    Dear all,

    I've been making Illumina truseq DNA library and I've noticed dimer/trimer peak on bioanalyzer (high sensitivity assay) after PCR amplification. Is there anyone seeing this kind of result before?



    To avoid this problem, I optimised PCR input, which was around 3ng total, and there was no dimer or trimer peak. Do you think it's gonna make less complex library? Is 3ng PCR input enough for final sequencing library?
    Attached Files
    Last edited by sehrrot; 06-07-2011, 06:16 PM.

  • #2
    Hey:
    Thanks for the reply on my separate post...
    Techniques and protocol discussions on sample preparation, library generation, methods and ideas


    I'm just going to clone that 550nt beast and let illumina (and everyone here) know what it is.

    I got around the issue by running 10 cycles of PCR, gel extracting the correct molecular weights, Quantifying via picogreen, and determined that 4 more PCR cycles would bring me up to the same amount of total DNA as I was seeing after the original 10 cycles--containing that beast 550 nt thing---
    That extra 4 cycles brought my DNA amount up to similar amounts (no surprise there), but the PCR did not contain the 550nt band.

    Off to cluster generation...

    Comment


    • #3
      Hey,

      How much amount of DNA are you starting with for librray prep?
      We had the same issue which was resolved by reducing the amount of adapter ligated dna into pcr.
      Also dilute the adapters accordingly. Try different ratios.

      Thanks!
      Ashwini

      Comment


      • #4
        I've taken this library prep through to completion with three different samples at this point, and I see that the ligation TIME is the screwy part...however I like the idea of decreasing adapter concentrations...
        However, I am getting this nasty Laddering phenomenon, whereby I am getting concatenated DNA...again, Ligase concentration is the suspected culprit, even though the DNA should not be self-ligate-able, nor the adapter ends.
        very strange.

        Thanks for the suggestion to dilute adapter.
        Take care.
        Z

        Comment


        • #5
          Originally posted by ashchin View Post
          Hey,

          How much amount of DNA are you starting with for librray prep?
          We had the same issue which was resolved by reducing the amount of adapter ligated dna into pcr.
          Also dilute the adapters accordingly. Try different ratios.

          Thanks!
          Ashwini
          Thanks Ashchin,

          I started off fragmentation with 2.4 ug (calculated by QuBit BR) in 120 ul and then purified with AMPure beads. Then, remeasured the concentration with QuBit HS and got 50ng/ul as the Truseq protocol recommends.

          I've tried different ligation time and also different PCR input several times. Different from the previous PE protocol, longer ligation time resulted in dimer issue. Also, too much PCR input did too. I strictly keep 10 minute ligation time and then put stop ligation buffer and also used 3-4 ng of input for amplficiatoin, which I am a bit worried about library complexity.

          Anyway, it might be interesting to sequence dimer library and non-dimer library.

          Comment

          Latest Articles

          Collapse

          • seqadmin
            Genetic Variation in Immunogenetics and Antibody Diversity
            by seqadmin



            The field of immunogenetics explores how genetic variations influence immune responses and susceptibility to disease. In a recent SEQanswers webinar, Oscar Rodriguez, Ph.D., Postdoctoral Researcher at the University of Louisville, and Ruben Martínez Barricarte, Ph.D., Assistant Professor of Medicine at Vanderbilt University, shared recent advancements in immunogenetics. This article discusses their research on genetic variation in antibody loci, antibody production processes,...
            11-06-2024, 07:24 PM
          • seqadmin
            Choosing Between NGS and qPCR
            by seqadmin



            Next-generation sequencing (NGS) and quantitative polymerase chain reaction (qPCR) are essential techniques for investigating the genome, transcriptome, and epigenome. In many cases, choosing the appropriate technique is straightforward, but in others, it can be more challenging to determine the most effective option. A simple distinction is that smaller, more focused projects are typically better suited for qPCR, while larger, more complex datasets benefit from NGS. However,...
            10-18-2024, 07:11 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by seqadmin, Today, 11:09 AM
          0 responses
          23 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, Today, 06:13 AM
          0 responses
          20 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 11-01-2024, 06:09 AM
          0 responses
          30 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 10-30-2024, 05:31 AM
          0 responses
          21 views
          0 likes
          Last Post seqadmin  
          Working...
          X