  • Coyk
    Member
    • Jun 2011
    • 15

    Paired end library, 3kb woes

    Hello all,
    I have made two attempts so far to make it through the paired end library protocol (Rev. June 2010 is the manual I've been referencing). I have made it through the circularization adaptor ligation, library span size selection, and fill-in reaction (up to section 3.5). After that there is a quantification step that I've done twice and both times gotten 22ng/ul which is more than enough to proceed. Apparently, though, I tend to lose everything by the time I run it on a 7500 chip after circularization and nebulization. The two most obvious possibilities to me are either the circularization adaptors aren't ligating properly or the circularization isn't happening and the subsequent DNase step chews my sample to bits. I was wondering if this has happened to anyone else and if there are any suggestions for troubleshooting.

    Thanks!
  • MissDNA
    Senior Member
    • Nov 2010
    • 146

    #2
    I have prepared my fair share of 3kbPE libraries, and after circularization I NEVER saw a nice peak, as it shows in the manual, on the DNA 7500 chip. Most of the times I have an almost flat line. Still I continue with the protocol and most of the times I get a library because further on the protocol there is a library amplification step. Usually what I do is continue with 2 amplifications, instead of only one, and I combine the product of the 2 amplification in the final column purification step.
    This week I prepared 2 libraries like that.
    I really wonder if someone really sees nice peaks after circularization, and what can be done to improve circularization. Once per Roche´s tech support recommendation I increased the time of the circularization a bit. I went as far as 40 min only, but I am not sure if helped.
    Last edited by MissDNA; 09-30-2011, 12:34 PM.

    Comment

    • Coyk
      Member
      • Jun 2011
      • 15

      #3
      Well this gives me much hope! Thank you so much. I will take up the charge again and hope for the best.

      Comment

      • MissDNA
        Senior Member
        • Nov 2010
        • 146

        #4
        Cool. Let me know how it goes.
        The 2 libraries I prepared this week, titrated very well. Hopefully sequencing will be good.

        Comment

        • newmoon
          Member
          • Oct 2010
          • 11

          #5
          You should check Cre recombinase activity using the control in the kit. It's not a stable enzyme.

          Comment

          • MissDNA
            Senior Member
            • Nov 2010
            • 146

            #6
            I´ve used different lots of Cre recombinase and I always see an almost flat line after nebulization. However, as I stated before I end up with library. I never checked its stability and since it takes forever to get enzymes in Brazil is hard to keep changing lots.
            I sequenced the libraries I made last week. For one of them we got almost 700k reads and median read legth of 418, which is excellent for PE. The other library was bad (barely 10e08 molec./ul) but since we were in a hurry we sequenced anyway. We got about 400k reads with median lenght of 382, most of the reads failed in the mix+dot.

            Comment

            • Dun
              Member
              • Nov 2011
              • 16

              #7
              Same problem here. During three rounds of trying, I NEVER see any peak after circularization, nor after 500bp fragmentation (I use Covaris). Absolutely no peak.
              BTW: I didn't use their 'carrier DNA'

              Basically, I didn't see anything after circularzation until PCR, which contains 25 cycles and yields some fragments ranging from 100~700bp. Although I've made 3 PE libraries, I am just simply not trust them, so I didn't put them into emPCR.

              I desperately need some direction.
              How do you guys do QC for circularzation and final PE library? THX

              Comment

              • MissDNA
                Senior Member
                • Nov 2010
                • 146

                #8
                Why don´t you follow the protocol, using nebulization, carrier DNA and 15 cycles? Even though some libraries to not look perfect on the RNA chip, they tend to sequence well. I have prepared and sequenced over 30 3kb PE libraries, which good results.

                We sequenced another 3kb PE library today. I will see the results tomorrow morning. The library was definitely not the greatest.

                Could you post the agilent trace from your libraries?

                Comment

                • Dun
                  Member
                  • Nov 2011
                  • 16

                  #9
                  thx MissDNA

                  So it's OK to have flat line after circularization?

                  The story after that is a little complicated. since we are trying to put RLMID adaptors and use rapid library protocol, so the fragments I was mentioned are dsDNA.
                  I am wondering how does the DNA trace look like right after PCR in original assay.

                  Comment

                  • MissDNA
                    Senior Member
                    • Nov 2010
                    • 146

                    #10
                    I see. I have never tried that and I am not really sure that would work.
                    I have posted in some other thread, traces after nebulization where there is an almost flat line and then the final library in the expected size range. I´ve never seen a nice peak after nebulization.
                    Do you have the trace of an HS chip?
                    It would be actually interesting to know if your libraries would sequence well.

                    Comment

                    • MissDNA
                      Senior Member
                      • Nov 2010
                      • 146

                      #11
                      My traces are on this thread: http://seqanswers.com/forums/showthread.php?t=16183

                      Comment

                      • Dun
                        Member
                        • Nov 2011
                        • 16

                        #12
                        Many thx. I will check out that post carefully, seems it's a common problem.
                        I don't have HS report with me right now, I can show a pic of PCR product which is done by DNA 7500 chip.

                        I will throw my PE library in anyway later this month.
                        Attached Files

                        Comment

                        • Dun
                          Member
                          • Nov 2011
                          • 16

                          #13
                          My trace after Covaris 500bp fragmentation looks very similar to your nebulized ones.

                          Comment

                          • MissDNA
                            Senior Member
                            • Nov 2010
                            • 146

                            #14
                            When you say PCR product, do you mean the amplified library?

                            Your trace shows a lot of short fragments. My barely there peaks start raising at about 300 bp and end at around 1000 bp.

                            I would like to see your final library and know the results of your sequencing, especially if the RLMID worked.

                            Comment

                            • Dun
                              Member
                              • Nov 2011
                              • 16

                              #15
                              Yes, it is amplified library by PCR.

                              Comment

                              Latest Articles

                              Collapse

                              • seqadmin
                                New Genomics Tools and Methods Shared at AGBT 2025
                                by seqadmin


                                This year’s Advances in Genome Biology and Technology (AGBT) General Meeting commemorated the 25th anniversary of the event at its original venue on Marco Island, Florida. While this year’s event didn’t include high-profile musical performances, the industry announcements and cutting-edge research still drew the attention of leading scientists.

                                The Headliner
                                The biggest announcement was Roche stepping back into the sequencing platform market. In the years since...
                                03-03-2025, 01:39 PM
                              • seqadmin
                                Investigating the Gut Microbiome Through Diet and Spatial Biology
                                by seqadmin




                                The human gut contains trillions of microorganisms that impact digestion, immune functions, and overall health1. Despite major breakthroughs, we’re only beginning to understand the full extent of the microbiome’s influence on health and disease. Advances in next-generation sequencing and spatial biology have opened new windows into this complex environment, yet many questions remain. This article highlights two recent studies exploring how diet influences microbial...
                                02-24-2025, 06:31 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by seqadmin, Yesterday, 05:03 AM
                              0 responses
                              16 views
                              0 reactions
                              Last Post seqadmin  
                              Started by seqadmin, 03-19-2025, 07:27 AM
                              0 responses
                              17 views
                              0 reactions
                              Last Post seqadmin  
                              Started by seqadmin, 03-18-2025, 12:50 PM
                              0 responses
                              18 views
                              0 reactions
                              Last Post seqadmin  
                              Started by seqadmin, 03-03-2025, 01:15 PM
                              0 responses
                              185 views
                              0 reactions
                              Last Post seqadmin  
                              Working...