  • Ahmadiut
    Junior Member
    • May 2015
    • 3

    #1

    RNA quality and quantity for RNA-seq

    Hi
    I have five samples and want to send them to an overseas institute for RNA-seq. My samples are from a native plant which RNA extraction of it is very difficult. Here I attached two files that show the agaros gel and nanodrop results. I am not sure about their qual and quant, Could you please tell me your opinions about them?
    Thanks in advance
    Attached Files
    Last edited by Ahmadiut; 05-17-2015, 07:13 AM.
  • dpryan
    Devon Ryan
    • Jul 2011
    • 3478

    #2
    The brightness of the wells suggests that you have significant genomic DNA contamination. You might want to throw some DNase on the samples (you'll need to repurify the results, likely just with a column).

    Comment

    • Ahmadiut
      Junior Member
      • May 2015
      • 3

      #3
      In the Past, I had some samples with DNA contamination, but they had a band between the well and 28s RNA. DNA move in the agarose gel, don't they?
      My samples have a lot of gum-resins and carbohydrates. Don't you think that the contaminants in the wells are gum-resin or carbohydrates with absorption in 230 nm? Are these kind of contaminates interfere in the process?
      Last edited by Ahmadiut; 05-17-2015, 01:19 PM.

      Comment

      • freestile
        Member
        • Aug 2014
        • 11

        #4
        I think that you have genomic DNA (the upper mark of each well). I recommend anyway use DNase previously start libraries preparation. About quality I can't see the the value of 260/280 ratio (must be near 2).
        Regards.

        Comment

        • nucacidhunter
          Jafar Jabbari
          • Jan 2013
          • 1250

          #5
          You RNA quality are good judged by nice rRNA bands (I assume these are from non-leave tissues) but NanoDrop is not accurate for quantification. You may try to use RNA specific dye such as Quant-iT RiboGreen RNA reagent for accurate quantification. Sticky materials in the wells most likely will inhibit or reduce enzymatic reactions efficiency during RNAseq library prep processing by preventing access of enzymes and reagents to RNA. While they may not cause library failure but they will impact the quality of data. Tolerance of such effect will be depending on your experiment aims. A column based post extraction RNA clean up kit can remove those from samples.

          Comment

          • pmiguel
            Senior Member
            • Aug 2008
            • 2328

            #6
            Originally posted by freestile View Post
            About quality I can't see the the value of 260/280 ratio (must be near 2).
            Regards.
            Well you can read the value at 260nm and 280nm as "4.5" and "2.0", respectively. But both of these numbers are meaningless as they reflect a small residual amount of phenol in the sample, not the nucleic acids.

            See:

            Techniques and protocol discussions on sample preparation, library generation, methods and ideas


            for common optically absorbing contaminants of DNA and RNA preps.

            The stuff stuck up in the well may or may not be DNA. It could be RNA complexed with glop that is hindering its migration into the gel. Or DNA, or a mixture of both.

            Genomic DNA if you don't beat it up too bad during prepping will tend to be in the 50-150kb range so it might not migrate into your gel much by the time rRNA had migrated as far as it had.

            --
            Phillip

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM
            • SEQadmin2
              New Genomics Technologies Take Aim at Long-Standing Limits
              by SEQadmin2


              Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

              We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
              ...
              09-28-2026, 10:25 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 10-15-2026, 10:52 AM
            0 responses
            14 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-14-2026, 05:39 AM
            0 responses
            21 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-13-2026, 09:56 AM
            0 responses
            19 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-09-2026, 11:14 AM
            0 responses
            27 views
            0 reactions
            Last Post SEQadmin2  
            Working...