Unconfigured Ad

Collapse
X
 
  • Filter
  • Time
  • Show
Clear All
new posts
  • IshSeq
    Junior Member
    • Jul 2014
    • 4

    Agilent Bioanalyzer vs AATI fragment analyzer

    Hello Seqers,

    We use the Agilent Bioanalyzer to perform QC on our DNA samples, pre and post library preparation.

    However, on several occasion we have had issues with this system, from data not being reproducible, one sample being affected by the other during the run and the fact that you have to use an entire chip even when there are only 5 samples to run (the chips and the reagents are quite expensive).

    Therefore, we are looking into a new system and are considering the AATI fragment analyzer.I would like to know if anyone has done a comparison of the bioanalyzer and the fragment analyzer and know how they stack against each other in terms of performance and cost.

    Thanks
  • NextGenSeq
    Senior Member
    • Apr 2009
    • 482

    #2
    The AATI Fragment Analyzer is more sensitive and can detect lower amounts of RNA and DNA.

    Illumina has switched to AATI from Agilent for all their in house library QC.

    Comment

    • der_eiskern
      Member
      • Jul 2009
      • 46

      #3
      re:

      i'm very concerned about AATI's inability to provide specs for their DNA quantitation abilities. Is it still recommended to use picogreen for library quantitation and only use microcapillaries for size range validation?

      Comment

      • NextGenSeq
        Senior Member
        • Apr 2009
        • 482

        #4
        We don't use the Agilent for quantification either. I do know that we can detect much lower amounts of Illumina libraries.

        Also, the AATI does not have an upper marker which interferes with Oxford Nanopore libraries. We make ONT libraries ranging from 8kB to 20kB and the 10 kB Agilent marker screws up the quantification on the Bioanalyzer.

        The AATI has a 100 kB marker. Since we are about to try to make 50 kB libraries this will be useful for sizing and quantification.

        Comment

        • der_eiskern
          Member
          • Jul 2009
          • 46

          #5
          Good to know!

          so you do use the AATI (not Agilent) for quantitation?

          Comment

          • NextGenSeq
            Senior Member
            • Apr 2009
            • 482

            #6
            ONT libraries are hard to quantify since the adapters have proteins tethered to them.

            The MinION is less sensitive to DNA quantity though since there is no cluster amplification or emulsion PCR.

            Basically if you have enough to see your library on the Bioanalyzer or AATI you have enough.

            Comment

            • northpaw
              Junior Member
              • Oct 2014
              • 8

              #7
              Illumina R&D has indeed transitioned away from chip-based systems and is now releasing new sample prep kits recommending the AATI Fragment Analyzer:



              Illumina recently announced Advanced Analytical as an automation partner for their Fragment Analyzer system:



              AATI Fragment Analyzer specs are available on their website:

              Comment

              • nucacidhunter
                Jafar Jabbari
                • Jan 2013
                • 1250

                #8
                Originally posted by northpaw View Post
                Illumina R&D has indeed transitioned away from chip-based systems and is now releasing new sample prep kits recommending the AATI Fragment Analyzer:



                Illumina recently announced Advanced Analytical as an automation partner for their Fragment Analyzer system:



                AATI Fragment Analyzer specs are available on their website:

                www.aati-us.com
                Your Illumina links are not working.

                Comment

                • northpaw
                  Junior Member
                  • Oct 2014
                  • 8

                  #9
                  Try these:





                  If links don't work you can find the Illumina Automation Partners page using Google.

                  The most recent technical note by Illumina recommending the AATI Fragment Analyzer is the TruSeq RNA Access Kit. They developed a new critical QC metric they call the DV200 using the Fragment Analyzer:

                  Google search:

                  [PDF]Evaluating RNA Quality from FFPE Samples - Illumina
                  res.illumina.com/documents/.../technotes/technote-truseq-rna-access.pdf
                  Technical Note: RNA Sequencing. Introduction. The TruSeq® RNA Access Kit provides an exon-capture, RNA-Seq approach for difficult samples such as RNA ...

                  Comment

                  • dannydesloover
                    Junior Member
                    • Nov 2012
                    • 5

                    #10
                    Aati

                    Had used the original system before the face-lift and thought it was great.

                    Now in a new role, just purchased the new AATI system and can't wait to put it through its paces.

                    Comment

                    Latest Articles

                    Collapse

                    • SEQadmin2
                      Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                      by SEQadmin2



                      Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                      ...
                      07-09-2026, 11:10 AM
                    • SEQadmin2
                      Cancer Drug Resistance: The Lingering Barrier to Rising Survival
                      by SEQadmin2



                      Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

                      There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
                      07-08-2026, 05:17 AM
                    • GATTACAT
                      Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                      by GATTACAT
                      Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
                      07-01-2026, 11:43 AM

                    ad_right_rmr

                    Collapse

                    News

                    Collapse

                    Topics Statistics Last Post
                    Started by SEQadmin2, 07-13-2026, 10:26 AM
                    0 responses
                    27 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 07-09-2026, 10:04 AM
                    0 responses
                    37 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 07-08-2026, 10:08 AM
                    0 responses
                    25 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 07-07-2026, 11:05 AM
                    0 responses
                    35 views
                    0 reactions
                    Last Post SEQadmin2  
                    Working...