    low efficiency of DSN normalization

    We have set up a protocol for RNA-seq based on adapter-ligation to RNA-fragments. For ribosome removal, I would like to use the DSN normalization method, which is said to work better than bead-based methods (according to Illumina).

    The problem is that the efficiency seems to be very low. Our genomics department is performing the normalization for us prior to sequencing the libraries on an Illumina GA-II. I mapped the results to the reference genome and found that still on average 90 - 95 % of the reads map to ribosomal loci. Untreated samples typically have 98-99%. Thus, it's working but not satisfyingly.

    The sequence is a large bacterial genome (6.2 Mb).

    Does anyone have experience with DSN and could give us a hint on how to improve the efficiency?
    What are the ribosome ratios that you can typically achieve with the protocol (Illumina states ~15%).
  • mnkyboy
    Member
    • Mar 2009
    • 87

    #2
    Do you have an accurate measurement of the ng mass of the library you are DSN treating? I have been doing a lot of DSN (granted with mammalian) and I find that you really need to be in that 80-100 ng range measured by picogreen.

    I still see probably 1/10 that fail but that is most likely due to me losing the CoT during treatment.

    Comment

    • nasobema
      Member
      • Jul 2010
      • 14

      #3
      Originally posted by mnkyboy View Post
      Do you have an accurate measurement of the ng mass of the library you are DSN treating? I have been doing a lot of DSN (granted with mammalian) and I find that you really need to be in that 80-100 ng range measured by picogreen.

      I still see probably 1/10 that fail but that is most likely due to me losing the CoT during treatment.
      Yes, DNA concentration was measured with Qubit, which should be reliable.

      Any more ideas?

      We have been tightly following the Illumina protocol for DSN. Losing a constant temperate may always be an explanation but the technicians performing the protocol are well aware of this issue.

      Comment

      • irit
        Junior Member
        • Oct 2010
        • 5

        #4
        A temperature drop should have the reverse effect - digestion of secondary structures.

        Have you tested the enzyme to see if it is still active? Is it possible your incubation temperature is too high?

        Comment

        • mnkyboy
          Member
          • Mar 2009
          • 87

          #5
          Run the provided control with the protocol they have and that will tell you if your enzyme is still working. It is a short protocol and you can check results on gel or BioAnalyzer.

          Comment

          • nasobema
            Member
            • Jul 2010
            • 14

            #6
            Originally posted by irit View Post
            A temperature drop should have the reverse effect - digestion of secondary structures.

            Have you tested the enzyme to see if it is still active? Is it possible your incubation temperature is too high?
            Yes, the enzyme works fine on control dsDNA.

            Incubation temperature was 68° according to the Illumina protocol.
            So, could it help to reduce Temperature by a few degrees?
            We have currently no idea on the effect of altering the parameters like incubation Temp./Time and Hybridization Temp./Time.
            We are doing some experiments on that but I would still highly appreciate input from experienced DSN users.
            Thank you all for the previous replies.

            Comment

            • QTLdum
              Junior Member
              • Feb 2011
              • 5

              #7
              We encountered very similar problems (low efficiency of normalization) when using bovine samples. Thus, we decreased the hybridization and incubation temperature from 68° to 66° and 64° C, but this did not improve normalization. We had also confirmed that the DSN worked by performing the test protocol

              Comment

              • monad
                Member
                • May 2008
                • 40

                #8
                We worked with more than 8,000 library prep until today, so we are not inexperienced users here for Illumina library prep.

                However, DSN is not the way to go if you want to remove ribosomal RNA!

                DSN works beautifully for small amount of input total RNA, even below picogreen detection level, but not an effective method to remove ribosomal RNA. Typically you will expect to have >60% of your reads are aligned to rRNA sequences.

                Go with ribo minus or ribo zero type of removal methods.

                Comment

                Latest Articles

                Collapse

                • SEQadmin2
                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                  by SEQadmin2



                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                  ...
                  10-14-2026, 09:55 AM
                • SEQadmin2
                  New Genomics Technologies Take Aim at Long-Standing Limits
                  by SEQadmin2


                  Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                  We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                  ...
                  09-28-2026, 10:25 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by SEQadmin2, 10-22-2026, 11:13 AM
                0 responses
                31 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-22-2026, 10:48 AM
                0 responses
                22 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-15-2026, 10:52 AM
                0 responses
                39 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-14-2026, 05:39 AM
                0 responses
                45 views
                0 reactions
                Last Post SEQadmin2  
                Working...