    Covaris parameters to obtain ~200bp fragments?

    I am using the Covaris to shear my samples during sample prep. My target fragment size is ~200bp, however practice shears assessed on the bioanalyser have shown smears which are very spread out, with fragment sizes of 150-250bp making up only 30% of the total fragmented DNA. The parameters used are:

    Duty Cycle: 10%
    Intensity: 5
    Cycles/Burst: 200
    Time:180

    We have also tested variations of these parameters: 10%dc, 5i, 200cpb, 120s ; 20%dc, 5i, 200cpb, 90s ; 20%dc, 5i, 200cpb, 120s and all produce very similar results. Could anyone tell me the best parameters to use in order to get a greater peak at around 200bp with a narrower spread?
  • Hamid
    Senior Member
    • Sep 2009
    • 108

    #2
    Dear ssg1,

    The settings of 10%dc/5i/200cpb/180 seconds is the correct settings for generating 200bp fragments. The same setting is also used and cited in the Agilent SureSelect protocol.
    Please check the following:

    1. Make sure that you are using 1-5ug/120ul of DNA in the Covaris MicroTube with the AFA fiber.
    2. The shearing should be carried out with the water tank temperature between 4-8 degrees Celsius.
    3. please check to make certain that the water level in the tank during processing is as suggested in the shearing protocol http://www.covarisinc.com/pdf/pn_400056.pdf
    4. please analyze an aliquot of the sheared DNA prior to any purification or concentration step.


    If possible can you send or post a gel image of bioanalyzer trace.

    Thank you

    hamid

    Comment

    • ssg1
      Junior Member
      • Jan 2010
      • 2

      #3
      Dear Hamid,

      Thanks for the response, we are already using 3ug/100ul, with a tank temperature between 4-8 degrees Celsius and the appropriate water level in the tank. I have attatched an image of the bioanalyser trace below.
      Attached Files

      Comment

      • Hamid
        Senior Member
        • Sep 2009
        • 108

        #4
        Dear ssg1,

        The wider distribution that you are noticing is an aritifact of the bioanalyzer electropharogram, and it is due to the following:
        1. Please check your DNA concentration. Typically when we run an 1-2ul aliquot of sheared DNA ( 3ug/120ul)on the Bioanalyzer, we obtain FU units of around 100-150.
        2. Please increased your sample load volume on the Bioanalyzer. Based on what you are currently seeing, I would suggest doubling or tripling the sheared DNA volume you are loading on the chip.
        3. Alternatively, you can always run an aliqote on a 2% agarose TBE gel, and post stain with EthBr.

        Thank you

        Hamid

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
          by SEQadmin2



          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
          ...
          10-14-2026, 09:55 AM
        • SEQadmin2
          New Genomics Technologies Take Aim at Long-Standing Limits
          by SEQadmin2


          Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

          We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
          ...
          09-28-2026, 10:25 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 10-22-2026, 11:13 AM
        0 responses
        49 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 10:48 AM
        0 responses
        37 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-15-2026, 10:52 AM
        0 responses
        46 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-14-2026, 05:39 AM
        0 responses
        46 views
        0 reactions
        Last Post SEQadmin2  
        Working...