I accidentally set my index read as 10x as many cycles as I would like to sequence, and I am through most of my 1st read. Does anyone know how to change the parameters for the index read while a run is in progress?
Unconfigured Ad
Collapse
X
-
I understand it is too late for you now, but in principle changing paramaters mid-run is easy: you kill HCS, change parameters in the config files, restart. Since you discovered problem during cycle 1 and the barcodes are read at the end changing these parameters should create no issues. It is more tricky if you want to change parameters that are important early on later in the run. If you are still interested, let me know to post details about config files to be changed
Comment
-
Yes, MiSeq should be in the same boat. If you kill the run, it shouldn't write out to the RFID's so the cartridge and flow cell would still be useable. However, you'll have to change the recipe fragments so that when it goes to perform cluster generation it only does something mild like a PR2 delivery. MiSeq won't let you resume a run like HiSeq will.
Comment
Latest Articles
Collapse
-
by SEQadmin2
Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?
Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating...-
Channel: Articles
10-14-2026, 09:55 AM -
-
by SEQadmin2
Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.
We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing...-
Channel: Articles
-
ad_right_rmr
Collapse
News
Collapse
| Topics | Statistics | Last Post | ||
|---|---|---|---|---|
|
Started by SEQadmin2, 10-22-2026, 11:13 AM
|
0 responses
61 views
0 reactions
|
Last Post
by SEQadmin2
10-22-2026, 11:13 AM
|
||
|
Started by SEQadmin2, 10-22-2026, 10:48 AM
|
0 responses
43 views
0 reactions
|
Last Post
by SEQadmin2
10-22-2026, 10:48 AM
|
||
|
Started by SEQadmin2, 10-15-2026, 10:52 AM
|
0 responses
76 views
0 reactions
|
Last Post
by SEQadmin2
10-15-2026, 10:52 AM
|
||
|
Started by SEQadmin2, 10-14-2026, 05:39 AM
|
0 responses
74 views
0 reactions
|
Last Post
by SEQadmin2
10-14-2026, 05:39 AM
|
Comment