  • a strange peak in LncRNA library

    Hi, all,
    We met a strange peak in prepared LncRNA library. The library construction procedure is listed below.
    1 micro gram total RNA after Ribozero treatment was used at starting material and ribosomal RNA was almost completely removed as PicoRNA chip map indicated and Ribozero- RNA is ~20-50ng. Library construction was performed following NEBNext® Ultra™ RNA Library Prep Kit for Illumina protocol with size selection for 300-400bp fragments. The PCR cycle number is ~15 cycle.
    In 12 of 16 samples, we found a strange sharp peak, which unlike those we usually obtained.
    Did you ever meet similar results? anyone could give us an explanation?
    Thanks!
    attachment 1: Strange peak in library 2100 map
    attachment 2: RNA after Ribozero treatment in Pico RNA map.
    Attached Files

  • #2
    It seems that you have used a gel based size selection method. If this is the case, peak is normal.

    Comment


    • #3
      Thanks for your reply, we used beads for size selection.
      Does the library could be sequenced and produce normal data?

      Comment


      • #4
        The kit that you have mentioned does not seem to include a size selection step and size selection of RNA-Seq libraries is uncommon. The library can be sequenced but data analysis will be difficult because somehow you have to account for the biases introduced by size selection that does not seem have worked well.

        Comment


        • #5
          Did you reduce the RNA fragmentation time to have longer fragments at the time of cDNA synthesis? It does look like the trace of a standard protocol RNA-Seq library that has had the lower MW fragments hacked off by a low ratio Ampure selection.
          Here's a thread about the Clontech Stranded kit that has a slide showing the effects of different RNA fragmentation times on the final library profile http://seqanswers.com/forums/showthread.php?t=44453. I believe the two kits have different divalent cation concentrations for the fragmentation, so you probably can't just lift the conditions from the slide, but the same principle would apply if you can afford the time/kits to do some empirical testing (If having this larger size library is actually important, which is probably only true if you're actually trying to assemble rather than count).

          Comment

          Latest Articles

          Collapse

          • seqadmin
            A Brief Overview and Common Challenges in Single-cell Sequencing Analysis
            by seqadmin


            ​​​​​​The introduction of single-cell sequencing has advanced the ability to study cell-to-cell heterogeneity. Its use has improved our understanding of somatic mutations1, cell lineages2, cellular diversity and regulation3, and development in multicellular organisms4. Single-cell sequencing encompasses hundreds of techniques with different approaches to studying the genomes, transcriptomes, epigenomes, and other omics of individual cells. The analysis of single-cell sequencing data i...

            01-24-2023, 01:19 PM
          • seqadmin
            Introduction to Single-Cell Sequencing
            by seqadmin
            Single-cell sequencing is a technique used to investigate the genome, transcriptome, epigenome, and other omics of individual cells using high-throughput sequencing. This technology has provided many scientific breakthroughs and continues to be applied across many fields, including microbiology, oncology, immunology, neurobiology, precision medicine, and stem cell research.

            The advancement of single-cell sequencing began in 2009 when Tang et al. investigated the single-cell transcriptomes
            ...
            01-09-2023, 03:10 PM

          ad_right_rmr

          Collapse
          Working...
          X