Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • Tali7
    Junior Member
    • Aug 2011
    • 8

    #1

    Help with smallRNA Truseq kit - Agilent results?

    Hi guys! I'm a newbie to this forum so please bear with me!

    I've prepared lots of Illumina libraries before (RADseq, RNAseq..) and am now doing miRNAseq using the TruSeq smallRNA kit, using total RNA input.

    I've got to the pre-gel purification stage (post-PCR) and have run an aliquot on the Agilent Bioanalyser as encouraged to do by the protocol. However, the on-site Bioanalyser isn't new enough to use the suggested High Sensitivity DNA chips, and I had to use a normal DNA chip. So basically I didn't really know what I was looking for in my results graphs.

    I do know that I am supposed to see a peak at ~145-160bp (the adapter-ligated miRNA size). And I do have small peaks in roughly that size region.

    HOWEVER, I also have maaaassive peaks at ~270bp. I have no idea what this could be, and nobody I've spoken to knows either! I know it isn't contamination, because a sample that is effectively my -ve control (don't ask) didn't have any peaks. I'm just concerned that, as the peaks are so big, the adapters (although designed to ligate to miRNA) have also ligated to bigger RNA too, and so have also been amplified in PCR.

    Would this be possible, and/or does anybody have any idea what these 270bp peaks are doing in my library prep?!

    Thanks so much
  • chaos81
    Member
    • Aug 2011
    • 10

    #2
    The most important thing is to have results in your target region. yes, you will get other ligated material, but after size selecting this should be primarily wiped out. As long as you see results in your target region then I would continue on with the protocol. and I use a DNA 1000 chip at that step and we have the capability of using a high sense chip. hope this helps.

    Comment

    • jpelletier
      Junior Member
      • Feb 2011
      • 5

      #3
      I have seen the same thing in my miRNA library prep too! I was told that it is probably adapter-ligated t-RNAs. They were indeed effectively removed by the gel size-selection.
      cheers!

      Comment

      • mnkyboy
        Member
        • Mar 2009
        • 87

        #4
        It is mainly tRNA. Run the BA again after you gel select.

        Comment

        • mboth
          Member
          • Oct 2010
          • 23

          #5
          I have done a few TruSeq small RNA library preps. You are adding ca 120bp of adapters/primers to your fragments, so the tRNA peak is at 180 - 190bp, not at 270. I suspect that 270bp represents rRNA. I also had a massive peak there, but after ribodepletion it was much smaller. Hope this helps

          Comment

          • pmiguel
            Senior Member
            • Aug 2008
            • 2328

            #6
            Or this is the common "bubble product/daisy chain" issue seen with TruSeq libraries. If so, you libraries are fine. Try denaturing them and running them on an RNA chip -- this will remove artifacts caused by ectopic reannealing of adapters. Here is what we do sometimes:

            Bridged amplification & clustering followed by sequencing by synthesis. (Genome Analyzer / HiSeq / MiSeq)


            --
            Phillip

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
              by SEQadmin2




              NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
              Today, 06:46 AM
            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 10-28-2026, 09:57 AM
            0 responses
            22 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-28-2026, 09:51 AM
            0 responses
            15 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 11:13 AM
            0 responses
            68 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 10:48 AM
            0 responses
            48 views
            0 reactions
            Last Post SEQadmin2  
            Working...