Post
    Read help(results)
    Hi D
    A quick question.
    So far I have processed a DE analysis with 3 replicates in 3 conditions as I told you. The collaborator is interested with the DE genes but not the transcripts. He will verify the DE genes may be by qPCR.
    What so far I get are the DE gene lists from Texudo/ DESeq2/ edgeR/ baySeq separately. And I got the Venn plot of the overlap among the DE genes lists.
    Is anything else I can do? e.g. decide the best pipeline, or some other DE analysis except only offering list?
    Or go to downstream analysis i.e. GO,pathway, etc.
    Pls give me some suggestion.
    Thank you!
    Last edited by super0925; 05-23-2014, 03:22 AM.

    Leave a comment:


  • super0925
    replied
    Originally posted by dpryan View Post
    Read help(results)
    Another question , is it critical if I don't filter the low counts by genefilter package? If I save the tags that are expressed in at least one of condition (like totourial in edgeR), is it enough?
    Last edited by super0925; 05-19-2014, 07:03 AM.

    Leave a comment:


  • dpryan
    replied
    Read help(results)

    Leave a comment:


  • super0925
    replied
    Originally posted by dpryan View Post
    DESeq (better would be DESeq2) would work fine as well.
    Hi D,
    I have tried what you said the multiple condition comparison in DESeq2.

    The input count matrix is (3 conditions with 3 samples in per condition)

    C1.R1 C1.R2 C1.R3 C2.R1 C2.R2 C2.R3 C3.R1 C3.R2 C3.R3
    ENSBTAG00000000003 0 0 0 0 0 0 0 0 0
    ENSBTAG00000000005 1 0 0 1 0 0 2 2 2
    ENSBTAG00000000008 2 2 1 0 2 0 1 1 2

    Here is my script:

    library("DESeq2")
    colData<- data.frame(condition=factor(c(rep("C1",3),rep("C2",3),rep("C3",3))))
    dds<-DESeqDataSetFromMatrix(countData= output2,colData= colData,design=~condition)
    dds$condition<-factor(dds$condition,levels=c("C1","C2","C3"))
    dds<-DESeq(dds)
    res<-results(dds)
    resOrdered<-res[order(res$padj),]
    head(resOrdered)
    write.csv(as.data.frame(resOrdered),file="C1_C2_C3_results.csv")

    But the output is only one table with the significant DE genes list ranked by P-value (is it C1vsC2vsC3?), but how do I know the ranking of C1vs C2,C2vsC3 and C1vsC3 which is what I want ?
    Sorry for this naive question
    Last edited by super0925; 05-19-2014, 03:02 AM.

    Leave a comment:

Latest Articles

Collapse

  • SEQadmin2
    16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
    by SEQadmin2




    NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
    10-30-2026, 06:46 AM
  • SEQadmin2
    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
    by SEQadmin2



    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
    ...
    10-14-2026, 09:55 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, 10-28-2026, 09:57 AM
0 responses
37 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-28-2026, 09:51 AM
0 responses
23 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-22-2026, 11:13 AM
0 responses
72 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-22-2026, 10:48 AM
0 responses
48 views
0 reactions
Last Post SEQadmin2  
Working...