Unconfigured Ad
Collapse
X
-
Looks good.
If you want to see your RIN score you can check for errors in the error pane. It may be something like the "5S Region Anomaly Threshold" error. Then you can increase the 5S Region Anomaly Threshold from 0.5 to 1 and it should clear the error and give you a RIN score. But the RIN score will be unduly low due to the amount of small RNAs you have. So it is probably not worth the trouble.
--
Phillip
-
Plant RNA per se, will not necessarily give you poor RIN scores. We have a recent set of wheat RNAs submitted that all gave RINs > 9.5.
I think, other than actual degradation, the factor that will most lower the RIN score is a high concentration of small RNAs. That is typical of some isolation methods -- the Trizol (or other acid phenol) methods allow much higher yields of low molecular weight RNAs. For some applications that is actually good (well, obviously for small RNA projects). But for normal RNAseq experiments it can cause issues.
But, yes, it is not uncommon for plants to have a bunch of weird peaks (frequently plastid rRNA).
--
Phillip
Comment
Latest Articles
Collapse
-
by SEQadmin2
NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...-
Channel: Articles
10-30-2026, 06:46 AM -
-
by SEQadmin2
Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?
Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating...-
Channel: Articles
10-14-2026, 09:55 AM -
ad_right_rmr
Collapse
News
Collapse
| Topics | Statistics | Last Post | ||
|---|---|---|---|---|
|
Genome Sequencing Reveals Large, Repeat-Rich DNA in Trapdoor Spider Named for Stephen Colbert
by SEQadmin2
Started by SEQadmin2, 10-28-2026, 09:57 AM
|
0 responses
45 views
0 reactions
|
Last Post
by SEQadmin2
10-28-2026, 09:57 AM
|
||
|
Started by SEQadmin2, 10-28-2026, 09:51 AM
|
0 responses
26 views
0 reactions
|
Last Post
by SEQadmin2
10-28-2026, 09:51 AM
|
||
|
Started by SEQadmin2, 10-22-2026, 11:13 AM
|
0 responses
74 views
0 reactions
|
Last Post
by SEQadmin2
10-22-2026, 11:13 AM
|
||
|
Started by SEQadmin2, 10-22-2026, 10:48 AM
|
0 responses
48 views
0 reactions
|
Last Post
by SEQadmin2
10-22-2026, 10:48 AM
|
Comment