Seqanswers Leaderboard Ad

Collapse

Announcement

Collapse
No announcement yet.
X
 
  • Filter
  • Time
  • Show
Clear All
new posts

  • Comment on my cDNA lib electropherograms - new to this

    Hi there,

    Just starting to make cDNA libraries and not sure how the electropherogram should look when I do fragment size analysis with Bioanalyzer High sensitivity DNA chip.

    So I used the KAPA Stranded mRNA-seq kit. Fragmentation is thermal, selecting for 200-300bp fragments.

    Goal is DE between drug treatments compared to control. Right now just troubleshooting to make sure our library prep is up to par before doing the real libraries.

    Sample 2: 2ug total eukaryotic RNA from mouse primary neuron culture with 10x PCR cycles
    Sample 4: 1ug same type of RNA with 10x PCR cycles
    Sample 5: 1ug same type of RNA with 8x PCR cycles

    My feeling is that Samples 2 & 4 are just over amplified/too much cDNA loaded for the high-sensitivity DNA chip.

    Please give me feedback on my electropherograms, would be very greatful for any positive or negative feedback.

    Cheers.
    Attached Files

  • #2
    Sounds reasonable. If your PCR reaction becomes depleted for reactants, then you end up with "bubble products". That is, the ends of the amplicons are all basically the same sequences, so they can anneal even if their inserts don't match. So after denaturation if no product strand is created, then the single-stranded amplicon will likely run into another single-stranded amplicons and their ends with hybridize. These "bubble products" run strange on bioanalyzer chips -- aberrantly large for their molecular weight.

    --
    Phillip

    Comment


    • #3
      Thanks for the response! Do you find though that the downstroke of the curve should be more abrupt similar to the upstroke of the curve?

      Comment


      • #4
        Dear PhDstudent,

        I am a Sr. Support and Applications Scientist with Kapa Biosystems. After reviewing your BioAnalyzer traces, it appears that your libraries may be slightly overamplified, most noticeably in Sample 2. With future samples, I recommend decreasing the amplification by a cycle or two.

        However, it is probable that the root cause of the bimodal peaks observed in Samples 2 and 4 are a result of sample overloading on the High Sensitivity chip. Increasing the dilution factor prior to assessment on the BioAnalyzer should alleviate this issue.

        I hope you find this information useful.

        Best,

        Jennifer Pavlica

        Comment

        Latest Articles

        Collapse

        • seqadmin
          Essential Discoveries and Tools in Epitranscriptomics
          by seqadmin




          The field of epigenetics has traditionally concentrated more on DNA and how changes like methylation and phosphorylation of histones impact gene expression and regulation. However, our increased understanding of RNA modifications and their importance in cellular processes has led to a rise in epitranscriptomics research. “Epitranscriptomics brings together the concepts of epigenetics and gene expression,” explained Adrien Leger, PhD, Principal Research Scientist...
          04-22-2024, 07:01 AM
        • seqadmin
          Current Approaches to Protein Sequencing
          by seqadmin


          Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
          04-04-2024, 04:25 PM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by seqadmin, 04-25-2024, 11:49 AM
        0 responses
        19 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-24-2024, 08:47 AM
        0 responses
        17 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-11-2024, 12:08 PM
        0 responses
        62 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-10-2024, 10:19 PM
        0 responses
        60 views
        0 likes
        Last Post seqadmin  
        Working...
        X