    Dissociation analysis for RNAseq library

    Dear all,

    I've been using qPCR assay with dissociation analysis to check primer dimers or other undesired fragments. I found bird-nesting peak (red plot) at higher temperature and am wondering what this peak means. (c.f. one single peak in the red plot is the one we expected as library)

    Attached Files
    Last edited by sehrrot; 08-25-2011, 03:14 PM.
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    Originally posted by sehrrot View Post
    Dear all,

    I've been using qPCR assay with dissociation analysis to check primer dimers or other undesired fragments. I found bird-nesting peak (red plot) at higher temperature
    I don't understand. The red peak overlaps the lower temp green peak. When you say "higher temperature" in this context, what do you mean?
    Originally posted by sehrrot View Post
    and am wondering what this peak means. (c.f. green plot is the one we expected)
    I see a "shoulder" red peak at 84 oC. Is that what you are looking for?

    I have not done much with melting curves. What temp would one expect >100 bp dsDNA to melt at? I would have thought <90 oC.

    --
    Phillip

    Comment

    • sehrrot
      Member
      • Jul 2010
      • 58

      #3
      Dear Phillip

      The peak around 79-80C indicates library. The one I am looking for is the peak (shoulder one) at 84C in the green peak. I just guess this is the fragment with high GC

      Comment

      • pmiguel
        Senior Member
        • Aug 2008
        • 2328

        #4
        Interesting. Do you have an example of one with a high amount of primer dimers?

        --
        Phillip

        Comment

        • sehrrot
          Member
          • Jul 2010
          • 58

          #5
          This is the first RNAseq samples I got. Before qPCR, I found primer dimer peaks at the bioA plot so repeated beads purification using 0.7:1 ratio (beadsNA library). Then I got primer-dimer-free library and supposed that there would be a single peak in dissociation curve analysis. But there's shoulder peak....

          Comment

          • pmiguel
            Senior Member
            • Aug 2008
            • 2328

            #6
            So you think that the higher melting temp peak is the primer dimer?
            Can you use another assay to confirm? Like running an RNA Agilent chip on the library after strand denaturation? (95 oC for 2 minutes followed by snap cool on ice until the sample is loaded.) If you run the library strand-denatured, the short amplicons (primer dimers) won't be able to "hide" by annealing to longer library molecules.

            --
            Phillip

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM
            • SEQadmin2
              New Genomics Technologies Take Aim at Long-Standing Limits
              by SEQadmin2


              Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

              We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
              ...
              09-28-2026, 10:25 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 10-22-2026, 11:13 AM
            0 responses
            28 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 10:48 AM
            0 responses
            20 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-15-2026, 10:52 AM
            0 responses
            38 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-14-2026, 05:39 AM
            0 responses
            44 views
            0 reactions
            Last Post SEQadmin2  
            Working...