    TruSeq DNA libraries of unusual size

    I'm preparing Illumina TruSeq DNA libraries from MNase digested DNA, and getting weirdly sized output.

    I start with MNAse-digested dsDNA that has been gel purified to contain 100-200 bp fragments, with a prominent peak at 147 bp. MNase-cleaved DNA should have 3' phosphates.

    I input 1 ug of this into the TruSeq DNA kit, performing end-repair, A-tailing, and adapter ligation according to the protocol.

    I run this on a gel. I'd expect peaks at 150 bp (no adapters ligated) and possibly 270 bp (adapters ligated). Instead I see very little material in the 100-400 bp range, but a smear between 500-800 bp with possible peaks near 500 and 800 bp.

    I can gel purify this large stuff and amplify it according to the TruSeq PCR protocol, producing a size XL library.

    Any ideas about what is going on here? Some kind of concatamerization? I was worried that the Ampure beads were size selecting against the relatively small MNAse digestion fragments, so I repeated the procedure substituting MinElute for Ampure, and got the same oversized library.

    Thanks for any and all input.

Latest Articles

Collapse

  • SEQadmin2
    Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
    by SEQadmin2



    Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
    ...
    07-09-2026, 11:10 AM
  • SEQadmin2
    Cancer Drug Resistance: The Lingering Barrier to Rising Survival
    by SEQadmin2



    Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

    There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
    07-08-2026, 05:17 AM
  • GATTACAT
    Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
    by GATTACAT
    Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
    07-01-2026, 11:43 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, 07-13-2026, 10:26 AM
0 responses
23 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 07-09-2026, 10:04 AM
0 responses
33 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 07-08-2026, 10:08 AM
0 responses
20 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 07-07-2026, 11:05 AM
0 responses
34 views
0 reactions
Last Post SEQadmin2  
Working...