Hi everybody, I just performed an amplicon run to sequence 16S RNA using fusion primers and the "one way read approach". The amplicons have ~700bp and the bioanalyzer profiles were perfect with no primer-dimer peak. Guess what? I obtained a high % of short read/short primer, reads with low quality. The processing was made using amplicon and shotgun pipelines giving the same poor result. Amplicon sequencing using 454 shouldn't be allowed. I would appreciate any advise from you. Thanks in advance.
Comment
-
Originally posted by LFMar View PostHi everybody, I just performed an amplicon run to sequence 16S RNA using fusion primers and the "one way read approach". The amplicons have ~700bp and the bioanalyzer profiles were perfect with no primer-dimer peak. Guess what? I obtained a high % of short read/short primer, reads with low quality. The processing was made using amplicon and shotgun pipelines giving the same poor result. Amplicon sequencing using 454 shouldn't be allowed. I would appreciate any advise from you. Thanks in advance.
The trick here is that small amounts of primer dimer anneal to the 700 bp strands during PCR. PCR clean-ups get rid of short products, but not the short products annealed to long products. Each time you do a small-product removal cycle, some more of the short products denature and re-anneal to other short products and can be removed via size selection. Or at least that is my interpretation of what is happening.
--
Phillip
Comment
-
We routinely do 16s runs and they almost always look good. I agree with Philip that primer-dimer seems to be a major problem in all 454 amplicon work. All our amplicon libraries undergo gel-cut post pooling. The one library where we didn't do this gave us problems.
FYI I've attached results from our typical run. We use software version 2.5.3 with everything set as default. You have to remember that the Bioanalyser is a plot of DNA mass against size. What's important is not mass of DNA but molarity, so small peaks that are low in length are actually a big problem as they contain a high number of molecules. These are also favoured in emPCR over the longer amplicons you want to sequence. Also I don't think the Bioanalyser is sensitive enough to pick up any low level dimerisation.
It's might also be a good idea to also keep the results of the first NaOH wash during the bead enrichment step. This will contain the reverse strands of the emPCR which you can then run on a gel or Bioanalyser. You will probably need to concentrate on a Qiagen column first after neutralising with a small volume (40ul) of 3M sodium acetate (make sure the Qiagen indicator solution is yellow).
Comment
-
Originally posted by TonyBrooks View PostWe routinely do 16s runs and they almost always look good. I agree with Philip that primer-dimer seems to be a major problem in all 454 amplicon work. All our amplicon libraries undergo gel-cut post pooling. The one library where we didn't do this gave us problems.
FYI I've attached results from our typical run. We use software version 2.5.3 with everything set as default. You have to remember that the Bioanalyser is a plot of DNA mass against size.
Originally posted by TonyBrooks View PostWhat's important is not mass of DNA but molarity, so small peaks that are low in length are actually a big problem as they contain a high number of molecules. These are also favoured in emPCR over the longer amplicons you want to sequence. Also I don't think the Bioanalyser is sensitive enough to pick up any low level dimerisation.
It is worse than that. There is nothing to stop a primer dimer strand from annealing to a full length amplicon strand. So even if you get rid of the low molecular weight primer dimers quantitatively your sample may still be contaminated with primer dimers. Denature the strands and they may be there.
--
Phillip
Comment
-
TonyBrooks,
I’ve been wondering if our 16S runs could get better, and then I looked at the thumbnails you posted; ours are performing very similarly. I’ve had trouble in the past getting the enrichment low enough to even bother with sequencing, but between 0.01 and 0.05 cpb seems to be the sweet spot (after quantifying with the KAPA qPCR kit) for 8-20% enrichment.
After following this thread, I’m wondering if it would be feasible to denature the DNA immediately prior to the AMPureXP purification? Will the AMPureXP beads bind ssDNA, or would the RNA SPRI beads be an option? I’ve had some issues with small fragments that don’t show up until sequencing, and AMPure-ing denatured DNA seems like it might alleviate the problem, if it would work.
This is a great thread, and you guys have given some great information!! Keep it up!
Anthony
Comment
-
Thanks, Anthony. Good to know we are not over cautious. Also, that most of 454 users seem to have problems with enrichment. We do repeat a emPCR every now and then but certainly nothing close to half. Most of our titration does not perfectly correlate with MV/LV yet we seem to have a good guess-O-meter. It is harder to find the right cpb for MVs.
Comment
-
Originally posted by Anthony.287 View PostWill the AMPureXP beads bind ssDNA, or would the RNA SPRI beads be an option? I’ve had some issues with small fragments that don’t show up until sequencing, and AMPure-ing denatured DNA seems like it might alleviate the problem, if it would work.
Originally posted by MissDNA View PostOn the enrichment topic, do you guys sequence stuff that enriched over 20%? Most of the times we get enrichment over 20%, we repeat emPCR. Sometimes I wonder if we are being over cautious.
Comment
-
22% is not too far from recommended. The highest we have sequenced was 27% and as far as I remember there was not much difference from the other runs we had performed with the same library. Last week we perform a 4-region 3kbPE run. All the libraries were equally good, 3 of them had enrichment around 5-6% (we decided to repeat emPCR using SV, because originally we had gotten over 20% on MV), the 4th library was at about 14% (they were product of the 1st MV we had done). The 3 libraries with low enrichment had over 70% passed filter, while the one with 14% had only 49%.
I do think that passed filter has a lot to do with the libraries quality, still I am not comfortable to potentially ruin a run by using super enriched beads.
I would not run amplicons that had over 30% enrichment.
Comment
-
I normally don't run anything above 20%, but in that case, I had all the others ready to go and couldn't wait any longer. The one that was 22% was a little high, but as you said, it wasn't much above 20% so I went ahead with it. With the other one, I thought the results would probably be not so great, but because of what it was and what I needed out of it, I figured that would be okay. As it turned out, I got the information I needed in spite of the low filter pass rate.
Comment
Latest Articles
Collapse
-
by seqadmin
The field of epigenetics has traditionally concentrated more on DNA and how changes like methylation and phosphorylation of histones impact gene expression and regulation. However, our increased understanding of RNA modifications and their importance in cellular processes has led to a rise in epitranscriptomics research. “Epitranscriptomics brings together the concepts of epigenetics and gene expression,” explained Adrien Leger, PhD, Principal Research Scientist...-
Channel: Articles
04-22-2024, 07:01 AM -
-
by seqadmin
Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...-
Channel: Articles
04-04-2024, 04:25 PM -
ad_right_rmr
Collapse
News
Collapse
Topics | Statistics | Last Post | ||
---|---|---|---|---|
Started by seqadmin, 05-02-2024, 08:06 AM
|
0 responses
16 views
0 likes
|
Last Post
by seqadmin
05-02-2024, 08:06 AM
|
||
Started by seqadmin, 04-30-2024, 12:17 PM
|
0 responses
20 views
0 likes
|
Last Post
by seqadmin
04-30-2024, 12:17 PM
|
||
Started by seqadmin, 04-29-2024, 10:49 AM
|
0 responses
25 views
0 likes
|
Last Post
by seqadmin
04-29-2024, 10:49 AM
|
||
Started by seqadmin, 04-25-2024, 11:49 AM
|
0 responses
28 views
0 likes
|
Last Post
by seqadmin
04-25-2024, 11:49 AM
|
Comment