    lib prep using SPRIworks

    Hi,

    We have been recently experiencing that sometimes library preps done using SPRIworks show up with a larger fragment size than expected on the Bioanalyzer. And if we go ahead and sequence those samples, the G signal intensity is very low for those samples which eventually leads to fewer G calls and a poor alignment rate.

    I am wondering if anyone else has been experiencing something similar and may have more insight into the cause of this kind of phenotype.

    Thanks for your help in advance.
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    I don't think SPRIworks controls the upper fragment size. Maximum amplicon length then would solely be be determined your input fragment size.

    Are your input fragments larger than normal?

    --
    Phillip

    Comment

    • sgupta
      Junior Member
      • Nov 2008
      • 6

      #3
      Hi pmiguel,

      Thanks for your reply. We are seeing this phenotype predominately with whole cell extracts. Basically, the sample is split into 2 parts, one becomes the WCE, and the other goes through the ChIP. The IP looks fine after the library prep but the WCE's have a larger average fragment size.

      We ran the original WCE (before library prep/ligation of adaptors) on bioanalyzer also and as expected, it looks like a smear.

      Thanks again. Any inputs will be greatly appreciated.

      Comment

      • pmiguel
        Senior Member
        • Aug 2008
        • 2328

        #4
        You could do a size selection after SPRIworks to eliminate the larger amplicons.

        What size do you see currently?

        --
        Phillip

        Comment

        • sgupta
          Junior Member
          • Nov 2008
          • 6

          #5
          Hi Phillip,

          That is actually the problem. We expected the fragment size to be between 200 and 400 (based on size selection). But we see larger fragment size. So, SPRIworks seems to not eliminate the larger fragment size. Basically the size selection on SPRIworks does not seem to work well for WCE's for some reason and we were hoping someone else might have experienced the same this.

          Do you do library prep's for WCE using SPRIworks? If so, do you not see such a issue with WCE samples?

          Thanks

          Comment

          • pmiguel
            Senior Member
            • Aug 2008
            • 2328

            #6
            Originally posted by sgupta View Post
            Hi Phillip,

            That is actually the problem. We expected the fragment size to be between 200 and 400 (based on size selection).
            Do you mean your size selection or the SPRIwork's size selection?
            Originally posted by sgupta View Post
            But we see larger fragment size. So, SPRIworks seems to not eliminate the larger fragment size.
            What is the size distribution of the fragments you put into your SPRIworks System?
            Originally posted by sgupta View Post
            Basically the size selection on SPRIworks does not seem to work well for WCE's for some reason and we were hoping someone else might have experienced the same this.

            Do you do library prep's for WCE using SPRIworks? If so, do you not see such a issue with WCE samples?

            Thanks
            We do not have a SPRIworks System.

            But I am familiar with SPRI and am unaware of any SPRI-only protocol that does an upper size selection. That would mean that if you put large DNA fragments in, you will get large DNA fragments out.

            What size are your WCE fragments before you put them into your SPRIworks?

            --
            Phillip

            Comment

            • pmiguel
              Senior Member
              • Aug 2008
              • 2328

              #7
              Looks like I was wrong and the SPRIworks is supposed to do an upper size cut as well.

              You will likely need to consult with SPRIworks tech support.

              --
              Phillip

              Comment

              • sgupta
                Junior Member
                • Nov 2008
                • 6

                #8
                Hi Phillip,

                Thanks for your efforts. I will speak with SPRIworks tech support again. In the past when we have spoken to them, they say that the size selection is not 100% accurate and 20% of higher size fragments "bleed over" and if one does not have enough DNA in the selected size range, the 20% "bleed over" stuff gets amplified. But I find it hard to believe that WCE has nothing or very little in that range because we ran it on bioanalyzer before library prep. Plus it is essentially the same sample that has been split into 2 parts (IP and WCE) as I mentioned in my 2nd post on this thread.

                I wonder though, since you are familiar with the system, if you know how much of starting material is used for a WCE, for library prep on SPRIworks. If you do not know it is fine, I thought I ask.

                Thanks again.

                Comment

                • rnal
                  Junior Member
                  • Jul 2011
                  • 6

                  #9
                  I think you can contact with the technical support of illumina, they say the large fragment(the large peak in agilent test) is the bead contaminate, not the dna fragment.

