Seqanswers Leaderboard Ad

Collapse

Announcement

Collapse
No announcement yet.
X
 
  • Filter
  • Time
  • Show
Clear All
new posts

  • #31
    Thank you all!! I was bit hesitant to post,until no option was left with me..
    Now I feel that lot many queries can be resolved by open discussions.

    About my library type..it has been genomic and metagenomic!!
    The one that generated 20% less cluster when quantitated with qPCR was metagenomic library!!I agree with Phillip...multiple reasons prop up while troubleshooting!!
    Conclusions were drawn that qPCR were set up incorrectly giving faulty concentration(R^2 values were 0.998 &E=85%).Later I came to know that the concentration that was taken for clustering was presumed by GA results and also a Hiseq Run(in which qPCR was not run for QC) but had given optimum cluster nos. that had been shooted for!!

    The second set of libraries were from some client,so don't know about the sample type.
    This behaved so weirdly!!Our Kapa sybr fast was over so decided to use ABI Sybr fast for qPCR.The concentrations were very high..so we repeated to check out for handling errors!!But second time concentration turned out to be even higher....got trapped in endless confusions.Then borrowed a Kapa Sybr fast to repeat the same libraries and all concentrations fell in place.Then concluded that ABi sybr fast was behaving differently with Illumina libraries so now we have stuck to Kapa!!

    Now with all these discussions can we draw a conclusion??Can anyone summarize the do's and don'ts for qPCR quantitation and what else should be taken care of to get accurate results??

    Also one more question is there any data available for cluster generated for PhiX standard dilution?Thanks

    Comment


    • #32
      Maybe here somebody has seen something similar like my results from the BioAnalyzer run following RNA library prep (and pcr) and can tell me where the huge peak comes from and how to proceed now?
      Attached Files

      Comment


      • #33
        That's your library. It looks like you slightly under fragmented. Maybe next time increase length of the fragmentation incubation. We saw this with our NEB Ultra RNA prep's. It still sequenced OK, but it's slightly harder to quantify.
        Not too sure about the shaper HWM peaks. Did you use any in-line controls?

        Comment


        • #34
          Originally posted by TonyBrooks View Post
          That's your library. It looks like you slightly under fragmented. Maybe next time increase length of the fragmentation incubation. We saw this with our NEB Ultra RNA prep's. It still sequenced OK, but it's slightly harder to quantify.
          Not too sure about the shaper HWM peaks. Did you use any in-line controls?
          Thank you!

          So, this means that I would have to do a whole new library prep or can I re-fragment these samples?

          I used the in-line controls for end-repair, a-tailing and ligation but as far as I understood, I can only check if they are present after sequencing?

          Comment


          • #35
            We saw this before with the in line controls
            Techniques and protocol discussions on sample preparation, library generation, methods and ideas


            The library will sequence OK as is. Quantification may be more of an issue as it's harder to pick the correct size to normalise to. If you want to improve the library and get better quantification, you'll need to reprep it. You can't fragment it now as the adapters have already been ligated.

            Comment


            • #36
              Tony already covered the main point. Lesser point, the secondary bulge at a higher molecular weight likely comprises bubble products. But these are also only really an issue for accurate titration of samples prior to clustering.

              --
              Phillip

              Comment


              • #37
                Okay, thanks!
                Our technician suggested running a NuSieve gel and re-purifying with beads.
                I'm not entirely sure if that will help any but I'll probably try and then re-do the libraries if that doesn't help.

                Comment


                • #38
                  Hi all,

                  Back again after three years of peaceHope people are still responding to the queries here.We have switched completely to Kapa Standards now.Assays working completely in sync except for few variability here n there.A quick question….

                  Have anyone of you experienced a variation in qPCR of cDNA Illumina libraries?

                  Has anyone seen a difference of two times in qPCR concentrations,when re quantitated after freezing those libraries for two months and then compared with its own qPCR concentration,which was done at the time when the libraries were made?

                  Thanks

                  Comment


                  • #39
                    No.
                    We, are not re-doing the qPCRs and have never seen loading problems.

                    Comment


                    • #40
                      Thanks Luc,

                      Waiting for others to respond as well?

                      Do any one get request on re sequencing same samples from their clients?Did anyone redo qPCR's on cDNA libraries to check its integrity over a period of few months?

                      Comment


                      • #41
                        We have the occasional request for sample resequencing, and often see a significant decrease in library concentration upon requantification. Nucleic acids adhere to plastic during freezing (a well-known phenomenon), thus the recommendation for using Lo-Bind or similar tubes. The effect is negligible at high concentration, but a real problem at the concentrations typical of sequencing libraries.

                        Comment


                        • #42
                          Hi HESmith,

                          Thanks that you took time out to reply!

                          That's what I would have expected when I repeated the qPCR for same libraries after two months.But I got alarmed,when I saw that my cDNA libraries increased in concentration (just double) in comparison to the previous qPCR done.
                          That is why I reached Seqanswers because cannot think of any possible reasons or explanations.
                          Can anyone give their opinions and insights on this?

                          Comment

                          Latest Articles

                          Collapse

                          • seqadmin
                            Recent Advances in Sequencing Analysis Tools
                            by seqadmin


                            The sequencing world is rapidly changing due to declining costs, enhanced accuracies, and the advent of newer, cutting-edge instruments. Equally important to these developments are improvements in sequencing analysis, a process that converts vast amounts of raw data into a comprehensible and meaningful form. This complex task requires expertise and the right analysis tools. In this article, we highlight the progress and innovation in sequencing analysis by reviewing several of the...
                            05-06-2024, 07:48 AM
                          • seqadmin
                            Essential Discoveries and Tools in Epitranscriptomics
                            by seqadmin




                            The field of epigenetics has traditionally concentrated more on DNA and how changes like methylation and phosphorylation of histones impact gene expression and regulation. However, our increased understanding of RNA modifications and their importance in cellular processes has led to a rise in epitranscriptomics research. “Epitranscriptomics brings together the concepts of epigenetics and gene expression,” explained Adrien Leger, PhD, Principal Research Scientist...
                            04-22-2024, 07:01 AM

                          ad_right_rmr

                          Collapse

                          News

                          Collapse

                          Topics Statistics Last Post
                          Started by seqadmin, 05-14-2024, 07:03 AM
                          0 responses
                          20 views
                          0 likes
                          Last Post seqadmin  
                          Started by seqadmin, 05-10-2024, 06:35 AM
                          0 responses
                          44 views
                          0 likes
                          Last Post seqadmin  
                          Started by seqadmin, 05-09-2024, 02:46 PM
                          0 responses
                          54 views
                          0 likes
                          Last Post seqadmin  
                          Started by seqadmin, 05-07-2024, 06:57 AM
                          0 responses
                          42 views
                          0 likes
                          Last Post seqadmin  
                          Working...
                          X