  • Maximum number of cycles from 300-cycle MiSeq kit...

    Just finished this run this morning...it was an attempt at 151x251 to determine just how many cycles are contained in the 300-cycle reagent cartridge.



    After a successful 308 cycle run (2x151,1x6index), there is quite a bit of reagents left over, mainly the incorporation, scan, and cleave mixes which is all that's needed to sequence and and image.

    The answer seems pretty clearly that at about cycle 360-370 (haven't looked at images yet), things fall off a cliff. So 2x175 would be pushing it but probably possible.

    Next time we have a run that fails early, I'm going to overfill a new reagent cartridge to see what's possible...2x200 seems really easy. 2x250 probably stretching it but we'll see!

  • #2
    Any quality scores available?

    Very cool!
    I wonder if you had any data on Q-scores during the run? From the Broad presentation (300bp) they ended at app 65% Q30, it would be interesting to see the drop off in your run too.

    Comment


    • #3
      Any reason to fear this could damage the instrument?

      Comment


      • #4
        miseq Q-scores

        Originally posted by hedas2303 View Post
        Very cool!
        I wonder if you had any data on Q-scores during the run? From the Broad presentation (300bp) they ended at app 65% Q30, it would be interesting to see the drop off in your run too.
        I also really would like to know about the Q-score at cycle150 or 200.

        Thanks advance.
        Cheers!
        Tanya

        Comment


        • #5
          Originally posted by krobison View Post
          Any reason to fear this could damage the instrument?
          No ill effects noticed yet.

          I wouldn't let it run dry and sit overnight...as (from reading the recipe xml files) it flushes the template position with incorporation buffer several times, likely to prevent template from drying on the sipper.

          Originally posted by Tanya12 View Post
          I also really would like to know about the Q-score at cycle150 or 200.
          I'll try to get the fastqc report up asap.

          Comment


          • #6
            Have you had a chance to get a fastqc profile?

            Comment


            • #7
              Yes! I just have to remember to post it! Setting reminder now!

              Comment


              • #8
                Finally dug it out...enjoy:

                Comment


                • #9
                  By the way, this run was from 12x libraries of 768 amplicons each (250bp long)...

                  Comment


                  • #10
                    Originally posted by ECO View Post
                    The answer seems pretty clearly that at about cycle 360-370 (haven't looked at images yet), things fall off a cliff. So 2x175 would be pushing it but probably possible.
                    About 372, perhaps? That's enough for 6 50bp runs with two barcode reads each (i.e. the newer dual-indexed Nextera barcodes).

                    Comment


                    • #11
                      Originally posted by gringer View Post
                      About 372, perhaps? That's enough for 6 50bp runs with two barcode reads each (i.e. the newer dual-indexed Nextera barcodes).
                      That would require hacking their RFID system...in the current form you can only use a cartridge once.

                      Comment


                      • #12
                        Any had luck hacking the RFID system?
                        I dont see any mention of it in the recipe XML files

                        Comment


                        • #13
                          Easiest way is just to break out the RFID chip (under the square "Illumina" pad on the top of the cartridge)...and when the scan fails, get a bypass code from icom.illumina.com. They are good for a week...I'm sure if you made a huge practice of this ILMN would crack down but it's easy enough for playing around.

                          Comment


                          • #14
                            I wonder if it would be possible to reuse the flow cell, not to run completely new sequencing, but to repeat sequencing with clusters already there. Sequencing errors are generated at random, so repeated sequencing will not generate the same dataset. Or a flow cell with the good cluster density of a good library could be used to obtain longer reads. Removing just synthesized strand should not be a problem as it is done anyway in paired end runs. I am not sure about paired end chemistry though, could the cell after a paired end run sequenced in the reverse order? Certainly the current RFID system disallows this. Any thoughts?

                            Comment

                            Latest Articles

                            Collapse

                            • seqadmin
                              The Impact of AI in Genomic Medicine
                              by seqadmin



                              Artificial intelligence (AI) has evolved from a futuristic vision to a mainstream technology, highlighted by the introduction of tools like OpenAI's ChatGPT and Google's Gemini. In recent years, AI has become increasingly integrated into the field of genomics. This integration has enabled new scientific discoveries while simultaneously raising important ethical questions1. Interviews with two researchers at the center of this intersection provide insightful perspectives into...
                              02-26-2024, 02:07 PM
                            • seqadmin
                              Multiomics Techniques Advancing Disease Research
                              by seqadmin


                              New and advanced multiomics tools and technologies have opened new avenues of research and markedly enhanced various disciplines such as disease research and precision medicine1. The practice of merging diverse data from various ‘omes increasingly provides a more holistic understanding of biological systems. As Maddison Masaeli, Co-Founder and CEO at Deepcell, aptly noted, “You can't explain biology in its complex form with one modality.”

                              A major leap in the field has
                              ...
                              02-08-2024, 06:33 AM

                            ad_right_rmr

                            Collapse

                            News

                            Collapse

                            Topics Statistics Last Post
                            Started by seqadmin, Yesterday, 06:12 AM
                            0 responses
                            19 views
                            0 likes
                            Last Post seqadmin  
                            Started by seqadmin, 02-23-2024, 04:11 PM
                            0 responses
                            67 views
                            0 likes
                            Last Post seqadmin  
                            Started by seqadmin, 02-21-2024, 08:52 AM
                            0 responses
                            75 views
                            0 likes
                            Last Post seqadmin  
                            Started by seqadmin, 02-20-2024, 08:57 AM
                            0 responses
                            66 views
                            0 likes
                            Last Post seqadmin  
                            Working...
                            X