                  Comment

                  • pmiguel
                    Senior Member
                    • Aug 2008
                    • 2328

                    #10
                    Okay, maybe you could post the electropherogram of the library from your bioanalyzer? Is there some reason you are steadfastly refusing to tell us the size range of your two libraries? You are making it very difficult to help you.

                    I don't actually know what the SPRIworks is doing exactly. But SPRI is a PEG cut, where molecules of a higher molecular weight are preferentially precipitated onto the surface of derivatized magnetic beads. So you can do one cut where you keep the stuff in solution and discard the precipitated stuff. That gets rid of the high molecular weight molecules. Then change the PEG concentration and precipitate fragments of the desired size on the beads and discard the lower molecular weight that remains in solution.

                    Solutes that effect the solubility of DNA in solution could conceivably shift your fragment size range under such a regimen. Does your WCE have a lot of salt in it? Maybe you could try desalting it prior to SPRIworks?

                    --
                    Phillip
                    Last edited by pmiguel; 07-23-2011, 08:18 AM.

                    Comment

                    • mgravina
                      Member
                      • Dec 2010
                      • 10

                      #11
                      Having similar problems

                      Our laboratory has been having problems that sound suspiciously similar to yours. Here's the link to the thread:
                      Bridged amplification & clustering followed by sequencing by synthesis. (Genome Analyzer / HiSeq / MiSeq)


                      I see that you were having this problem back in July. Have you resolved it yet?

                      I am also in Cambridge, at MIT - we may be neighbors.

                      Comment

                      • sgupta
                        Junior Member
                        • Nov 2008
                        • 6

                        #12
                        Update

                        We have discontinued the use of SPRIworks for now. I think it is a issue with the bead purification/size selection process and the amount of sample in the 200 to 400 region after sonication. Collectively, it leads to a significant loss of sample.

                        It may affect ChIP samples only (or the cause-outcome relation is more apparent) because underlying characteristics of the sample type is different from others given sonication steps, antibody quality etc.

                        As a solution, we have moved to manual preps with size selection using pippin prep until we find a better solution. At this point, we do not have much confidence in SPRIworks prep and will not use it until Beckman resolves this issue.

                        Comment

                        • dottomarco
                          Member
                          • Jul 2009
                          • 32

                          #13
                          Hi, we are now evaluating the SPRIworks System in order to do the library prep for our 454 and for HiSeq. I read your posts about the issues you faced with the system. We will be sequencing mostly exomes (TrueSeq/Nimblegen enriched), RNAseq and miRNA. Do you think the system works well with these types of samples or do you see this problems possibly affecting all the different types of libraries ? As you can immagine the main reason to use SPRIworks is to avoid labour intensive steps like size selection.. so if it is an issue with Beckman systems, I would not go for that. Does anyone know if they solved this issues? Any other feedback from SPRIworks users that can give some help regarding this choice would be greatly appreciated ! Thanks in advance to who ever could help

                          Comment

                          Latest Articles

                          Collapse

                          • SEQadmin2
                            Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                            by SEQadmin2


                            Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                            The systematic characterization of the human proteome has
                            ...
                            07-20-2026, 11:48 AM
                          • SEQadmin2
                            Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                            by SEQadmin2



                            Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                            ...
                            07-09-2026, 11:10 AM
                          • SEQadmin2
                            Cancer Drug Resistance: The Lingering Barrier to Rising Survival
                            by SEQadmin2



                            Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

                            There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
                            07-08-2026, 05:17 AM

                          ad_right_rmr

                          Collapse

                          News

                          Collapse

                          Topics Statistics Last Post
                          Started by SEQadmin2, 07-24-2026, 12:17 PM
                          0 responses
                          15 views
                          0 reactions
                          Last Post SEQadmin2  
                          Started by SEQadmin2, 07-23-2026, 11:41 AM
                          0 responses
                          15 views
                          0 reactions
                          Last Post SEQadmin2  
                          Started by SEQadmin2, 07-20-2026, 11:10 AM
                          0 responses
                          23 views
                          0 reactions
                          Last Post SEQadmin2  
                          Started by SEQadmin2, 07-13-2026, 10:26 AM
                          0 responses
                          37 views
                          0 reactions
                          Last Post SEQadmin2  
                          Working...