Commonly, I mix pools from various projects on a typical 500 cycle MiSeq run. Specifically, I have groups submitting 16S amplicon libraries constructed using either the Pat Schloss laboratory protocol or a TruSeq adapter 16S protocol I helped design. The former requires custom sequencing primers, while the latter does not. However, if I add 3.4 µL of a 100 µM stock of the appropriate custom primer to the MiSeq slots containing the standard Illumina primers, everything works fine for both types of libraries.
When I attempt the equivalent process on the Aviti, it fails in a puzzling way: the Schloss-type libraries work fine, but the other 16S amplicon libraries are significantly suppressed (30-50 times lower percentage yield than on the MiSeq with the same superpool).
It took me a number of runs to pin this down, but having done so I can see why those 16S custom primers might interfere with the sequencing of the TruSeq 16S amplicons. Both the TruSeq and the 16S custom primers can anneal to the TruSeq 16S amplicons. My question is why doesn't this cause a problem on the MiSeq?
--
Phillip
When I attempt the equivalent process on the Aviti, it fails in a puzzling way: the Schloss-type libraries work fine, but the other 16S amplicon libraries are significantly suppressed (30-50 times lower percentage yield than on the MiSeq with the same superpool).
It took me a number of runs to pin this down, but having done so I can see why those 16S custom primers might interfere with the sequencing of the TruSeq 16S amplicons. Both the TruSeq and the 16S custom primers can anneal to the TruSeq 16S amplicons. My question is why doesn't this cause a problem on the MiSeq?
--
Phillip