    SR Adapters on a PE flowcell?

    Does this work? Of course only for the first read...
    Or do you think colony formation will not work at all?

    Thanks
  • ETHANol
    Senior Member
    • Feb 2010
    • 308

    #2
    Use the search function and look through recent posts before asking a question. Thanks.
    --------------
    Ethan

    Comment

    • andibody
      Member
      • Mar 2008
      • 45

      #3
      Thanks for your comment but I haven't found the exact answer I'm looking for.
      I know single read adapters can not be used for paired end sequencing, the question is if I get only one read or if colony formation is totally inhibited.

      Comment

      • ein_io
        Member
        • Nov 2011
        • 14

        #4
        Indeed andibody

        I just asked this question yesterday.

        I also called our Seq. rep. She said that she could make a single read from a library made with PE adapters on a single read flow cell. But of course she cannot do PE seq on a single read flow cell.

        hope that helps

        let me know what you think

        ein_io

        Comment

        • Robby
          Member
          • Mar 2011
          • 68

          #5
          You can sequence the first read, but in the second read are of course just N's (we tested that). Furthermore you might have problems to recognize the Index-sequence => so it might be better to add an index at the beginning of the first read

          Comment

          • pmiguel
            Senior Member
            • Aug 2008
            • 2328

            #6
            Originally posted by Robby View Post
            Furthermore you might have problems to recognize the Index-sequence => so it might be better to add an index at the beginning of the first read
            Not sure what you mean here. TruSeq Index reads are possible (they do work) on SR or PE flow cells. Do you mean there is another non-inline indexing scheme that might not work?

            --
            Phillip

            Comment

            • andibody
              Member
              • Mar 2008
              • 45

              #7
              We had to fill up a PE flowcell and took therefore this SR library (small RNA adapters no Truseq).
              It didn't really work, very few, very small clusters. I was expecting to get the first read only for that sample as Robby mentioned...

              Comment

              • Robby
                Member
                • Mar 2011
                • 68

                #8
                As far as I understand the adapter-sequences, the difference between the small RNA adapter and the pe-adapter is just the P7-part. Because of that the bridge amplification can't be performed in the correct way, which causes small clusters. Also due to the different P7-part, just the P5-part can bind to the flowcell and the sequencing for the second read can't start. Nevertheless the sequencing primer and indexing primer can still bind.

                In theory I think the sequencing of a single end library with a paired end kit should work very bad, but in practice we observed that the sequencing of the forward reads works quite good (many reads in good quality), although the quality of the index/barcode read was very bad and the sorting was impossible.

                Comment

                • andibody
                  Member
                  • Mar 2008
                  • 45

                  #9
                  Great, thanks a lot, that's exactly what I wanted to know.

                  Best Wishes
                  Andi

                  Comment

                  • TonyBrooks
                    Senior Member
                    • Jun 2009
                    • 303

                    #10
                    As I understand it, the oligos on the SR flowcell are similar as on the PE flowcell, only shorter (at the 3' end). This means PE libraries can bind to SR flowcells and amplify, but SR libraries can't bind to PE flowcells (as the 3' end of the flowcell oligo is unbound to the library and therefore can't extend - or does so very poorly).
                    Read 1 sequencing oligo is the same for all applications (except small RNA).

                    Comment

                    • protist
                      Senior Member
                      • Jan 2009
                      • 101

                      #11
                      Tony is correct re the length differences - they can be seen in the original customer letter re adapters from 2008-2009. Also the linearisation reaction is different for PE versus SR flowcells. The single read P5 undergoes a periodate mediated linearisation whereas the PE P5 is enzymatic based. So when clustering the LMX mix from a PE cluster kit cannot be substituted for the mix in SR kit. I have attached a screen-shot from a very nice BROAD/Illumina presentation (https://www.broadinstitute.org/files...PrepSlides.pdf). PE libraries for PE/SR flowcells, SR libraries for SR flowcells.
                      Attached Files

                      Comment

                      • TonyBrooks
                        Senior Member
                        • Jun 2009
                        • 303

                        #12
                        So theoretically, it may be possible to do PE sequencing on a SR flowcell, as long as you use the PE cluster gen reagents? Or maybe the ribose residue (I'm assuming this is how the cleavage site is marked) used to mark the cleavage site in the SR flowcell would interfere with the enzymatic linearisation in the PE kit?

                        Comment

                        • ECO
                          --Site Admin--
                          • Oct 2007
                          • 1360

                          #13
                          Originally posted by TonyBrooks View Post
                          So theoretically, it may be possible to do PE sequencing on a SR flowcell, as long as you use the PE cluster gen reagents? Or maybe the ribose residue (I'm assuming this is how the cleavage site is marked) used to mark the cleavage site in the SR flowcell would interfere with the enzymatic linearisation in the PE kit?
                          Nope. The magic of the PE chemistry is a P5 primer that can be cleaved (via a uracil) and leave behind a cluster compatible primer for the paired end turnaround, in addition to the periodate-labile other primer. For PE seq you need to be able to linearize twice...at first cluster gen, then after the PE-turnaround.

                          Please correct me if I'm wrong.

                          Comment

                          • HESmith
                            Senior Member
                            • Oct 2009
                            • 512

                            #14
                            Originally posted by ECO View Post
                            Please correct me if I'm wrong.
                            You are 100% correct.

                            Comment

                            • DJ Flowcell
                              Member
                              • Apr 2013
                              • 13

                              #15
                              If I accidentally clustered PE libraries with a PE cluster kit on the SR flowcell, would the reads be good or bad? Let's forget about index and turnaround and second read, what would the first few cycles look like? Weird question but I'm troubleshooting a weird problem.

                              Thanks!

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM
                              • SEQadmin2
                                New Genomics Technologies Take Aim at Long-Standing Limits
                                by SEQadmin2


                                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                                ...
                                09-28-2026, 10:25 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, Yesterday, 10:52 AM
                              0 responses
                              9 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-14-2026, 05:39 AM
                              0 responses
                              19 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-13-2026, 09:56 AM
                              0 responses
                              16 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-09-2026, 11:14 AM
                              0 responses
                              27 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...